Wanted: Instructions for Leybold PGA 100 mass spectrometer

AI Thread Summary
A user is seeking alignment instructions for the Leybold PGA 100 mass spectrometer, noting that the instrument requires alignment when components are swapped. Despite having the manual, the user is unable to find the necessary supplement online or through the manufacturer, as the device is outdated. Other participants in the discussion suggest searching for similar manuals and share links, but these are deemed unsuitable for the specific needs of the PGA 100. Ultimately, the user successfully repairs the instrument and adjusts the potentiometers, restoring functionality. The discussion highlights the challenges of sourcing information for older scientific equipment.
Michael Koch
Messages
7
Reaction score
8
I have a Leybold PGA 100 mass spectrometer, including the instruction manual. The instrument is paired with a gauge head. If either the instrument or the gauge head is exchanged, then an alignment is required. It's mentioned in the instruction manual there exists a supplement, which contains the alignment instructions. Does anybody have these alignment instructions? I'm ready to pay for a photocopy.
Thanks,
Michael
 
Engineering news on Phys.org
Why would you have to pay? Is it not available online? How about from the company?
 
I wouldn't ask here if I hadn't already tried Google and the manufacturer. It's an old instrument and out of service for many years.
 
Can you post a copy of the cover page of the manual you have, and also a shot of where is mentions the alignment supplement? That might help us to try to look for it. Also paging @dlgoff @davenn @jim hardy
 
Here is the cover page of the manual and the second page, where the required supplement is mentioned.
leybold_1.jpg
leybold_2.jpg
 

Attachments

  • leybold_1.jpg
    leybold_1.jpg
    23.1 KB · Views: 557
  • leybold_2.jpg
    leybold_2.jpg
    44.5 KB · Views: 569
  • leybold_1.jpg
    leybold_1.jpg
    23.1 KB · Views: 541
  • leybold_2.jpg
    leybold_2.jpg
    44.5 KB · Views: 581
  • Like
Likes davenn and berkeman
Michael Koch said:
I wouldn't ask here if I hadn't already tried Google and the manufacturer.

no probs, :smile: @berkeman had to ask the obvious, as a VERY high percentage of people ask here without ever doing any searches on even the most basic things

Michael Koch said:
Here is the cover page of the manual and the second page, where the required supplement is mentioned.

Thanks for that ...
I searched through a few old manual repositories that I am familiar with and didn't have any luck either
even searched youtube :rolleyes:

the only way you may get info is from some one who has one of these units

best of luck

Dave
 
  • Like
Likes berkeman and Michael Koch
Unfortunately these manuals aren't close enough. The leak detectors have a mass separator with a magnetic field, and they are tuned only for helium with mass 4. However the PGA 100 has a quadrupole separator and is tunable from mass 1 to 99. I have two PGA 100 instruments and one of them is defect. The measuring head was paired with the instrument that is now dead. When I use the working instrument with this head, then the mass spectrum is off by about one. The Argon peak is shown at 41, instead of 40 where it should be. The HF module has 14 potentiometers for adjustment. That's why I really need the instruction supplement. If there were only 2 or 3 potentiometer, I would simply try them out.
Michael
 
  • Like
Likes jim hardy
Michael Koch said:
The HF module has 14 potentiometers for adjustment.
Ouch! About 50 yrs ago I tuned a large Helium Leak Detector, which is rather simple... it took only a couple hours! You have a head start on yours, it's "only off by a little bit." :wink:

Try a search for Leybold PGA 100 using duckduckgo.com. https://duckduckgo.com/?q=leybold+pga+100
Of course the first hit is this thread! But if you read the results you will find an outfit in Germany that does servicing and further down the list there are several papers that mention they used that model in their research. As @davenn suggested above, hopefully you can find someone that is willing to send you a copy of the needed manual.

Good Luck! And keep us posted.
Tom
 
  • #10
I did already contact the servicing company in Germany last week, but got no reply so far. May be they don't want to share the manual.
 
  • #11
Good news. The repair was successful. The defect component revealed itself by a smoke signal.
Michael

IMG_9809.jpg
 

Attachments

  • IMG_9809.jpg
    IMG_9809.jpg
    55.6 KB · Views: 507
  • IMG_9809.jpg
    IMG_9809.jpg
    55.6 KB · Views: 481
  • Like
Likes berkeman and davenn
  • #12
I also figured out which of the 14 potentiometers must be adjusted for shifting the mass result. Now my two mass spectrometers are both working fine and have the peaks exactly where they should be.
Michael
 
  • Like
Likes berkeman, dlgoff and jim hardy
  • #13
Nothing like the sweet taste of success Congratulations! This is how we learn. Look how much more you know about your machine now.
 
  • Like
Likes dlgoff and Michael Koch
  • #14
Gotta love that smoke. :thumbup:
 
  • Like
Likes jim hardy and Michael Koch
  • #15
That's the 'good' thing about Tantalum capacitors, their failure is spectacular! (assuming a hefty enough power supply.)

Congratulations!
 
  • Like
Likes Michael Koch and dlgoff