Evaluate Used Wittig W2022 Scope - Ideas & Procedures

  • Thread starter Thread starter Windadct
  • Start date Start date
  • Tags Tags
    Scope
AI Thread Summary
The discussion focuses on evaluating a Wittig W2022 Scope acquired from eBay, emphasizing the need for thorough testing procedures to ensure its precision. Participants suggest obtaining a service manual for specifications and using a high-quality audio signal generator for performance testing. Recommendations include comparing DC accuracy with a digital voltmeter and assessing the scope's response to square waves. Additionally, testing standard signal diodes and analyzing square wave responses can help confirm the scope's functionality. The conversation highlights the importance of firmware and software in the performance of low-end oscilloscopes.
Windadct
Messages
1,452
Reaction score
411
Hello PF'ers...

I just picked up a Wittig W2022 Scope on E-bay.

I would like to give it a thorough evaluation, however, this is to be the highest precision item on the bench - does anyone have some good ideas or procedures for running this through its paces?

Perhaps even an eval form?

Thanks!
Paul
 
Engineering news on Phys.org
Hi Windact

have you got or have tried to get a service manual for it ?
if not you could try here ebaman

he has masses of manuals for all sorts of gear
A manual is at least going to give you the specs the scope should be able to achieve
(taking its age/amount of use into account)

a decent quality audio sig-gen/function generator with calibrated level output
would be an essential signal source for testing any scope's performance

cheers
Dave
 
Windadct said:
Hello PF'ers...

I just picked up a Wittig W2022 Scope on E-bay.

I would like to give it a thorough evaluation, however, this is to be the highest precision item on the bench - does anyone have some good ideas or procedures for running this through its paces?

Perhaps even an eval form?

Thanks!
Paul

You should have a DVM that is higher precision, at least for DC measurements and sine wave RMS measurements... :smile:

Compare the DC accuracy against your DVM using your DC power supply. Then use an AC square wave source to be sure that you can compensate the 'scope probes to flat signal response.

Does the 'scope have any measurement capabilities of its own, or is it all visual measurements made by you?
 
Yes - I'll download the manuals- I and many have some self cal, my DMM is OK - but a few features like bandwith would be tricky.

I was thinking of something like comparing VI curve for a few standard, signal diodes. Their performance should be pretty consistent. Many times when you look at cal you are checking each parameter at once, but IMO many of the digital devices, there are often one or two "tests" that can confirm that pretty much everything is working.

Another woud be Square Wave response - both unfiltered and then filtered. By overlapping results of different test - shortcomings of the test equipment can be ruled out and attributed to the UUT.

Just thinking out loud

Also - thanks for the firmware project - it looks interesting - probably the most common issue/complaint with low end scopes is the software, but in general all software projects are far more difficult then consumers ( and many manufactures) understand. So I want / Need the PC connection - but will see how well the OEMs works and then check the FW Project out.
 
Thread 'Weird near-field phenomenon I get in my EM simulation'
I recently made a basic simulation of wire antennas and I am not sure if the near field in my simulation is modeled correctly. One of the things that worry me is the fact that sometimes I see in my simulation "movements" in the near field that seems to be faster than the speed of wave propagation I defined (the speed of light in the simulation). Specifically I see "nodes" of low amplitude in the E field that are quickly "emitted" from the antenna and then slow down as they approach the far...
Hello dear reader, a brief introduction: Some 4 years ago someone started developing health related issues, apparently due to exposure to RF & ELF related frequencies and/or fields (Magnetic). This is currently becoming known as EHS. (Electromagnetic hypersensitivity is a claimed sensitivity to electromagnetic fields, to which adverse symptoms are attributed.) She experiences a deep burning sensation throughout her entire body, leaving her in pain and exhausted after a pulse has occurred...