Medical Grade UPS 3000VA - Meet UL 60601-1 Requirements

  • Thread starter Thread starter Jetach
  • Start date Start date
  • Tags Tags
    Medical
AI Thread Summary
The discussion centers on the search for a medical-grade UPS that meets UL 60601-1 requirements, specifically one that can output 2kW for about 30 minutes. The user has identified a Tripp Lite model that provides 1920 Watts for 40 minutes but is seeking more options. There is a suggestion to use Google Images for better search results and to explore other manufacturers that may offer compliant units. Additionally, it is noted that UL 544 has been replaced by UL 60601-1, which imposes stricter testing standards. The emphasis is on finding a reliable UPS that adheres to these regulations for medical equipment use.
Jetach
Messages
9
Reaction score
0
Hello PF,

After doing much research, I've decided that buying a retail UPS would be a much better idea than building one from scratch even though the cost is higher.

I need to output 2kW for roughly 30 minutes.

I am looking for a 3000 VA UPS that meets medical grade requirements.

I believe that regulation is UL 60601-1, however there may be more things I may be missing.

Any help would be appreciated.
 
Engineering news on Phys.org
Jetach said:
Hello PF,

After doing much research, I've decided that buying a retail UPS would be a much better idea than building one from scratch even though the cost is higher.

I need to output 2kW for roughly 30 minutes.

I am looking for a 3000 VA UPS that meets medical grade requirements.

I believe that regulation is UL 60601-1, however there may be more things I may be missing.

Any help would be appreciated.

Sounds like a good decision. :smile:

What have you found so far via Google? I often will use Google Images for searches like this, to get better results.
 
berkeman said:
Sounds like a good decision. :smile:

What have you found so far via Google? I often will use Google Images for searches like this, to get better results.

So far I've only found one UPS that fits my needs which would be a Tripp Lite

http://www.tripplite.com/en/products/model.cfm?txtSeriesID=748&txtModelID=5440

with a battery it can run 1920 Watts for 40 minutes.

It appears this is the only option available, I called up APC and their customer service didn't appear to know what the UL 60601-1 regulation was.
 
Jetach said:
So far I've only found one UPS that fits my needs which would be a Tripp Lite

http://www.tripplite.com/en/products/model.cfm?txtSeriesID=748&txtModelID=5440

with a battery it can run 1920 Watts for 40 minutes.

It appears this is the only option available, I called up APC and their customer service didn't appear to know what the UL 60601-1 regulation was.

I did a Google Images search on +"60601-1" ups, and got some reasonable looking units from different manufacturers. Maybe try that search to see if the other manufacturers have medical grade units in your power range...
 
In your earlier thread, on 24 June, berkeman wrote, “The pertinent UL standard for medical equipment is UL544”
Just do a Google search using the term “UL 544 ups” and many suitable commercial units and their standards appear.
 
Well, our device in development will be used on patients, and anything coming within 6 feet radius of a patient needs to have UL 60601-1. I believe that UL544 was withdrawn sometime in 2009 and replaced with UL 60601-1. The testing for this is much more stiffer, which is ultimately what gave me the decision to find one in a marketplace.
 
Very basic question. Consider a 3-terminal device with terminals say A,B,C. Kirchhoff Current Law (KCL) and Kirchhoff Voltage Law (KVL) establish two relationships between the 3 currents entering the terminals and the 3 terminal's voltage pairs respectively. So we have 2 equations in 6 unknowns. To proceed further we need two more (independent) equations in order to solve the circuit the 3-terminal device is connected to (basically one treats such a device as an unbalanced two-port...
suppose you have two capacitors with a 0.1 Farad value and 12 VDC rating. label these as A and B. label the terminals of each as 1 and 2. you also have a voltmeter with a 40 volt linear range for DC. you also have a 9 volt DC power supply fed by mains. you charge each capacitor to 9 volts with terminal 1 being - (negative) and terminal 2 being + (positive). you connect the voltmeter to terminal A2 and to terminal B1. does it read any voltage? can - of one capacitor discharge + of the...
Thread 'Weird near-field phenomenon I get in my EM simulation'
I recently made a basic simulation of wire antennas and I am not sure if the near field in my simulation is modeled correctly. One of the things that worry me is the fact that sometimes I see in my simulation "movements" in the near field that seems to be faster than the speed of wave propagation I defined (the speed of light in the simulation). Specifically I see "nodes" of low amplitude in the E field that are quickly "emitted" from the antenna and then slow down as they approach the far...

Similar threads

Replies
18
Views
4K
Replies
7
Views
2K
Replies
2
Views
2K
Replies
5
Views
4K
Replies
56
Views
53K
Replies
6
Views
5K
Replies
14
Views
16K
Back
Top