Infinity corrected lens - expanding before tube lens?

AI Thread Summary
The discussion revolves around challenges faced in a home-built setup combining Raman microscopy and optical tweezers. The user converted to a Nikon infinity-corrected 100x objective but is experiencing rapid divergence of white light, capturing only 25% of it with a tube lens positioned 200 mm away. The divergence is attributed to the need for the objective to image off-center objects, and there are suggestions to adjust the tube lens position or use a larger diameter lens. The user also seeks advice on ensuring the trapping beam is centered and perpendicular to the objective's back aperture. Overall, the conversation highlights the complexities of integrating optical components in microscopy setups.
donroy81
Messages
7
Reaction score
0
Hello folks, I am on a project that involves combining Raman microscopy + optical tweezers, and I made a setup over the past year. All of this is home built on minimal funds so its not the neatest of all setups. It involves an upright microscope (coupled to trapping laser) and collecting the back reflected Raman signal at 180o and also the optical image along the same axis.

To make things easy i converted to a NIKON infinity corrected 100x objective and added a tube lens. While the back reflected laser beam from the laser seems to be collimated when the sample is at focus, the white light (illuminated from the bottom) diverges very fast so I catch only 25% of l the light on my tube lens (which needs to be 200 mm away anyway). Should not the white light collected be the same size as the back aperture of the objective (8mm) : I need to catch all the light on 25mm dichroic filters (which eliminate the laser).

Why is the white light expanding out of the objective - Is something wrong? We are a spectroscopy group and not a microscopy one, so all this is pretty new to me and my research group. Any help is appreciated.
 
Last edited:
Science news on Phys.org
How are you using the white light? Is there a condenser (a lens that focuses the light onto the sample)?
 
Hi Andy, thanks for replying. I have a 10x (.25NA)objective as a condenser, I was using a 1.25 NA condenser before with the same effect.
I know its not Kohler illumination (since the objective is 1.25 oil) but this is also what Thorlabs uses in their "optical trapping kit".
 
I'm not surprised that the white light is diverging- it has to, in order for the objective to image objects that are off-center. Some of the diverging light may be 'spurious', in that it would normally be rejected by the tube lens- objectives have a limited field of view. Have you (without the trapping beam) simply looked at the image your setup produces?

I'm not sure how much control you have over the divergence angle (does it change between using the 0.25 NA and 1.25 NA condensers? Can you ring up a field stop?), so one solution may simply be to move the tube lens closer (or use a larger diameter tube lens). AFAIK, the tube lens on my Leica is about 100mm off the back aperture of the objective.
 
Andy, sorry for the delay, I wanted to review everything once. I also went over your conversation with Kubikat. I would need to show you some pictures to explain. Slide1 is a closeup of the problem region.I have 1" optics, the filter and the mirrors before the tube lens. DSC09636 gives you another take on it. Everything was fine when I was using a 160mm objective and a complete Nachet, we decided that the infinity will help with the raman and hence this. 2" dichroics run too many $$. In anycase, I don't see why a parallel flux of light from an infinity objective with a back aperture would expand so fast to 2".

Additionally, if I may, I would like to know a few more things, for starters : how do you check that the trapping beam is on center of the back aperture of the objective and perfectly perpendicular to it?
 

Attachments

  • Slide1.jpg
    Slide1.jpg
    40.1 KB · Views: 576
  • DSC09636.JPG
    DSC09636.JPG
    35 KB · Views: 607
Thread 'A quartet of epi-illumination methods'
Well, it took almost 20 years (!!!), but I finally obtained a set of epi-phase microscope objectives (Zeiss). The principles of epi-phase contrast is nearly identical to transillumination phase contrast, but the phase ring is a 1/8 wave retarder rather than a 1/4 wave retarder (because with epi-illumination, the light passes through the ring twice). This method was popular only for a very short period of time before epi-DIC (differential interference contrast) became widely available. So...
I am currently undertaking a research internship where I am modelling the heating of silicon wafers with a 515 nm femtosecond laser. In order to increase the absorption of the laser into the oxide layer on top of the wafer it was suggested we use gold nanoparticles. I was tasked with modelling the optical properties of a 5nm gold nanoparticle, in particular the absorption cross section, using COMSOL Multiphysics. My model seems to be getting correct values for the absorption coefficient and...