Evaporation of organic solvent and leftover residue

AI Thread Summary
The discussion centers on the effects of residual organic solvents after evaporation on the analysis of dry samples, particularly in thin-layer chromatography (TLC) and infrared (IR) spectroscopy. It is generally accepted that simple evaporation may not completely remove all solvent, especially if the compound does not crystallize. However, in most cases, residual solvents do not significantly interfere with spectral analysis unless specific solvents are used that could create overlapping peaks in the IR spectrum. For TLC, the solvent used for dissolving the sample typically does not pose a problem, as the compounds are analyzed from solution. Concerns arise primarily when residual solvents might be detected during spectral analysis, particularly if they could confuse the interpretation of functional groups. The consensus is that while complete removal of solvents is ideal, minor residues are often manageable and do not usually hinder analysis unless they leave an oily residue, which could cause significant interference.
~christina~
Gold Member
Messages
714
Reaction score
0
I have a question about evaporation of organic solvents and the leftover sample after this is done. (sample dissolved in the organic solvent)

Does the sample, leftover after the evaporation of a organic solvent, (used to dissolve the sample) affect the analysis of the dry sample? (basically, would there still be leftover organic solvent left, or does it all evaporate off? and thus not affect the sample analysis)

This question applies to two situations I have:

1. I have to run a TLC plate and I was wondering if the organic solvent used to dissolve the compound mattered, since you evaporate it off before you run the plate. Does it affect the run?

2. Would there be a residual leftover of organic solvent on the sample after evaporation that would affect the spectrum of the compound when run on a KBr pellet in IR spectroscopy?

Thank you.
 
Chemistry news on Phys.org
In general there should be no problems.

It wouldn't be very surprising if in some very rare cases simple evaporation will be not enough, but IMHO you will be very (un)lucky to have to deal with such a situation.
 
Borek said:
In general there should be no problems.

It wouldn't be very surprising if in some very rare cases simple evaporation will be not enough, but IMHO you will be very (un)lucky to have to deal with such a situation.

Thanks a lot Borek :smile:
 
~christina~ said:
Would there be a residual leftover of organic solvent on the sample after evaporation that would affect the spectrum of the compound when run on a KBr pellet in IR spectroscopy?

Thank you.

In my experience, and I'm a professional chemist, you almost always have residual organic solvent in your sample unless you take pains to remove it thoroughly. Simple evaporation almost never removes all of the solvent unless the compound crystallizes.
 
OK, I stand corrected :redface:

What I was aiming at was that in my experience (past experience I must add) solvent was never a problem when preparing spectra. Sure, if you are using spectroscopy to determine contaminants, you may easily detect solvent left, but - if that's the risk - you will never use this solvent in the first place. When you use solvent for extraction, or for reaction - it usually doesn't interfere.

Or is my experience abnormal?
 
You need to dry a compound for at least 24 hours over night in a vacuum oven if you want a clean NMR without any traces of solvent.
 
chemisttree said:
In my experience, and I'm a professional chemist, you almost always have residual organic solvent in your sample unless you take pains to remove it thoroughly. Simple evaporation almost never removes all of the solvent unless the compound crystallizes.

http://img5.imageshack.us/img5/4596/shockedb.gif Darn, I thought that it would leave a residue.

Situation: unknown compound(s) dissolved in solvent(s) and thus I have to evaporate solvent off before being able to analyse compound(s). I want to use TLC to run, solvent + compound to determine number of compounds in liquid. Then of course evaporate the solvent off if it's one compound and extract if it's more than one compound. Finally I want to, evaporate solvent off and make KBr pellet and run in IR spectrometer.

Note: Using AIR evaporation in fume hood

The big question is, (for anyone who can answer this)
1. Will it make it difficult to determine the spectra of an unknown compound if there is
residual solvent leftover, when analysing the spectra through KBr pellet method?

2. Does solvent used to dissolve compound and run TLC plate matter since it evaporates? (I
want to run the TLC plate preferably without using GC to find exact solvent first, to see
how many drugs are in unknown liquid)

Borek said:
What I was aiming at was that in my experience (past experience I must add) solvent was never a problem when preparing spectra. Sure, if you are using spectroscopy to determine contaminants, you may easily detect solvent left, but - if that's the risk - you will never use this solvent in the first place. When you use solvent for extraction, or for reaction - it usually doesn't interfere.
I don't know what the solvent is and I really haven't encountered having to evaporate liquid samples to obtain a powder form of a compound. This is the reason I don't know how a IR spectra will be affected by leftover solvent residue from just air evaporation.
gravenewworld said:
You need to dry a compound for at least 24 hours over night in a vacuum oven if you want a clean NMR without any traces of solvent.
Well the problem is that I don't have the availability to place something in an oven overnight and we don't have a vacuum oven, so I'm planning to use air evaporation in the fume hood.

Thanks
 
Last edited by a moderator:
Borek said:
OK, I stand corrected :redface:

What I was aiming at was that in my experience (past experience I must add) solvent was never a problem when preparing spectra. Sure, if you are using spectroscopy to determine contaminants, you may easily detect solvent left, but - if that's the risk - you will never use this solvent in the first place. When you use solvent for extraction, or for reaction - it usually doesn't interfere.

Or is my experience abnormal?


Christina asked about IR pellets which implies that she is working with a solid. You are correct that the residual solvent won't usually interfere with an IR determination. There are some exceptions I could come up with but I'd be reaching if I called that the usual experience. If a student were trying, for example, to make a ketone or carboxylic acid from an alcohol for the first time, he or she might be confused in their interpretation of an IR spectrum if solvent peaks are present. An alcohol (starting material) has an OH stretch that would interfere with the OH absorption of water if that was used in the workup - extraction. Likewise the carbonyl stretch of a solvent like ethyl acetate might be confused with the carbonyl stretch from the expected product likely producing two closely-spaced absorbances and leading to confusion in the interpretation.
 
~christina~ said:
The big question is,(for anyone who can answer this)
1. Will it make it difficult to determine the spectra of an unknown compound if there is
residual solvent leftover, when analysing the spectra through KBr pellet method?

For IR, probably not but be aware that you are usually only looking at certain functional groups in the IR. Keep in mind what type(s) of solvents you used that might interfere.

2. Does solvent used to dissolve compound and run TLC plate matter since it evaporates? (I want to run the TLC plate preferably without using GC to find exact solvent first, to see how many drugs are in unknown liquid)[/B]

I don't know what the solvent is and I really haven't encountered having to evaporate liquid samples to obtain a powder form of a compound. This is the reason I don't know how a IR spectra will be affected by leftover solvent residue from just air evaporation.
If the evaporation leaves an oily residue, expect significant interferences.

TLC is usually no problem since you are spotting them from solution anyway.
 
  • #10
chemisttree said:
For IR, probably not but be aware that you are usually only looking at certain functional groups in the IR. Keep in mind what type(s) of solvents you used that might interfere.
We are using the fingerprint region of the spectrum and comparing that with known fingerprint region spectra of known compounds in a database.

If the evaporation leaves an oily residue, expect significant interferences.

Define, "oily residue." I'm not sure how that would look like.

Thank you
 
  • #11
Like goo.
 
  • #12
Borek said:
Like goo.

i.e. not powdery?
 
  • #13
Generally - not solid. Most likely you will have a drop of something that looks like oil.
 
  • #14
Borek said:
Generally - not solid. Most likely you will have a drop of something that looks like oil.

okay, Thanks Borek. :smile:
 