Self-Mixing Interferometer vs Capacitive Sensor

AI Thread Summary
For building a non-contact displacement sensor sensitive to micrometer to millimeter motion, the self-mixing laser interferometer is generally considered a better choice than capacitive sensors, despite the latter being less expensive. Capacitive sensors can face issues like offset shifts over time, making them unreliable for absolute position measurements, while laser interferometers provide relative displacement information but require a limit stop for absolute positioning. Capacitive sensors need to be positioned very close to the target, and their performance can be affected by humidity, although this typically results in offset rather than sensitivity errors. The choice between these technologies depends on specific requirements for resolution, accuracy, and environmental conditions. Ultimately, understanding the strengths and weaknesses of each sensor type is crucial for making an informed decision.
Isaiah Gray
Messages
18
Reaction score
0
I'm trying to build a non-contact displacement sensor, sensitive to motion in the micrometer to millimeter range, for a research project. My two best choices are a self-mixing laser interferometer and a capacitive displacement sensor. Has anyone had enough experience with these types of sensors to know if they are doable?
 
Engineering news on Phys.org
Isaiah Gray said:
I'm trying to build a non-contact displacement sensor, sensitive to motion in the micrometer to millimeter range, for a research project. My two best choices are a self-mixing laser interferometer and a capacitive displacement sensor. Has anyone had enough experience with these types of sensors to know if they are doable?

I'm not able to say that I have a lot of experience, but the laser interferometer would seem to the the better choice, IMO.

Capacitive sensing involves lots of problematic issues, in my limited experience. It may be that your measurement distances are short enough that capacitive sensing may be practical, though. Can you provide more complete specs on the physical space available for the sensor (capacitor area) and accuracy requirements? What are related EMI issues (are there any things nearby the space that would be interfered with by the capacitive sensing voltage waveform)? Can you control the humidity of the gas in the gap of the variable capacitive sensor?
 
BTW, you probably already know that the laser interferometer measurement gives you relative displacement information. It does not give you any absolute displacement information. You need a "limit stop" type of input to give you a starting position, before you can use the interferometer's information to give you the displacement from that limit stop.
 
You must determine your desired resolution/accuracy to make a good selection. All technologies have their weaknesses and you can only choose when you know when the those weaknesses are beyond your tolerance.

Regarding Capacitive Sensors:
I would say capacitive sensors are also relative measurement devices (again, depending on tolerance). Capacitive sensors usually exhibit offset shifts over long periods of time (months/years) making them poor absolute position devices at the sub-micron/nm level.

The humidity issue for capacitive sensors mentioned by berkeman is real but not usually a problem. Humidity-related errors are offsets, not sensitivity errors, so they remain accurate over a wide range of humidity provided humidity doesn't change significantly during measurement. These errors are usually in the submicron range.

Larger ranges require larger sensing areas. Millimeters is a large range for cap sensors. Up to 2mm can be achieved with a 3/8" diameter probe. If you want 5mm you're now using an 18mm diameter probe. The target diameter should be 130% of the probe diameter, or special calibration will be required.

Capacitive probes must be very near the target compared to interferometers.

Cap sensors cost much less than interferometers.

I've never heard of a cap probe causing EMI because of the sensing field. Our probes have 1MHz, 100Vp-p drive voltage present on the sensing surface and extremely low currents. They are used in electron microscopy positioning stages in the presence of high DC voltages (2-3kV) without any problem.
 
Last edited by a moderator:
It's always best to get what you can off the shelf.
I haven't worked with interferometry systems, but according to people I worked with, they purchased the required materials from HP.
I used to design capacitance based sensors for a living. They could give you amazing resolution, without the fuss of lasers (i.e. 10nm), but we never attempted to use them more than about 1/20 of a mm away. Also, you either needed to be able to design exceptional capacitance measurement or buy it from another instrument house (Andeen-Hagerlan).
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
I have an engine that uses a dry sump oiling system. The oil collection pan has three AN fittings to use for scavenging. Two of the fittings are approximately on the same level, the third is about 1/2 to 3/4 inch higher than the other two. The system ran for years with no problem using a three stage pump (one pressure and two scavenge stages). The two scavenge stages were connected at times to any two of the three AN fittings on the tank. Recently I tried an upgrade to a four stage pump...