Hi all,
Having said the machine works, I’m now not so sure. Here it is, on the bench:
The third meter is labelled “Megohms at 1 kV ac”. The top scale runs from inf to 0.4 (Mohm) non-linearly, and the bottom runs from 0 to 2.5mA linearly.
This, I think, is measuring the insulation resistance between L+N and Earth for an appliance plugged into the test socket (off photo, to right). You press the button marked “FLASH...IR”, dial the second meter scale to 1 kV with the black knob, and read off the insulation resistance value from the third meter.
However, I can’t get the readings correct. If I measure the voltage between L or N and Earth with my multimeter, it reads 1000V ac, and the scale correctly reads the 10 M ohm internal impedance of the multimeter.
But putting, say, two 1 megohm resistors between L or N and Earth gives a reading of 1 Mohm. Three resistors gives 1.5 Mohm. It’s always half. If I twiddle the meter cal pot shown here to the right of the lower blue cap:
... I can alter the scale reading, but that throws off the 10 Mohm reading from the multimeter.
Also, with only 1 Mohm resistor between L or N and earth, the machine trips, suggesting a leakage current >5mA, when at 1 kV the leakage should be 1mA. From this, I assume there’s a problem with the measurement circuitry, rather than the meter itself. The very same resistor is measured correctly by my other 1 kV (DC) insulation tester. Is this an AC/DC thing? Peak voltage at 1 kV is 1414 kV. This doesn’t give enough extra current for the trip, but could the resistor be breaking down? They are standard 1/4W.
Another problem is the “set inf” pot under the meter is very jumpy - when you turn it either way, the needle jumps up before settling. Could this be a clue to the fault?
In summary: The voltage applied is measured correctly (1 kV), but the current measured and resistance calculated is off for values below 10 Mohm. The meter control board shown above tests OK on a component level, but the tripping is handled by the board which was the original subject of this thread. Here is the backside:
Legs of the suspect component (mystery diode) marked with some chalk. The frontside:
... has a meter cal pot for the kV scale (working), and a trip cal pot. The tripping at 1mA fault is out of range of the pot.
A long-winded post, for which apologies. Sadly there is no schematic, service manual or even user manual online.