RMS value secondary voltage electronic halogen transformer

  • Thread starter Merlion
  • Start date
  • #1
9
1

Main Question or Discussion Point

I am trying to derive a scaling factor for an analog voltmeter for the purpose of measuring the secondary voltage of an electronic halogen transformer (EHT).

http://www.ledbenchmark.com/faq/Transformers-Output-and-Compatibility.html

The output voltage of these things is a "high frequency" square wave (30-100kHz) with a 100 Hz (2x line frequency) envelope (see above link). Digital multimeters often have problems measuring this unless you opt for high bandwidth, True-RMS models with a considerable price tag, so why not try an analog meter.

The RMS-voltage specs of the DUT (Osram HTM70) is 11.5V. Peak of the envelope lies at 19V (checked this with an oscilloscope) which equates theoretically to a 0.605 Peak to RMS ratio, different from the 0,707 for a regular sine wave. Maybe my reasoning is flawed but I believe this can be explained by the gaps between the envelopes. The oscillator inside the transformer needs a minimum voltage to run.

Tried 3 different analog AC-volt/multimeters, including 2 wide-band audio VOMs one with a BW of 1MHz and all 3 show a 10V reading. Analog meters measure the average value and apply a scaling factor of Pi/2sqrt(2) to adjust the reading to the RMS value of a pure sine wave. When the voltage of the EHT passes through the rectifier inside the analog multimeter the average value must be the same, since the positive and negative part are symmetrical, at least to my understanding. I therefore expected the same reading, which is apparently not the case.

Where's the flaw in my logic?
 

Answers and Replies

  • #2
tech99
Gold Member
1,843
647
The analogue voltmeters probably respond to average voltage and use a scaling factor from average to RMS. You have already found that the scaling factor is wrong due to the wave shape.
To obtain a true reading you need a square law rectifier for the meter. This gives an output proportional to the mean square voltage and needs to be calibrated, which can be done using a sine wave.
 

Related Threads on RMS value secondary voltage electronic halogen transformer

  • Last Post
Replies
4
Views
1K
  • Last Post
Replies
1
Views
7K
  • Last Post
Replies
4
Views
23K
Replies
9
Views
5K
  • Last Post
Replies
6
Views
2K
  • Last Post
Replies
1
Views
2K
Replies
3
Views
2K
  • Last Post
Replies
3
Views
3K
  • Last Post
Replies
5
Views
19K
  • Last Post
Replies
2
Views
2K
Top