The mechanical index (MI) of applied ultrasound is calculated using the formula MI = P_ra / √f, where P_ra is the peak rarefactional pressure and f is the frequency. The pressure is adjusted by an attenuation factor of 0.3 dB/(cm·MHz), which varies with depth and frequency. In a specific example using a 4 MHz pulse at 6 cm and 2 MPa, the MI is derived to be approximately 0.44 after accounting for attenuation. The normalization constant, C_MI, is crucial for accurate calculations, as it affects the units and final value.
  • Understanding of mechanical index (MI) calculations
  • Familiarity with acoustic pressure and intensity concepts
  • Knowledge of attenuation coefficients in ultrasound
  • Basic proficiency in logarithmic calculations (dB)
  • Study the role of normalization constants in ultrasound calculations
  • Learn about the effects of attenuation on ultrasound imaging
  • Explore the relationship between acoustic pressure and intensity in ultrasound
  • Review the paper "Tutorial paper: thermal and mechanical indices" for practical examples
Researchers, physicists, and medical professionals involved in ultrasound technology and its applications, particularly those focused on calculating mechanical indices and understanding acoustic properties.
rwooduk
On the surface, the equation is simply the peak rarefractional pressure divided by the root of the applied frequency:
##MI = \frac{P_{ra}}{\sqrt{f}}##
But the pressure is reduced/derated by an attenuation factor/coefficient that is dependent on depth and frequency e.g. ##0.3 \ dB / (cm \cdot MHz)##
An example calculation is given in the paper "Tutorial paper: thermal and mechanical indices". Where it takes a 4 MHz pulse, at 6cm, 2 Mpa, ##0.3 \ dB / (cm \cdot MHz)## and says the MI would be (2*0.44)/SQRT(4) = 0.44. I can't get the 0.44 because the dimensional analysis with units doesn't seem right.
##MI = \frac{MPa}{\sqrt{MHz}}\cdot \frac{dB}{cm\cdot MHz}##
I can easily calculate the attenuation, which is simply the attenuation coefficient multiplied by distance and frequency. Which for his example is 7.2 dB, I can also calculate intensity from this from I = I * 10^(-dB/10), but still don't get the 0.44.
Am I overcomplicating things?
jim mcnamara said:
This is probably some help - note their approach seems a little different from yours-
https://www.ncbi.nlm.nih.gov/pmc/articles/PMC5336845/
Thanks, I did see that paper but it seems they do not factor in the attenuation due to tissue / depth at all. I should have probably put this in the physics sections above as I think I'm having a calculation problem, rather than a definition problem. But I really appreciate the reply, thanks.
[Thread moved from the Medical forum to the Physics forums by a Mentor]
There is a normalization factor in the formula. If the factor is written in the denominator the units are ##\frac{MPa}{MHz^{1/2}}##. The formula I use is ##MI=\frac{p_{r.03}}{C_{MI}\sqrt{f}}## where ##C_{MI}## is this normalization constant. This is concerning the units.
For the value, the attenuation refers to the intensity. The intensity is proportional to the square of the acoustic pressure. So you have ##10 log(I_1/I_2)= 7.2 dB ## and ## 20 log(p_1/p_2)= 7.2 dB## so the pressure at 6 cm is derated to ##\frac{2 Mpa}{10^{0.36}} ## which is about 0.87 MPa. Divide this by ##\sqrt{4}## and you get about 0.44.
Edit.
In some papers they don't put the normaliation factor in the formula but it is assumed to be there.
It does not change the values.
Edit.
Thank you very much nasu, I had completely messed up my calculation. Extremely helpful!