How Does the Micsig DP2003 Handle High Voltage Measurements?

  • Thread starter Thread starter KyleGranger
  • Start date Start date
  • Tags Tags
    Differential Probe
AI Thread Summary
The Micsig DP2003 high voltage differential probe is rated for a maximum differential voltage of 5.6 kV and a common mode voltage of 1 kV, which can lead to confusion regarding its measurement capabilities. The common mode voltage refers to the average voltage of the two terminals with respect to ground, while the differential voltage is the voltage difference between the two probes. In practical scenarios, measuring high voltage systems, such as a 20 kV bus from capacitors in series, may exceed the common mode voltage limit, risking damage to the probe. Users need to understand that the differential voltage rating does not allow for measuring voltages that exceed the common mode specification without floating the oscilloscope. It is advisable to consult the manufacturer's data sheet for precise specifications and consider alternative probes if higher voltages are required.
KyleGranger
Messages
37
Reaction score
2
TL;DR Summary
I am trying to understand different ratings on a differential probe
Hi,

I am looking at a Micsig DP2003 high voltage differential probe. It's rated for 5.6 kV differential voltage. I understand that the differential votlage is measuring across two terminals which can be floating (not referenced to GND). The specs say it is rated for 1 kV common mode voltage, but I don't understand that. I thought the common mode voltage was difference in voltage between one terminal being measured and GND. I don't think my understand of this is right because if so, how would you ever measure >1 kV with your scope referenced to GND? If this is correct, does that mean I would have to float my scope to truly measure >1 kV differential?

Assuming my understand is incorrect, how do I determine the maximum voltage that I can measure? For instance, let's say that I have 10 capacitors in series with 2 kV across each. I would have a 20 kV bus. If I wanted to measure the voltage across the top capacitor, it would only have a differential voltage of 2 kV so would be fine by the DM voltage spec. But wouldn't the CM voltage be 20 kV and blow up? What's the highest cap that I can measure in this 10 capacitor string using this specific probe? A link to this probe with specs is below.

https://www.amazon.com/dp/B07DKT61LD/?tag=pfamazon01-20
 
Engineering news on Phys.org
KyleGranger said:
Assuming my understand is incorrect, how do I determine the maximum voltage that I can measure?
“Max differential test voltage (DC+AC PK-PK): 560V (200X); 5600V (2000X)”
“Max input common mode voltage: 1000V CATIII”
That suggests to me the specifications have been confused.

The differential voltage is the voltage between the two probes.
But that is also the common mode voltage, specified as 1000 V max.
So you will not be able to measure 2 kV across a capacitor.

So the 560V or 5600 V must be the voltage on either probe relative to the ground.
Maybe you need to go to the manufacturers data sheet to get the correct details.
http://micsig.com/html/41.html
 
Baluncore said:
“Max differential test voltage (DC+AC PK-PK): 560V (200X); 5600V (2000X)”
“Max input common mode voltage: 1000V CATIII”
That suggests to me the specifications have been confused.

The differential voltage is the voltage between the two probes.
But that is also the common mode voltage, specified as 1000 V max.
So you will not be able to measure 2 kV across a capacitor.

So the 560V or 5600 V must be the voltage on either probe relative to the ground.
Maybe you need to go to the manufacturers data sheet to get the correct details.
http://micsig.com/html/41.html
The unfortunate thing is their user manual says the exact same thing. What I am looking for is the isolation voltage I think. I've used these probes to measure about 1.25 kV differential and about 10 kV common mode (with respect to ground). It just hit me that I should check what the actual CM rating is before I push it any higher. I won't bash their manual here but it looks like I'll just have to find a different probe that lists the proper ratings. I was really hoping it was just me not understanding what I was reading.
 
Sorry, I didn't actually look at your device data. The thing to keep in mind with common mode voltages is that they are referenced to a common ground, which is normally the Earth ground of your oscilloscope or equivalent. So for two voltages V1 and V2 referred to ground, the common mode is Vcm = (V2+V1)/2. The difference mode is normally Vdm = (V2-V1)/2, but some assume it is just the difference V2-V1.
 
DaveE said:
Sorry, I didn't actually look at your device data. The thing to keep in mind with common mode voltages is that they are referenced to a common ground, which is normally the Earth ground of your oscilloscope or equivalent. So for two voltages V1 and V2 referred to ground, the common mode is Vcm = (V2+V1)/2. The difference mode is normally Vdm = (V2-V1)/2, but some assume it is just the difference V2-V1.
So if I understand you correctly, using the example of 10 series capacitors at 2 kV each for a 20 kV bus...if I were to measure the top capacitor one terminal would be 18 kV WRT GND. Each capacitor has a differential voltage of 2 kV so the 2nd terminal would be 20 kV WRT GND. You're saying the differential voltage, relating to ratings, would be only (20kV-18kV)/2= 1 kV?

As far a CM voltage, it would be (20kV+18kV)/2=18 kV?

That's all news to me because I would expect the DM voltage to be 2 kV and CM voltage to be 20 kV.
 
KyleGranger said:
So if I understand you correctly, using the example of 10 series capacitors at 2 kV each for a 20 kV bus...if I were to measure the top capacitor one terminal would be 18 kV WRT GND. Each capacitor has a differential voltage of 2 kV so the 2nd terminal would be 20 kV WRT GND. You're saying the differential voltage, relating to ratings, would be only (20kV-18kV)/2= 1 kV?

As far a CM voltage, it would be (20kV+18kV)/2=18 kV?

That's all news to me because I would expect the DM voltage to be 2 kV and CM voltage to be 20 kV.
The CM voltage is the average voltage wrt ground Vcm = (20kV+18kV)/2=19 kV.

The differential voltage (according to this definition) is Vdm = (20kV-18kV)/2= 1 kV, the voltage away from the common mode voltage.

The voltage difference is (20kV-18kV)= 2 kV.

Note that each individual voltages can be reconstructed from (Vcm + Vdm) and (Vcm - Vdm). So it's a nice linear decomposition to a different basis.

But these are just definitions that turn out to be useful in analysis (noise & differential amplifiers, mostly). So, make your own definitions if you like. But be aware that experienced EEs think you mean their "normal" version.
 
Thread 'Weird near-field phenomenon I get in my EM simulation'
I recently made a basic simulation of wire antennas and I am not sure if the near field in my simulation is modeled correctly. One of the things that worry me is the fact that sometimes I see in my simulation "movements" in the near field that seems to be faster than the speed of wave propagation I defined (the speed of light in the simulation). Specifically I see "nodes" of low amplitude in the E field that are quickly "emitted" from the antenna and then slow down as they approach the far...
Hello dear reader, a brief introduction: Some 4 years ago someone started developing health related issues, apparently due to exposure to RF & ELF related frequencies and/or fields (Magnetic). This is currently becoming known as EHS. (Electromagnetic hypersensitivity is a claimed sensitivity to electromagnetic fields, to which adverse symptoms are attributed.) She experiences a deep burning sensation throughout her entire body, leaving her in pain and exhausted after a pulse has occurred...