Four Point Probe Measurement Help for 500 nm Ni Coating

  • Thread starter Thread starter physicistX
  • Start date Start date
  • Tags Tags
    Point Probe
AI Thread Summary
The discussion centers on a user's experience with four-point probe measurements of a 500 nm nickel coating on a glass substrate, which yielded a sheet resistance of 13.56 mohm/square at 5 mA. The user questions the accuracy of this measurement, as it results in a resistivity significantly lower than nickel's bulk value. Forum members suggest verifying the measurement setup, ensuring the probe configuration is appropriate for thin films, and confirming the thickness of the coating. They recommend conducting a manual measurement to cross-check results and using a multimeter to validate current and voltage readings. The conversation emphasizes the importance of proper equipment setup and measurement techniques in obtaining accurate results.
physicistX
Messages
3
Reaction score
0
Hi,

I am new to four point probe measurements and I have a question.

I've deposited 500 nm of Ni on glass substrate to check out the equipment we bought. When I measured its sheet resistance the display showed 13.56 mohm/sqare for 5 mA current.

To check out the measurement I multiplied the sheet resistance with the coating thickness which gave about 1/10 of Ni's bulk resistivity (96.3 nohm.meter). I couldn't figured it out if it's a faulty measurement or is there a logic I am missing. Any help would be appreciated.

thanks.
 
Physics news on Phys.org
Hi physicistX, welcome to PF. A few questions: is your test equipment set to analyze a thin conductive layer (instead of a bulk sample)? Are the length and width of the Ni film much larger than the spacing between the probes? Are you sure of the thickness?

You could also confirm the current and voltage with a multimeter. In fact, a four-point-probe measurement is quite easy to do yourself; the only advantage of the tool is the precise spacing of the probes.

Are you familiar with the equation used to turn current, voltage, and geometry into sheet resistance? See http://www.ecse.rpi.edu/~schubert/Course-Teaching-modules/A37-Four-point-probe-measurement-of-semiconductor-sheet-resistance.pdf" , for example.
 
Last edited by a moderator:
Could it be that your film is actually much thicker?

Have you measured at smaller currents and established linearity?

I don't see an error in your logic. The film resistivity for 500nm should be pretty close to the bulk value... possibly a little bigger.
 
Hi Mapes,

Thank you for your answer. I am quite sure of the thickness although I did not check it out after the deposition process but it was deposited with magnetron sputter equipped with a thickness monitor. The thickness monitor is healty and calibrated. So I am sure it is 500 nm. I checked the linearity of the resistance by using different currents. The current value I used is the current value where I could get stable voltage and sheet resistance.
 
Then I think a useful next step is to try the measurement yourself: hold four electrodes in contact with the sample (perhaps with a friend to help) and apply current, measure voltage, apply equation. Also, use a multimeter to confirm what your prober is doing. Finally, check the prober manual to see if you've set it up right. Good luck!
 
So I know that electrons are fundamental, there's no 'material' that makes them up, it's like talking about a colour itself rather than a car or a flower. Now protons and neutrons and quarks and whatever other stuff is there fundamentally, I want someone to kind of teach me these, I have a lot of questions that books might not give the answer in the way I understand. Thanks
I think it's easist first to watch a short vidio clip I find these videos very relaxing to watch .. I got to thinking is this being done in the most efficient way? The sand has to be suspended in the water to move it to the outlet ... The faster the water , the more turbulance and the sand stays suspended, so it seems to me the rule of thumb is the hose be aimed towards the outlet at all times .. Many times the workers hit the sand directly which will greatly reduce the water...