High voltage capacitor defibrillator

Click For Summary
SUMMARY

The discussion centers on the operational protocols and safety considerations of high voltage capacitor defibrillators, particularly Automated External Defibrillators (AEDs). Participants emphasize that defibrillators charge only when a shockable rhythm is detected via EKG, and the capacitor discharges if the patient stabilizes before defibrillation. The ZOLL AEDPlus model is mentioned, which charges in 2-3 seconds, highlighting the urgency in defibrillation scenarios. Suggestions for energy recovery systems are discussed, but safety concerns regarding accidental discharges during unnecessary charging are paramount.

PREREQUISITES
  • Understanding of Automated External Defibrillators (AEDs)
  • Knowledge of EKG rhythm analysis
  • Familiarity with high voltage capacitor operation
  • Awareness of defibrillation protocols and safety measures
NEXT STEPS
  • Research the operational protocols of the ZOLL AEDPlus
  • Explore the safety features of modern AEDs
  • Investigate energy recovery systems for medical devices
  • Study the implications of high voltage capacitor design in defibrillators
USEFUL FOR

Medical professionals, emergency responders, biomedical engineers, and students studying cardiac care and defibrillation technology.

ernd59
Messages
28
Reaction score
0
If defibrillation is not required is there an alternative method to save the energy for the next needed defibrillation rather than to discharge the capacitor using a resistor ?
 
Engineering news on Phys.org
If you add something to convert the high voltage back to low voltage to charge a battery, sure. Adding that costs much more than one capacitor charge for the rare cases where the defibrillator is used.
 
  • Like
Likes   Reactions: ernd59
And the defibrillator is only charged up if the EKG shows a shockable rhythm. I've never seen an aborted defibrillation -- have you?
 
  • Like
Likes   Reactions: ernd59
berkeman said:
And the defibrillator is only charged up if the EKG shows a shockable rhythm. I've never seen an aborted defibrillation -- have you?
it takes some time to charge the high voltage capacitor, and in a defibrillation every second counts. Normally while the EKG is working the capacitor is charging, I've never seen an aborted defibrillation too but theoretically it is possible
 
I guess the charge is discarded if the patient gets a pulse before getting shocked.
 
  • Like
Likes   Reactions: ernd59
ernd59 said:
it takes some time to charge the high voltage capacitor, and in a defibrillation every second counts. Normally while the EKG is working the capacitor is charging, I've never seen an aborted defibrillation too but theoretically it is possible

Not where I work. You always read the EKG to see if it is shockable rhythm, then it if is, you push the charging button. Where do you work where the defibrillator is charged before reading the EKG?
 
  • Like
Likes   Reactions: ernd59
berkeman said:
Not where I work. You always read the EKG to see if it is shockable rhythm, then it if is, you push the charging button. Where do you work where the defibrillator is charged before reading the EKG?
i'm a student working on bachelor thesis, the Professor told me or what i think he told me, that in some AED "Automated external defibrillator" the charging of the capacitor and the EKG reading are done in the same time to minimise waiting period and assure the effectiveness of the defibrillation
 
mfb said:
I guess the charge is discarded if the patient gets a pulse before getting shocked.
yes, that's what i meant :)
 
ernd59 said:
i'm a student working on bachelor thesis, the Professor told me or what i think he told me, that in some AED "Automated external defibrillator" the charging of the capacitor and the EKG reading are done in the same time to minimise waiting period and assure the effectiveness of the defibrillation

I am not aware of any AEDs that do that. Can you find out a specific brand and model?

I suppose you could propose that it could be a new feature, if you could return the energy to the battery if no defibrillation was needed. On the other hand, it is pretty dangerous to be charging the defibrillation capacitor if you aren't going to use it. All it takes is one slip near the controls to discharge the capacitor into the patient and whoever else is in contact with them. I'm not sure you could get safety approvals for such a "feature". My AED charges in just 2-3 seconds, and that amount of time is not going to make a big difference in mortality or morbidity...

ZOLL AEDPlus:
http://www.dremed.com/catalog/images/zoll_aed_plus2_lg.jpg
zoll_aed_plus2_lg.jpg
 
  • Like
Likes   Reactions: ernd59
  • #10
How energy recovery might be done will depend on the capacitor voltage and how the charger works. If it is an oscillator–transformer–rectifier charger then only the rectifier diode needs to be HV rated. It would be difficult to find a semiconductor to switch the energy back through the transformer to the rechargeable batteries.

If the HV pulse was generated by a Marx Generator, the energy recovery could be done more easily since the parallel capacitor connection has a much lower voltage.

But either way it is probably not worth the extra complexity. The equipment will still need it's battery recharged when idle.
 
  • Like
Likes   Reactions: ernd59
  • #11
berkeman said:
I am not aware of any AEDs that do that. Can you find out a specific brand and model?

I suppose you could propose that it could be a new feature, if you could return the energy to the battery if no defibrillation was needed. On the other hand, it is pretty dangerous to be charging the defibrillation capacitor if you aren't going to use it. All it takes is one slip near the controls to discharge the capacitor into the patient and whoever else is in contact with them. I'm not sure you could get safety approvals for such a "feature". My AED charges in just 2-3 seconds, and that amount of time is not going to make a big difference in mortality or morbidity...

