Simple effective resistance of a circuit at low voltages

  • Thread starter Thread starter zapnthund50
  • Start date Start date
  • Tags Tags
    Circuit Resistance
AI Thread Summary
The discussion revolves around measuring effective resistance in a circuit at low voltages (0-5V) and the discrepancies observed compared to known resistance values. The varying effective resistances are attributed to factors like temperature changes in the wire and inaccuracies in measurement equipment, particularly at low current levels. Participants emphasize that resistance can change due to temperature coefficients and suggest that the measuring setup may introduce errors, especially when dealing with microamp currents. There is a consensus that using lower resistor values and higher currents could yield more accurate results. The conversation highlights the importance of proper measurement techniques and equipment specifications in obtaining reliable data.
zapnthund50
Messages
31
Reaction score
6
I passed small voltages (0-5V) through a known resistance, measured the current each time, and got different "effective resistances" for each case. However, when graphed, the effective resistances appear to approach the known resistance of the circuit as a limit at higher voltages.

What is the reason for this? Nobody I ask can give me a satisfactory response. I've been told resistance never changes ... is it due to the inner workings of the ammeter I'm measuring it with?

Thanks!
 
  • Like
Likes Asymptotic
Engineering news on Phys.org
It's not entirely true that resistance can't vary. Resistance of a wire depends on the temperature of the wire as well as it's temperature coefficient of resistance (which indeed depends on the material of the wire).
Rt=R0(1 + αt)

where Rt is the resistance at t°C, R0 is the resistance at 0°C,
t is the temperature of the wire when the current of known potential difference is applied
α is the temperature coefficient of resistance depending on the material of the wire.

So, as the potential difference varies, the temperature of the wire changes slightly and deviates slightly from the experimental value
 
You've asked us to evaluate data that you haven't shown..

"I've got a secret" is a game, not a request for help.
 
  • Like
Likes NTL2009 and PhysicoRaj
jim hardy said:
You've asked us to evaluate data that you haven't shown..

"I've got a secret" is a game, not a request for help.

Sorry didn't mean to be secretive, lol. Here's a pic of the data and the graph I constructed from it (I just uploaded it here as a jog)
 

Attachments

  • input resistance.jpg
    input resistance.jpg
    33.5 KB · Views: 528
HPPAS said:
It's not entirely true that resistance can't vary. Resistance of a wire depends on the temperature of the wire as well as it's temperature coefficient of resistance (which indeed depends on the material of the wire).
Rt=R0(1 + αt)

where Rt is the resistance at t°C, R0 is the resistance at 0°C,
t is the temperature of the wire when the current of known potential difference is applied
α is the temperature coefficient of resistance depending on the material of the wire.

So, as the potential difference varies, the temperature of the wire changes slightly and deviates slightly from the experimental value

Hmm, so the resistance depends linearly on temperature...but this doesn't match my results, I'm still confused
 
HPPAS said:
It's not entirely true that resistance can't vary. Resistance of a wire depends on the temperature of the wire as well as it's temperature coefficient of resistance (which indeed depends on the material of the wire).
Rt=R0(1 + αt)

where Rt is the resistance at t°C, R0 is the resistance at 0°C,
t is the temperature of the wire when the current of known potential difference is applied
α is the temperature coefficient of resistance depending on the material of the wire.

So, as the potential difference varies, the temperature of the wire changes slightly and deviates slightly from the experimental value
Wait a second I might have stumbled on part of my answer. P=I2R, so if temperature itself is dependent on the current (as I think it should be, since the temperature increase in a resistor arises from the resistance to current flow), then this means that the temperature is proportional to power, but itself approaches an upper temperature limit. But this would only work for a nonlinear resistor, so I'm still in the dark lol. Thoughts?
 
Last edited:
  • Like
Likes baldbrain
Where are you measuring the voltage across the resistor? PD should always be measured independently across the terminals and 'inside' the connections to the power supply. You have to eliminate contact resistance and other such nasties.
What sort of temperatures are you experiencing? A tungsten filament can increase in resistance by a factor a ten, when it is at full power. What metal is the wire made of?
In fact, it would improve your chances of a good answer if you gave a full description of your experiment (circuit diagram and even a photo of the layout). You can't assume we know exactly what you have ben doing and the details can be very important.
 
  • Like
Likes Asymptotic and baldbrain
sophiecentaur said:
Where are you measuring the voltage across the resistor? PD should always be measured independently across the terminals and 'inside' the connections to the power supply. You have to eliminate contact resistance and other such nasties.
What sort of temperatures are you experiencing? A tungsten filament can increase in resistance by a factor a ten, when it is at full power. What metal is the wire made of?
In fact, it would improve your chances of a good answer if you gave a full description of your experiment (circuit diagram and even a photo of the layout). You can't assume we know exactly what you have ben doing and the details can be very important.

