The discussion centers on using the gamma constant to calculate the absorbed dose from a point source of photons. It highlights the relationship between exposure rate, dose rate, and the gamma constant, referencing a specific NRC document for guidance. The conversation also touches on the conversion fluence dose factor and its relevance in calculating dose per activity (Gy/Bq/hr) for brachytherapy sources. Participants suggest using ICRP Publication 119 and software like RadMap for cross-verification of results. The goal is to validate dose calculations against the TG-43 protocol, which has been updated in the AAPM Report #84.
khary23
Messages
92
Reaction score
6
It is unclear from my text. Can you use the gamma constant to find the absorbed dose for a point source of photons?
 
Engineering news on Phys.org
khary23 said:
It is unclear from my text. Can you use the gamma constant to find the absorbed dose for a point source of photons?
What text is one using? What is the context?

According to this document - https://www.nrc.gov/docs/ML1122/ML11229A688.pdf (see slides 15-33) - one can use a gamma constant, Γ, to calculate exposure rate dX/dt and exposure, X, and subsequently dose rate, dD/dt and dose, D.
 
I am using Turner's Atoms, Radiation and radiation protection.
I want to calculate the dose per activity (Gy/Bq/hr) of a point source of photons to theoretically check an MCNP result. The only way I could see to do it so that the units work, with the exception of the m^2, is with the gamma constant.
 
Last edited:
  • Like
Likes khary23
Are you referring to the mass attenuation coefficient?
 
no,
with H/phi factor,
with mass attenuation it is :
K (mGy/h)=5.74e-4*(mu/rho)*phi*E
same reference
 
I have not heard of the dose factor, but will look it up. What are its units?

Also edited my original post to clarify what I am doing.
 
Last edited:
the conversion fluence dose factor is in pSv.cm2
I Don't know "
gamma constant " but the "
conversion fluence dose factor" is defined in the ICRP
 
Yes, Use ICRP PUBLICATION 119.
For extra safety you can use another software like RadMap (radmap.tech) to cross check your result. It has a nice gui with a Geant4 backend and you can get a time limited free license.
 
  • #10
I assume that you want to calculate the dose in tissue due to a source within the tissue.
khary23 said:
I want to calculate the dose per activity (Gy/Bq/hr) of a point source of photons to theoretically check an MCNP result.
What MCNP result do you want to check. Your own simulation or another? Can you be more specific?
 
  • #11
KlasM said:
Yes, Use ICRP PUBLICATION 119.
For extra safety you can use another software like RadMap (radmap.tech) to cross check your result. It has a nice gui with a Geant4 backend and you can get a time limited free license.

Thank you for the suggestion. I will check it out!
 
  • #12
gleem said:
I assume that you want to calculate the dose in tissue due to a source within the tissue.

What MCNP result do you want to check. Your own simulation or another? Can you be more specific?

My own simulation. I am calculating the dose to water for a brachytherapy source to validate that it conforms to the TG-43 protocol.
 
  • #13
For your information the TG-43 report has been updated to the AAPM Report #84. I recommend you check it out.
 