So how does this work.
You check the patient - possible cardiac arrest or similar and irregular rhythm determined.
Defibrillation is chosen
Patient zapped
EKG tested - turns out OK
Defribulator turned to zero position
Paddles returned to their holding location on machine

Would it be a safety issue to have another KNOB or POSITION of the knob for Discharge to Battery.
I would think most operators in the spirit of the moment would find this distracting and perhaps confusing.
One would have to wait to return the paddles to their holding location for Discharge to Battery setting so that one can be sure arcing would not occur.
At least that is the way I see the safety issue.

Your thoughts.
 
  • Like
Likes   Reactions: ernd59
  • #12
berkeman said:
I am not aware of any AEDs that do that. Can you find out a specific brand and model?

I suppose you could propose that it could be a new feature, if you could return the energy to the battery if no defibrillation was needed. On the other hand, it is pretty dangerous to be charging the defibrillation capacitor if you aren't going to use it. All it takes is one slip near the controls to discharge the capacitor into the patient and whoever else is in contact with them. I'm not sure you could get safety approvals for such a "feature". My AED charges in just 2-3 seconds, and that amount of time is not going to make a big difference in mortality or morbidity...

ZOLL AEDPlus:
http://www.dremed.com/catalog/images/zoll_aed_plus2_lg.jpg
zoll_aed_plus2_lg.jpg
look at this Patent of a Portable defibrillators, they had a different goal but the idea is the same https://www.google.com.ar/patents/US7236823
 
  • #13
ernd59 said:
look at this Patent of a Portable defibrillators, they had a different goal but the idea is the same https://www.google.com.ar/patents/US7236823

Yeah, if you're going to be doing self-tests that charge up the output circuit, it makes sense to try to recover the energy. Keep in mind that AEDs generally use non-rechargeable batteries. Professional monitors with defibrillation capability use rechargeable batteries.
 
  • Like
Likes   Reactions: ernd59
  • #14
256bits said:
So how does this work.
You check the patient - possible cardiac arrest or similar and irregular rhythm determined.
Defibrillation is chosen
Patient zapped
EKG tested - turns out OK
Defribulator turned to zero position
Paddles returned to their holding location on machine

Each shock is separately charged. When you see the shockable rhythm, you push the charge button and wait a couple seconds for the light to come on. Then you call "everybody clear" (very important step! :smile: ), and press the discharge button. I don't know what happens if you don't press the discharge button. But I wouldn't want to be working on a patient while the HV was still charged up...
 
  • Like
Likes   Reactions: ernd59
  • #15
berkeman said:
Each shock is separately charged. When you see the shockable rhythm, you push the charge button and wait a couple seconds for the light to come on. Then you call "everybody clear" (very important step! :smile: ), and press the discharge button. I don't know what happens if you don't press the discharge button. But I wouldn't want to be working on a patient while the HV was still charged up...
OK thanks. Seen them in action but only in the movies, where they always do say "Clear", so I was wondering about the actual protocol.
 
  • #16
berkeman said:
Each shock is separately charged. When you see the shockable rhythm, you push the charge button and wait a couple seconds for the light to come on. Then you call "everybody clear" (very important step! :smile: ), and press the discharge button. I don't know what happens if you don't press the discharge button. But I wouldn't want to be working on a patient while the HV was still charged up...
can we consider not pressing the discharge button as self-tests?
 
  • #17
ernd59 said:
can we consider not pressing the discharge button as self-tests?

Not on most AEDs that I'm familiar with.

However, on manual Monitor/Defibrillators like the ZOLL X-Series, there are some self-tests that involve charging the HV power supply (both for Pacing and Defibrillating). AEDs can be used effectively by trained laypeople, but Monitors are for trained and licensed EMS and Hospital folks. See the section of the manual on Self-Testing:

http://www.zoll.com/medical-products/product-manuals/

:smile:
 
  • Like
Likes   Reactions: ernd59
  • #18
i found this in a master thesis dating back to 2003, it is called Automatic discharge :
The automatic discharge is needed to avoid the main capacitor being left at
high voltages when not operated. For instance, if the capacitor is charged and
the medical personnel finds it unnecessary to complete the defibrillation, the
capacitor should be discharged without connecting the load (the patient).
This
protection is most easily achieved by letting the Microcontroller monitor the time
passed since the charging stopped.
 

Similar threads

  • · Replies 7 ·
Replies
7
Views
2K
Replies
1
Views
1K
  • · Replies 15 ·
Replies
15
Views
2K
  • · Replies 32 ·
2
Replies
32
Views
3K
  • · Replies 49 ·
2
Replies
49
Views
5K
  • · Replies 78 ·
3
Replies
78
Views
6K
  • · Replies 19 ·
Replies
19
Views
2K
  • · Replies 21 ·
Replies
21
Views
2K
  • · Replies 28 ·
Replies
28
Views
15K
  • · Replies 7 ·
Replies
7
Views
2K