Can Heating/Freezing Blankets Manage Extreme Temperature Testing?

  • Thread starter Thread starter acbel
  • Start date Start date
AI Thread Summary
Heating and freezing blankets may not be sufficient for the extreme temperature testing required for a small item measuring 2m x 1m x 1m, which needs to be tested at -50°C, ambient temperature, and +220°C. Users recommend specialized test chambers, such as those from Sun Electronics, which can handle significant temperature ramps and may accommodate both heating and cooling needs. Liquid CO2 models are suggested for achieving rapid cooling, while modifications to industrial refrigerators may be necessary for effective freezing. The discussion highlights the challenges of achieving reliable freezing solutions compared to heating. Overall, a combination of advanced test chambers and potential modifications to existing equipment is recommended for optimal testing conditions.
acbel
Messages
2
Reaction score
0
Hi guys, I'm new to this forum and first of all would like to say Hello to all of you.

I'm an R&D Engineer ( not long since I've graduated ) and I share a great interest in engineering therefore will be following the forum at regular basis.

Today I write to you with a problem I need to solve.

I need to test a relatively small item (2m length by 1m width by 1m height) however the testing needs to be done at different temperatures. The first one would be roughly -50deg C, second at ambient temperature and third at +220deg C. Now previously we have used heating fans for the heat test and liquid nitrogen for the freeze tests (all this wrapped up in a steel box with insulation).

My question is: what would you recommend to improve my tests, I've looked into heating blankets and industrial refrigerators however is there a device that would do both?

Are there heating/freezing blankets available or perhaps refrigerators that allow both positive and negative temperatures?
 
Engineering news on Phys.org
acbel said:
Hi guys, I'm new to this forum and first of all would like to say Hello to all of you.

I'm an R&D Engineer ( not long since I've graduated ) and I share a great interest in engineering therefore will be following the forum at regular basis.

Today I write to you with a problem I need to solve.

I need to test a relatively small item (2m length by 1m width by 1m height) however the testing needs to be done at different temperatures. The first one would be roughly -50deg C, second at ambient temperature and third at +220deg C. Now previously we have used heating fans for the heat test and liquid nitrogen for the freeze tests (all this wrapped up in a steel box with insulation).

My question is: what would you recommend to improve my tests, I've looked into heating blankets and industrial refrigerators however is there a device that would do both?

Are there heating/freezing blankets available or perhaps refrigerators that allow both positive and negative temperatures?

Welcome to the PF.

We use the test chambers from Sun Electronics in our R&D Lab:

http://www.sunelectronics.com/

We use the Liquid CO2 models to be able to support -10C/minute temperature ramps (we use +/-10C/minute temperature ramps in our product reliability testing). Our oven top out at 175C failsafe switches, though. But you might be able to get some models to go up to +220C (wow, that's hot!). :smile:
 
Hi Berkeman, Thanks for your quick reply,

I will have a look through their website. I've spend good few hours looking for ways of freezing and heating things up and as much as heating is relatively easy, freezing brings many problems. Most likely we will need to get one of the fridges (as large as possible) and modify it.
 
Thread 'How can I find the cleanout for my building drain?'
I am a long distance truck driver, but I recently completed a plumbing program with Stratford Career Institute. In the chapter of my textbook Repairing DWV Systems, the author says that if there is a clog in the building drain, one can clear out the clog by using a snake augur or maybe some other type of tool into the cleanout for the building drain. The author said that the cleanout for the building drain is usually near the stack. I live in a duplex townhouse. Just out of curiosity, I...