Thermal imaging and radiometry

AI Thread Summary
Thermal imaging cameras can detect bodies cooler than the detector if the signal from the target exceeds the noise level of the detector. The key factor is the signal-to-noise ratio; if the uncertainty in the detector's thermal signal is too high, it may mask the signal from a cooler object. Radiometers face a similar challenge, where their own thermal emissions can create a floor effect, complicating the detection of lower temperature bodies. Techniques like chopping and signal integration are employed to differentiate between instrument noise and actual thermal signals. Understanding these principles is crucial for effective thermal detection in various applications.
ArthurReader
Messages
1
Reaction score
0
Can a thermal imaging camera detect a body that is cooler than the detector? Or does the temperature of the detector put an absolute floor on the temperatures of objects that can be detected?

How do radiometers detect radiating bodies that are lower temperature than the radiometer? Or does the radiometer have a similar floor problem where the noise from the radiometer itself overwhelm the signal?
 
Science news on Phys.org
ArthurReader said:
How do radiometers detect radiating bodies that are lower temperature than the radiometer? Or does the radiometer have a similar floor problem where the noise from the radiometer itself overwhelm the signal?
it's all a question of signal to noise.
It's not the thermal signal from the detector that is a concern but the uncertainty in it.
If you know the detector had a signal of exactly 100units from it's own thermal emission an you measured a signal of 150units you know there was a target with a signal of 50units and so at a lower temperature.

But if the detector has a thermal signal of 100 +/- 20 then you wouldn't be able to detect a source that only had a signal of 20.
Because thermal emission is statistical for a given temperature there is a minumum uncertainty in that background.

in practice you do things like chopping (have the sensor look at the scene and then a at a fixed temperature reference) any difference in the two signal must come from the target scene, any signal that is the same is from the instrument. And integration to smooth out variations in the signal from the instrument.
 
I need to calculate the amount of water condensed from a DX cooling coil per hour given the size of the expansion coil (the total condensing surface area), the incoming air temperature, the amount of air flow from the fan, the BTU capacity of the compressor and the incoming air humidity. There are lots of condenser calculators around but they all need the air flow and incoming and outgoing humidity and then give a total volume of condensed water but I need more than that. The size of the...
Thread 'Why work is PdV and not (P+dP)dV in an isothermal process?'
Let's say we have a cylinder of volume V1 with a frictionless movable piston and some gas trapped inside with pressure P1 and temperature T1. On top of the piston lay some small pebbles that add weight and essentially create the pressure P1. Also the system is inside a reservoir of water that keeps its temperature constant at T1. The system is in equilibrium at V1, P1, T1. Now let's say i put another very small pebble on top of the piston (0,00001kg) and after some seconds the system...