- #1
Guineafowl
- 791
- 383
- TL;DR Summary
- As title.
This is for safely powering up devices being tested/repaired. Supply is 240V rms single phase, protected by 32A MCB and 30 mA RCD/GFCI, then fused at 13A.
There doesn’t seem to be a consensus online, and my EE and repair books just say I need both, without specifying order.
I think the isolation tx should go first, since it is designed to input/output 240V, and that is exactly what it’ll be doing. That way, the variac will take in 240V and output a variable voltage, again as per design.
If you put the variac first, the isolation transformer will be given a variable input voltage, from 0-270V.
Perhaps someone could confirm if I’m right, wrong, or it doesn’t matter. Mains Earth will of course be taken straight through without interruption.
There doesn’t seem to be a consensus online, and my EE and repair books just say I need both, without specifying order.
I think the isolation tx should go first, since it is designed to input/output 240V, and that is exactly what it’ll be doing. That way, the variac will take in 240V and output a variable voltage, again as per design.
If you put the variac first, the isolation transformer will be given a variable input voltage, from 0-270V.
Perhaps someone could confirm if I’m right, wrong, or it doesn’t matter. Mains Earth will of course be taken straight through without interruption.