Ignition coil HV pulse on oscilloscope?

AI Thread Summary
The discussion revolves around safely measuring high voltage (HV) pulses from a step-up transformer using an oscilloscope. Users explore the feasibility of using a lower-rated probe with a voltage divider made of high-voltage resistors to reduce the voltage input to the oscilloscope. Calculations indicate that a significant resistance value is needed to safely measure 10kV, which poses challenges in sourcing appropriate resistors. Participants emphasize the importance of isolating all components and ensuring proper safety precautions to avoid electrical hazards. The conversation concludes with suggestions for high-voltage probes and DIY solutions while stressing the necessity of careful design to prevent damage to equipment.
BHY-BK
Messages
18
Reaction score
1
Hi all,
I've been using a HV step up transformer (10kV-20W on secondary) lately and I'm wanting to look at the HV pulses on an oscilloscope. The coil is isolated from ground. I know connecting a probe directly to it is dangerous and would ruin my scope. I want to use a differential probe but they are crazy expensive so this got me wondering...

Instead of using a differential probe rated at 10+kV could I use a lower rated probe (say 1kV) then a voltage divider on the circuit using some HV resistors across the components I want to take measurements across? Would this be safe to do as long as its done right. I should mention also the pulsing frequency is 25kHz so I shouldn't need to worry about parasitic elements on the divider, or do I?
 
Engineering news on Phys.org
BHY-BK said:
Instead of using a differential probe rated at 10+kV could I use a lower rated probe (say 1kV) then a voltage divider on the circuit using some HV resistors across the components I want to take measurements across?
Yes, but you need to isolate every bit of wire and then put the whole resistor chain inside an isolating sleeve. 10kV creates sparks at every sharp corner (or wire end).

Now for a bit of calculation: If you have a 10:1 probe connected to your scope, it will have an input impedance of 10MΩ (that is fairly standard). Assuming that your scope can handle a 10V input, the 10:1 probe expands that to 100V. In order to look at 10kV, you need a series resistance R given by \frac{100V}{10kV}=\frac{10M\Omega}{R+10M\Omega}, giving a value for R = 990MΩ. That kind of resistance is hard to come by! For an example of a commercial high voltage probe, see http://www.elexp.biz/tst_vp40.htm.
 
  • Like
Likes BHY-BK
Svein said:
Yes, but you need to isolate every bit of wire and then put the whole resistor chain inside an isolating sleeve. 10kV creates sparks at every sharp corner (or wire end).

Now for a bit of calculation: If you have a 10:1 probe connected to your scope, it will have an input impedance of 10MΩ (that is fairly standard). Assuming that your scope can handle a 10V input, the 10:1 probe expands that to 100V. In order to look at 10kV, you need a series resistance R given by \frac{100V}{10kV}=\frac{10M\Omega}{R+10M\Omega}, giving a value for R = 990MΩ. That kind of resistance is hard to come by! For an example of a commercial high voltage probe, see http://www.elexp.biz/tst_vp40.htm.
Thanks for the help Svein,

So I'm looking at buying this diff probe
https://www.globalmediapro.com/dp/A01LT0/Pintek-DP-25-Differential-Probe-25MHz-1000V/#Compatibles

The specs say the probe has:
Input Impedance :
2M Omh, 2.3 PF between inputs and ground
4M Omh, 1.2 PF between inputs

and

Attenuation : x 20, x 50, x 200 (Into 1M Ohm scope) ; x 40, x 10, x 400 (Into 50 Ohm load)

So how do I calculate the resistance values I would need? Just to be on the safe side I would want to use the divider to reduce the voltage across the diff probe to 500V from 10kV. I know how to calculate the voltage drop across the resistances to get 500V but how do I know what resistance values to use?

I just don't understand the equation you showed...How did you get 990M ohms?Ok, I think I figured it out. I never thought to impedance match the scope. Glad I asked on this forum.
 
Last edited:
BHY-BK said:
Attenuation : x 20, x 50, x 200 (Into 1M Ohm scope) ; x 40, x 10, x 400 (Into 50 Ohm load)
So, if your scope has an input of 1MΩ and a max input level of 10V, using a 200x attenuation means a max input level of 200⋅10V = 2kV. If your scope can handle 50V input, you are just barely within range. Otherwise, look for a probe with a higher attenuation.
 
It all depends on how much you want to load down the signal.

A series resistor on each leg, with a small resistor across them will get the job done. I wouldn't depend on the probe as a calibrated load.

For example two 1K resistors feeding a 1 ohm resistor gives 5V,(10KV/2001 ohms) but I'm sure that is too high a load (2000 ohms 10KV is 5 amps)
10K 10 ohms
100K 100 ohms
1M 1000 ohms will draw 5ma, which still may be too high.
etc
I found these 10M 10KV resistors for $113 each (lol) https://www.amazon.com/dp/B00Q7VQ9NI/?tag=pfamazon01-20

Here is a page of ohmite high voltage resistors
http://www.ohmite.com/search.php?appl=High Voltage&function=results

Here is a high voltage divider (I think you need two):
http://www.ohmite.com/contact.php?function=form&which=divider&set=slimmox
 
Last edited by a moderator:
meBigGuy said:
It all depends on how much you want to load down the signal.
Exactly. 20W at 10kV is only 2mA of current so your Pot Chain needs to take significantly less than that if you don't want your measurement to affect the circuit operation. If you want better than 10% error then your chain must take less than 200μA. That would need a resistance higher than 10M. I looked at "high voltge probes" and they come at a couple of hundred GBP and have a resistance in the region of 10M (confirming my idea). I think that (safe) idea could be the best way to go. You could always try eBay . . . . In eBay UK, there are some for sale at around 50GBP.
 
check this series of resistors

http://www.ohmite.com/cat/res_minimox.pdf

they're not very expensive

a string of 10 or 100 meg ersistors to make a divider sounds practical enough

pay attention to svein's advice about high voltage precautions
 
It might be a good idea, if you are going for a DIY solution, to put a Zener diode across your scope input to short out any unwanted over volts due to breakdown. At least, you won't do a lot of damage with only 20mA,
 