w.shockley said:
thank you.
when I multiply the spectrum of the diluted solution, the absorbance is higher than the maximum (1).
how is it possible?
It sounds like you either didn't put a reference in your spectrometer or you used the wrong reference. Most absorption spectrometers work by using a "reference sample". One generally places two cuvettes in the spectrometer. One cuvette contains the unknown sample and one cuvette contains a reference material.
What one is generally looking for is not the absorbance of the entire unknown cuvette, but an absorbance of one particular component of the unknown sample. For instance, you may not be interested in the component of absorbance of the glass that the cuvette is made of. Therefore, you want to subtract the absorbance of the glass from the absorbance of the unknown cuvette. However, the reflectivity of the glass in the cuvette can also cause a false "gain" in the absorbance of your unknown sample. Therefore, one would want to fill the cuvette with a liquid of the same index of refraction as the index of refraction in your unknown sample.
This logic can continue. If want to know the absorbance of a solute molecule in water, you may not be interested in the absorbance of the water molecules. Therefore, you should fill your reference cuvette with water.
Most spectrometers automatically subtract the absorbance of the reference from the absorbance of the unknown. If the reference isn't chose appropriately, the uninteresting components of the reference will absorb more light than the uninteresting components of the unknown. That will read as a negative absorbance.
You may not need to take all he data again if that is your problem. You can measure the absorbance of an appropriate reference and subtract it from the measured absorbance of your unknown. That would eliminate the absorbance artifact in your absorbance data. You would be doing the arithmetic that the spectrometer does automatically.
One source of "gain artifact" would be if the width of the reference cuvette was narrower than the width of the unknown cuvette. Then, more light can pass by the reference cuvette than passes through the sample cuvette.
All cuvettes are not equal. Another cause of absorbance artifact is photoluminescence is a physical flaw in the cuvettes. Or maybe one of the cuvettes is dirty and the other is clean. Maybe you are using a glass cuvetter on the reference side and a silica cuvette on the sample side. Or maybe one cuvetter has glass one millimeter thick and the other cuvette has glass that is two millimeters thick.
You should chose cuvettes that are as similar as possible with the exception of the analyte that you are studying. It is okay to use a virtual reference if you have a very good idea of what the background in your unknown is going to be. By virtual reference, I mean you use theory to determine part of the background that you can't reproduce otherwise. However, it is usually better to find a physical reference that comes as close as possible to the background absorbance.