Emission spectrum changes with microscope objective?

In summary, the emission spectrum should definitely not change when I use different objective. However, I met with a very basic problem. When I use the same material (say plain silicon), and use different objectives (20X, 50X, 100X), then I will get totally different emission spectra. The spectra from each objective are consistent. My understanding is that the emission spectrum should definitely not change when I use different objective. This is driving me crazy. Any possible explanations?
  • #1
Yinxiao Li
46
0
I am using a Renishaw inVia confocal microscope to study emission spectrum of molecules. The excitation wavelength of laser that I use is 532nm. However, I met with a very basic problem. When I use the same material (say plain silicon), and use different objectives (20X, 50X, 100X), then I will get totally different emission spectra. The spectra from each objective are consistent. My understanding is that the emission spectrum should definitely not change when I use different objective. This is driving me crazy. Any possible explanations?
 
Physics news on Phys.org
  • #2
I don't know what is causing it, but here are some ideas to narrow it down:
What is the wavelength range of your spectra?
What do you get without any sample?
What do you get without laser?
 
  • #3
mfb said:
I don't know what is causing it, but here are some ideas to narrow it down:
What is the wavelength range of your spectra?
What do you get without any sample?
What do you get without laser?

The wavelength that I tested is 550nm-740nm. The excitation wavelength is 532nm.
Without any sample, I will only get some noise.
Currently I use 5% of laser power. If I change it to 0.00001% of laser power (close to without laser), I will only get noise.
 
  • #4
What is different about the emission spectra? How big of a change are we talking about?
 
  • #5
Yinxiao Li said:
I am using a Renishaw inVia confocal microscope to study emission spectrum of molecules. The excitation wavelength of laser that I use is 532nm. However, I met with a very basic problem. When I use the same material (say plain silicon), and use different objectives (20X, 50X, 100X), then I will get totally different emission spectra. The spectra from each objective are consistent. My understanding is that the emission spectrum should definitely not change when I use different objective. This is driving me crazy. Any possible explanations?

The lenses may have slightly different (spectral) transmissions. Do you have any information about the objective lenses?
I'm slightly familiar with the inVia Raman microscope and I looked up the spectrum of Silicon, but I can't easily parse the change from wavenumber shift to wavelength. Can you tell us the actual wavelength range you are looking at?

Edit- I just saw Post #3. Hang on...

Edit #2: Ok. I'd still like more information about the objective lenses, since you are looking out into the far red, past the wavelength range most lenses are chromatically corrected for.
 
  • #6
Different spectral transmissions, and probably different lensing behavior as well, if the lenses are not designed for this wavelength range. That could explain some trends. If completely new lines show up or vanish then things are more complicated. It would help to see example spectra.
 
  • #7
mfb said:
Different spectral transmissions, and probably different lensing behavior as well, if the lenses are not designed for this wavelength range. That could explain some trends. If completely new lines show up or vanish then things are more complicated. It would help to see example spectra.

I need to correct what I said before. When there is no sample, I can still get some signal with the same settings (5% power, 10s expose), although the signal is weaker.
I attach three example spectra here. Two of them correspond to the silicon substrate----one is absolute counts, and the other one is normalized. The third one correspond to no-sample based on absolute counts. Hopefully these will help!

For the objective lens, I am not sure how to check it. Maybe I should call the company for more information?
 

Attachments

  • compare_20X_50X_100X_1.jpg
    compare_20X_50X_100X_1.jpg
    33.1 KB · Views: 591
  • compare_20X_50X_100X_normalized.jpg
    compare_20X_50X_100X_normalized.jpg
    25.5 KB · Views: 622
  • nosample_compare_20X_50X_100X.jpg
    nosample_compare_20X_50X_100X.jpg
    25.3 KB · Views: 579
  • #8
Drakkith said:
What is different about the emission spectra? How big of a change are we talking about?
In my last post, I attach the example spectra. You can clearly see there is a big difference in peak. I would say the spectra are totally different.
 
  • #9
I don't know anything about how the whole system is set up. Is it possible you need to change the distance between different elements in the optical train when you change lenses?
 
  • #10
The two peaks around 560 nm are visible in all three spectra and they are at the same place, that is nice.
Also, 50 and 100 look very similar. 20 is odd, even without sample.
Yinxiao Li said:
For the objective lens, I am not sure how to check it. Maybe I should call the company for more information?
They certainly have some data on transmission and refractive index of their lenses as function of the wavelength. It could be in the documentation as well.
 
  • #11
Andy Resnick said:
The lenses may have slightly different (spectral) transmissions. Do you have any information about the objective lenses?
I'm slightly familiar with the inVia Raman microscope and I looked up the spectrum of Silicon, but I can't easily parse the change from wavenumber shift to wavelength. Can you tell us the actual wavelength range you are looking at?

Edit- I just saw Post #3. Hang on...

Edit #2: Ok. I'd still like more information about the objective lenses, since you are looking out into the far red, past the wavelength range most lenses are chromatically corrected for.

The objective lens is PLAN EPI--so it is designed for epi illumination. The 100X has a numerical aperture of 0.85 while 50X has a numerical aperture of 0.75. These information can be found directly on the body of the lens. They are the information you are looking for or not?
 
  • #12
Yinxiao Li said:
The objective lens is PLAN EPI--so it is designed for epi illumination. The 100X has a numerical aperture of 0.85 while 50X has a numerical aperture of 0.75. These information can be found directly on the body of the lens. They are the information you are looking for or not?

Not exactly- I was hoping for the manufacturer name and model number of each- Renishaw doesn't fabricate objective lenses.

