X-Ray Shielding (Scanning Electron Microscope)

Click For Summary

Discussion Overview

The discussion revolves around X-ray shielding in scanning electron microscopes (SEMs), particularly focusing on the effectiveness of materials like aluminum and mu-metal in blocking characteristic radiation generated during electron interactions with specimens. Participants explore concerns about the adequacy of shielding in modified SEM chambers and the implications for safety protocols.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested

Main Points Raised

  • One participant expresses concern about the adequacy of aluminum and mu-metal coatings for shielding X-ray radiation generated at energies up to 10 keV, citing aluminum's half-value thickness.
  • Another participant emphasizes that the effectiveness of shielding depends on the configuration and integrity of the materials used, suggesting that a thorough inspection is necessary.
  • A third participant agrees on the importance of assessing radiation safety and recommends consulting health physics experts for radiation leak checks.
  • A later reply discusses the potential for X-ray generation up to 30 keV in SEMs and questions the dangers associated with such radiation, while acknowledging the need for professional detection tools.

Areas of Agreement / Disagreement

Participants generally agree on the need for careful consideration of X-ray shielding and the importance of consulting safety protocols. However, there is no consensus on the specific adequacy of the shielding materials or the calculations for radiation exposure in sieverts.

Contextual Notes

Participants note that the effectiveness of shielding can vary based on the configuration and potential gaps in coverage. There is also mention of the need for calculations based on maximum electron energy rather than just characteristic lines, indicating a complexity in determining safety measures.

Who May Find This Useful

Individuals working with scanning electron microscopes, health physics professionals, and those involved in radiation safety protocols may find this discussion relevant.

Karido
Messages
4
Reaction score
0
Hallo everyone,

I have a question about X-Ray shielding in a scanning electron microscope and I hope you can help me!
The topic is that when electrons hit the specimen, characteristic radiation up to 10keV is generated. This radiation is blocked/decelerated my the surrounding tower and chamber that are shielded with mu-metal (how thick is this coating). If I would just use pure Aluminium the X-Ray shielding wouldn't be enough. So far, is this correct?
Because Aluminium has a half-value thickness of about 112mm for 10keV x-radiation.
I am a bit concerned that some new-built parts of our SEM-chamber-cover aren't enough to shield the generated x-radiation. They are made out of alluminium with a thin mu-metal coating. It may be enough for magnetic field but I don't know if it is enough for the x-rays.. Can anyone tell me if my concerns are legitimate or is the X-radiation easily shielded with such layers?

Thank you for your help!
 
Physics news on Phys.org
Karido said:
Hallo everyone,

I have a question about X-Ray shielding in a scanning electron microscope and I hope you can help me!
The topic is that when electrons hit the specimen, characteristic radiation up to 10keV is generated. This radiation is blocked/decelerated my the surrounding tower and chamber that are shielded with mu-metal (how thick is this coating). If I would just use pure Aluminium the X-Ray shielding wouldn't be enough. So far, is this correct?
Because Aluminium has a half-value thickness of about 112mm for 10keV x-radiation.
I am a bit concerned that some new-built parts of our SEM-chamber-cover aren't enough to shield the generated x-radiation. They are made out of alluminium with a thin mu-metal coating. It may be enough for magnetic field but I don't know if it is enough for the x-rays.. Can anyone tell me if my concerns are legitimate or is the X-radiation easily shielded with such layers?

Thank you for your help!

It is most impossible to give you any accurate answer to this, because it also depends on the configuration. You can easily use shielding of an adequate thickness, but if the coverage is like swiss cheese, then it is of no use either. So without actually looking at what you have, it is almost impossible to say.

You have a very legitimate concern, and if your organization is following the proper safety protocol, they should have someone do a radiation survey before operation, especially when an instrument has been modified. Relay your concern to your supervisor, and request a radiation survey be performed. If you have a nagging feeling that something isn't quite right, do not ignore your instinct.

If you are working at a US Nat'l Lab, this is a valid "Stop Work" order before you continue working any further.

Zz.
 
  • Like
Likes   Reactions: Andy Resnick, Karido and M Quack
I agree that this has to be taken seriously. The information you provide is not sufficient to give a yes/no answer.

10 keV is not too hard, but it still has some penetration power. In case of doubt, ask someone from health physics to check for radiation leaks with a hand held detector. In general, shielding should be calculated for the maximum voltage/electron energy in the system, not for characteristic lines.

To get an estimate of the absorption of different materials you can look at this database:

http://henke.lbl.gov/optical_constants/filter2.html
 
  • Like
Likes   Reactions: Karido
Thanks for your answers! You helped me a lot!

I will check this out. But I have another theoretical question:

So in my SEM x-radiation up to 30keV can be generated because I accelerate my electrons with a voltage up to 30 keV.
If I have a aluminium plate that is 10mm thick, I have a transmission of 5% of the radiation. (http://web-docs.gsi.de/~stoe_exp/web_programs/x_ray_absorption/index.php)
Now I ask myself, how dangerous is x-radiation that has an energy of 30keV?
I know that it depends on the amount of the x-rays and the time you are exposed to them!
But can you tell me a quick overview about how to calculate an amount in sievert?
Because most of hand-held detectors start at a detectable energy level of 100keV.
Nevertheless I will check this with a professional tool that is able to detect x-rays at 10keV and higher energy levels if possible!
 

Similar threads

  • · Replies 8 ·
Replies
8
Views
2K
  • · Replies 2 ·
Replies
2
Views
3K
  • · Replies 10 ·
Replies
10
Views
2K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 7 ·
Replies
7
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 5 ·
Replies
5
Views
4K
  • · Replies 1 ·
Replies
1
Views
1K