The discussion revolves around a problem encountered in Raman spectroscopy analysis of Propranolol HCl, where a specific batch fails to produce a detectable spectrum, unlike other batches that yield normal results. The user has successfully used other techniques like infrared and NMR for identification but seeks to understand the Raman spectroscopy issue. Suggestions include testing the Raman spectrometer with known standards, such as polystyrene, and considering the physical characteristics of the batch, including granulometry. The laser wavelength used is 785 nm, which cannot be changed. The possibility of low Raman intensity due to the material's characteristics is raised, along with recommendations to adjust experimental parameters like exposure time and optical power to improve signal detection.
GuiHn
Messages
3
Reaction score
0
Hello,

Recently I had a problem in the analysis of a raw material called Propranolol HCl, when performing Raman identification for the specific batch the obtained spectrum does not generate signal and, consequently, identity of the material.

By testing other techniques like infrared and NMR, I got the results without problems.

It turns out that I need to understand why the problem occurs only in Raman spectroscopy. Searching the internet did not find anything that could explain, since the problem happens only with this lot in question, other lots of the same product provide perfect spectra.

Anyone who has good knowledge of the technique could help me?

Thank you.
 
Chemistry news on Phys.org
I'm far from an expert in the field, but my first intuition is to test the Raman spectrometer. Have you tried producing other spectra, such as for ethanol or acetone, which you probably have in your lab?
 
DrClaude said:
I'm far from an expert in the field, but my first intuition is to test the Raman spectrometer. Have you tried producing other spectra, such as for ethanol or acetone, which you probably have in your lab?
Hello, thanks for the help. In the laboratory we have a polystyrene standard that is made before any reading to evaluate the equipment. In addition, in the same equipment we made other batchs of the same product that had the spectra with normal intensity, only for this batch that we did not obtain spectra with product identity.
Initially I imagined that it could be a physical characteristic of the batch, but the granulometry and other aspects are similar with the other batchs that approved in the identification
 
That's strange. What wavelength is the laser?
 
HAYAO said:
That's strange. What wavelength is the laser?
Hello, the wavelenght is the standart of equipament 785nm, I can't change this value.
I've been thinking that this raw material has a smaller fluorescence signal, but I don't have the equipment to prove it.
 
Some material have low raman intensivity. Do you have big thickness this material ? Do you have long exposition? Can you growing optical power up or wide slit?
 
I want to test a humidity sensor with one or more saturated salt solutions. The table salt that I have on hand contains one of two anticaking agents, calcium silicate or sodium aluminosilicate. Will the presence of either of these additives (or iodine for that matter) significantly affect the equilibrium humidity? I searched and all the how-to-do-it guides did not address this question. One research paper I found reported that at 1.5% w/w calcium silicate increased the deliquescent point by...
I was introduced to the Octet Rule recently and make me wonder, why does 8 valence electrons or a full p orbital always make an element inert? What is so special with a full p orbital? Like take Calcium for an example, its outer orbital is filled but its only the s orbital thats filled so its still reactive not so much as the Alkaline metals but still pretty reactive. Can someone explain it to me? Thanks!!
I'm trying to find a cheap DIY method to etch holes of various shapes through 0.3mm Aluminium sheet using 5-10% Sodium Hydroxide. The idea is to apply a resist to the Aluminium then selectively ablate it off using a diode laser cutter and then dissolve away the Aluminium using Sodium Hydroxide. By cheap I mean resists costing say £20 in small quantities. The Internet has suggested various resists to try including... Enamel paint (only survived seconds in the NaOH!) Acrylic paint (only...