How to estimate the measurement uncertainty of an oscilloscope ?

AI Thread Summary
Estimating measurement uncertainty for a Tektronix oscilloscope, such as the DPO7054 or DPO7254, can be challenging due to insufficient information in the manual and calibration sheet. The calibration sheet provides uncertainty values in divisions but lacks percentage representation, complicating the calculation. To determine measurement uncertainty for pulse area, users should consider combining error specifications for vertical amplitude and horizontal time base jitter, along with frequency and voltage amplitude ranges. The manual indicates vertical accuracy ranges from 5.2 to 6 bits, translating to approximately 2.7% of the full-scale reading at 2.5 GHz. Understanding these parameters is crucial for accurate measurement uncertainty assessment.
nordmoon
Messages
65
Reaction score
0
TL;DR Summary
How to estimate the measurement uncertainty of an Textronics 500 MHz oscilloscope ?
I have an Textronics oscilloscope with 500 MHz sampling rate and using an 50 ohm cable for the measurement.

I can’t find the measurement uncertainty in the manual or the calibration sheet.

In the calibration sheet it says an ‘uncertainty’ for the used channel in divisions, for example 1m up to 10m (the scale) and with highest value being 1. It states the nominal value, upper and lower value and the measurement value. It also state an uncertainty (value between 1-2m) but not in %. I don’t understand how to determine the measurement uncertainty from this data. Anyone understand this?

The manual is even more cryptic .. no words on measurement uncertainty.
 
Engineering news on Phys.org
Do you have a model number?
 
Averagesupernova said:
Do you have a model number?

If I remember correctly it’s a DPO7054 or DPO7254.
 
nordmoon said:
Yes but I don’t understand how to determine the measurement uncertainty, for the measured pulse area, with this oscilloscope ?
Pulse area? Are you checking Eye Diagrams for communication circuits? What is the application?

I would think that you would combine the error specs for the vertical amplitude and horizontal time base jitter, but maybe there are other considerations. What frequency and rise/fall times are involved? What range of voltage amplitudes?
 
In the manual, for vertical accuracy see Appendix A, page A-8 "Effective bits, typical." Depending on frequency, the accuracy ranges between 5.2 and 6 bits.

Therefore, at 2.5GHz the vertical accuracy would be:
1/(25.2) = 1/36.76 = 2.7% of Full Scale reading

User Manual .pdf is available at:
https://download.tek.com/manual/071087903.pdf

Cheers,
Tom
 