Power supply for testing: Variac or isolation transformer first?

Click For Summary

Discussion Overview

The discussion revolves around the optimal configuration for using a variac and an isolation transformer in a test setup for safely powering devices. Participants explore the implications of the order in which these components are connected, considering factors such as voltage tolerances and operational safety.

Discussion Character

  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • One participant suggests that the isolation transformer should be first in the setup to ensure it receives a stable 240V input, allowing the variac to output a variable voltage as designed.
  • Another participant agrees with the idea of placing the isolation transformer first, citing the mains voltage tolerance and the risk of exceeding 270V into the isolation transformer if the variac is placed first.
  • A different participant recalls that in their experience with large telecom switch-mode power supplies, the isolation transformer was always placed first to facilitate testing under over-voltage conditions.
  • One participant mentions that either configuration could work but leans towards placing the isolation transformer first to prevent magnetizing current from affecting the variac.
  • Another participant notes that the isolation transformer in use is a rewound battery charger transformer, likely around 300 VA, and discusses the overall design of their test supply panel.
  • Concerns are raised about the size and rating of the isolation transformer, with one participant questioning the appropriateness of a 13A rating for the transformer in this context.

Areas of Agreement / Disagreement

Participants generally agree that the isolation transformer should be placed first, but there are differing opinions on the implications of each configuration and whether it truly matters. The discussion remains unresolved regarding the best practice.

Contextual Notes

Participants mention various assumptions about voltage tolerances and the specific ratings of their equipment, which may affect their recommendations. There is also a lack of consensus on the impact of magnetizing currents from either device.

Who May Find This Useful

This discussion may be useful for electronics hobbyists, engineers, or technicians involved in testing and repairing electrical devices, particularly those working with high voltage or sensitive equipment.

Guineafowl
Messages
910
Reaction score
418
TL;DR
As title.
This is for safely powering up devices being tested/repaired. Supply is 240V rms single phase, protected by 32A MCB and 30 mA RCD/GFCI, then fused at 13A.

There doesn’t seem to be a consensus online, and my EE and repair books just say I need both, without specifying order.

I think the isolation tx should go first, since it is designed to input/output 240V, and that is exactly what it’ll be doing. That way, the variac will take in 240V and output a variable voltage, again as per design.

If you put the variac first, the isolation transformer will be given a variable input voltage, from 0-270V.

Perhaps someone could confirm if I’m right, wrong, or it doesn’t matter. Mains Earth will of course be taken straight through without interruption.
 
Engineering news on Phys.org
I would put the Iso transformer first. The 240Vac mains has a tolerance on it (10%?), so if you put the Variac first, you could get more than 270V into the Iso. (It still would probably be fine.)

BTW, 13A at 240V is a mongo Iso transformer, no? How big is that thing?
 
In a past life working on large telco SMPS, the iso transformer was first, then variac, I think from memory it was that way because the variac could provide somewhere like +20% or 25% voltage so we could test over voltage operation etc.
 
Guineafowl said:
Supply is 240V rms single phase, protected by 32A MCB and 30 mA RCD/GFCI, then fused at 13A.
berkeman said:
BTW, 13A at 240V is a mongo Iso transformer, no? How big is that thing?
Does your Iso transformer look something like this?

https://www.ato.com/3-kva-isolation-transformer
245039
 
You can do it either way. I would probably put the transformer first. That way the magnetizing current from the transformer (that's the primary current that flows in addition to the currents caused by the load) won't have to go through the variac. Although I suppose you could make the same argument about the viriac's magnetizing current too.
As others have said the variac being downstream will keep the transformer from seeing the extra 10-20% over-voltage when the variac is turned all the way up (if it does that). However, I' sure the transformer would be ok, it will just get a little hotter.
If either is rated for much larger power than the other, I would probably put that one closer to the source.
 
Thanks all. Sounds like iso first.

berkeman said:
I would put the Iso transformer first. The 240Vac mains has a tolerance on it (10%?), so if you put the Variac first, you could get more than 270V into the Iso. (It still would probably be fine.)

BTW, 13A at 240V is a mongo Iso transformer, no? How big is that thing?

Ah, no - the supply is UK standard, but the isolation transformer is a rewound battery charger transformer, probably 300 VA or so. It’s just for powering up devices for testing, so I can scope them and flip the boards around without too much danger.

This and the variac (which is 2.4 kVA) will be used as part of my test supply panel:
0D5441F4-5E88-4417-8FF2-3F719DD9372B.png


Master switch turns everything off; second one toggles mains/isolated, and isolation tx can be turned off; next section allows a series bulb or complete bypass (‘missile’ cover so I don’t accidentally...); next three switches add in parallel bulbs to allow more current through in limit mode.

Finally, a decent socket with DP switch and a kill-a-watt type meter so I can measure what’s going on.

This will be for careful powering on of valve(tube) radios, and general solid-state boards with faults. This panel has been tested to UK standard:
Continuity/polarity.
Earth bond (send 30A through each Earth path (including switch bodies) to check integrity).
Insulation resistance (500Vdc between L+N and E).

Hopefully a useful addition to the bench!
 
Looks like a useful setup! :smile:
Guineafowl said:
probably 300 VA or so
In that case, I'd probably downsize the fuse for your setup...
Guineafowl said:
then fused at 13A
 
  • Informative
Likes   Reactions: Guineafowl

Similar threads

  • · Replies 8 ·
Replies
8
Views
3K
  • · Replies 1 ·
Replies
1
Views
2K
  • · Replies 36 ·
2
Replies
36
Views
5K
  • · Replies 11 ·
Replies
11
Views
3K
Replies
61
Views
10K
Replies
19
Views
3K
Replies
3
Views
3K
  • · Replies 2 ·
Replies
2
Views
2K
  • · Replies 47 ·
2
Replies
47
Views
5K
  • · Replies 52 ·
2
Replies
52
Views
4K