I agree, there are significant differences between the 20X and the other two: the 50X and 100X could potentially be brought into agreement fairly easily with 'reasonable' spectral weighting. OTOH, if the 20X lens was specifically designed for far-red transmittance, that could also explain the difference- again, if you can provide the manufacturer and model number of each lens, that's how to start chasing down objective lens-dependent changes.

Now, here's a question since I don't know that much about Raman imaging: what is the angle and/or polarization dependence of the Raman-shifted light? I ask because the numerical aperture of the lenses are different, so you could be detecting differences due to changes in either incident angle and polarization state (which will vary azimuthally for a high NA lens)
 
  • #13
Andy Resnick said:
Not exactly- I was hoping for the manufacturer name and model number of each- Renishaw doesn't fabricate objective lenses.

I agree, there are significant differences between the 20X and the other two: the 50X and 100X could potentially be brought into agreement fairly easily with 'reasonable' spectral weighting. OTOH, if the 20X lens was specifically designed for far-red transmittance, that could also explain the difference- again, if you can provide the manufacturer and model number of each lens, that's how to start chasing down objective lens-dependent changes.

Now, here's a question since I don't know that much about Raman imaging: what is the angle and/or polarization dependence of the Raman-shifted light? I ask because the numerical aperture of the lenses are different, so you could be detecting differences due to changes in either incident angle and polarization state (which will vary azimuthally for a high NA lens)

The objective lenses are LEICA 566066 N PLAN EPI 20X/0.4, LEICA 566072 N PLAN EPI 50X/0.75, and LEICA 566073 N PLAN EPI 100X/0.85. Honestly I cannot find any useful information from these information...Do you have any clues?

To the best of my knowledge, there should be no angle or polarization dependence of the Raman-shifted light...
 
  • #14
Yinxiao Li said:
The objective lenses are LEICA 566066 N PLAN EPI 20X/0.4, LEICA 566072 N PLAN EPI 50X/0.75, and LEICA 566073 N PLAN EPI 100X/0.85. Honestly I cannot find any useful information from these information...Do you have any clues?

To the best of my knowledge, there should be no angle or polarization dependence of the Raman-shifted light...

Thanks for the info... Leica's database is currently down, tho... bleah...
 
  • #15
Yinxiao Li said:
The objective lenses are LEICA 566066 N PLAN EPI 20X/0.4, LEICA 566072 N PLAN EPI 50X/0.75, and LEICA 566073 N PLAN EPI 100X/0.85. Honestly I cannot find any useful information from these information...Do you have any clues?

To the best of my knowledge, there should be no angle or polarization dependence of the Raman-shifted light...

The Leica database is partially up and running today- still no spectral transmission curves- so I was only able to get partial information. It appears that the objectives require use of a #0 coverslip (as opposed to no coverslip or a #1.5 coverslip)- are you using those? Use of the proper coverslip is important because it is an optical element and the objective lens designers take it into account for the design of the lens. Incorrect coverslip thickness leads to uncorrected spherical aberration, which could account for your results.

I hesitate to create additional work, but if you are able to use 1 objective and 2 or 3 different coverslip thicknesses (say #0, #1.5, and no coverslip) you may see a difference.
 

1. How does the microscope objective affect the emission spectrum?

The microscope objective can affect the emission spectrum in several ways. First, it can affect the amount of light that reaches the sample, which in turn can affect the intensity of the emission spectrum. Second, the quality and design of the objective lens can affect the resolution and clarity of the spectrum. Lastly, different objective lenses may have different spectral properties, such as chromatic aberration, which can alter the emission spectrum.

2. Can the material of the microscope objective impact the emission spectrum?

Yes, the material of the microscope objective can have an impact on the emission spectrum. Some materials may absorb certain wavelengths of light, causing a decrease in the intensity of those wavelengths in the emission spectrum. Additionally, different materials may have different refractive indexes, which can affect the path and distribution of light, leading to changes in the emission spectrum.

3. Are there any specific objective lenses that are better for obtaining emission spectra?

The best objective lens for obtaining emission spectra depends on the specific experiment and sample being studied. Generally, a high-quality, low-magnification objective lens is preferred for obtaining a broad range of emission wavelengths. However, for more specific and precise measurements, a higher magnification lens may be necessary.

4. How can we minimize changes in the emission spectrum caused by the microscope objective?

To minimize changes in the emission spectrum caused by the microscope objective, it is important to use a high-quality objective lens that is well-suited for the sample and experiment. Additionally, proper calibration and alignment of the microscope can help reduce any distortions or aberrations in the emission spectrum. It is also important to use consistent and appropriate lighting conditions to minimize any external factors that may affect the emission spectrum.

5. Can the microscope objective affect the emission spectrum differently for different types of samples?

Yes, the microscope objective can affect the emission spectrum differently for different types of samples. This is because different samples may have varying characteristics, such as thickness, refractive index, and fluorescence properties, which can interact differently with the objective lens. It is important to consider the unique properties of each sample when choosing and using a microscope objective to obtain accurate emission spectra.

Similar threads

Replies
4
Views
2K
  • Other Physics Topics
Replies
3
Views
985
Replies
13
Views
1K
Replies
7
Views
1K
Replies
8
Views
2K
  • Other Physics Topics
Replies
18
Views
2K
  • DIY Projects
Replies
3
Views
2K
  • Materials and Chemical Engineering
Replies
1
Views
1K
  • Other Physics Topics
Replies
1
Views
1K
  • Introductory Physics Homework Help
Replies
2
Views
1K
Back
Top