The circuit was built on this breadboard by Digilent Inc (http://store.digilentinc.com/electr...-one-usb-oscilloscope-multimeter-workstation/). It was simply a few resistors in series, the voltage was measured from one resistor leg to the other, across the whole resistor chain. Since I was only working on 0-5V with only 0.02-0.18 mA current, the temperature rise must be so tiny its insignificant. I know the explanation must be something fundamental that I'm just missing.

Again, I'll post the pic of my results, which has the resistance graph on it.
 

Attachments

  • input resistance.jpg
    input resistance.jpg
    33.5 KB · Views: 505
Here's a plot of your data with a polynomial degree 1 fit:

p1_fit.png


Doesn't look that bad. Fit gives an effective resistance of 30.85 kohm.

I'd take a guess that your measurements are a bit inaccurate down at those microamp levels. Have a look at the datasheet for your equipment for its accuracy at whatever range setting you're using.

Alternatively, lower the values of your resistors.
 
  • Like
Likes anorlunda
  • #10
milesyoung said:
Here's a plot of your data with a polynomial degree 1 fit:

View attachment 197561

Doesn't look that bad. Fit gives an effective resistance of 30.85 kohm.

I'd take a guess that your measurements are a bit inaccurate down at those microamp levels. Have a look at the datasheet for your equipment for its accuracy at whatever range setting you're using.

Alternatively, lower the values of your resistors.

Yes, that is the plot I got as well ... and I do understand that we get resistance from the slope. I guess what I'm failing to grasp is why the Input Resistance varies, while the real resistance remains constant. My professor could not really explain it to me, he studied mechanical, not electrical lol. Quite simply, here is my confusion. Take the first data point, 0.1 volts, 0.02 milliamps. Then according to V=IR, the resistance should be 5000 ohms. But obviously the resistance was 31 kohms, so ... what is the concept I'm failing to comprehend?
 
  • #11
zapnthund50 said:
Take the first data point, 0.1 volts, 0.02 milliamps.
The equipment you're using to measure this isn't perfect. Typically, the lower the value of something you're measuring, the more inaccurate your equipment gets. That's only 20 microamps!

Try redoing your experiment with lower values of resistors. Aim for an effective resistance of a couple hundred ohm or something like it. Make sure you don't go too low and fry your resistors, although this is a rite of passage that you'll inevitably go through at some point.
 
  • Like
Likes Asymptotic
  • #12
milesyoung said:
The equipment you're using to measure this isn't perfect. Typically, the lower the value of something you're measuring, the more inaccurate your equipment gets. That's only 20 microamps!

Try redoing your experiment with lower values of resistors. Aim for an effective resistance of a couple hundred ohm or something like it. Make sure you don't go too low and fry your resistors, although this is a rite of passage that you'll inevitably go through at some point.

Okay thanks, I will retry the experiment with lower resistors and higher current. The thing is though, I think what I'm describing is an attribute of Linear Resistors. Ohms law is just a description of linear resistors, right? V=IR, this can be looked at in the basic linear form y=mx+b, where b=0 and m=R. Then y=Rx. We both graphed voltage on the x-axis, and current on the y-axis, so the equation of our line is actually x = y*(1/R), or I=V*(1/R). So no matter what value resistor I measure, I should always get the linear VI relationship, right? Because that's why we call them linear resistors.

Is it a property of linear resistors to have a very small resistance at low voltages, but at high enough voltages, they approach their rated resistance value? That's the question that's bugging me. I've used the fitted equation that we both got from the data to extrapolate the current out to about 100 volts, and I've attached the resulting graph. As you can see, resistance starts at zero, but quickly climbs to the rated value. Thoughts on this?
 

Attachments

  • input resistance 2.jpg
    input resistance 2.jpg
    18 KB · Views: 500
  • #13
zapnthund50 said:
Is it a property of linear resistors to have a very small resistance at low voltages, but at high enough voltages, they approach their rated resistance value?
No, there are significant nonlinear effects that depend on temperature and (high) voltage, but none you should be worried about here. You're probably just seeing the effects of measuring a very low level of current with a piece of equipment that's not designed for measuring that range accurately. If you had such a piece of equipment, I'd wager you'd get the results you initially expected.

There doesn't actually seem to be a datasheet with metering accuracies listed available for the kit you're using, which is a bit disappointing. I guess they expect you to go through the schematics and components yourself, or maybe I'm just not looking in the right place.

Edit: I see there's a voltmeter function in the kit. How are you measuring current?
 
  • Like
Likes Asymptotic
  • #14
milesyoung said:
No, there are significant nonlinear effects that depend on temperature and (high) voltage, but none you should be worried about here. You're probably just seeing the effects of measuring a very low level of current with a piece of equipment that's not designed for measuring that range accurately. If you had such a piece of equipment, I'd wager you'd get the results you initially expected.

There doesn't actually seem to be a datasheet with metering accuracies listed available for the kit you're using, which is a bit disappointing. I guess they expect you to go through the schematics and components yourself, or maybe I'm just not looking in the right place.

Edit: I see there's a voltmeter function in the kit. How are you measuring current?

With a fluke (on ammeter setting) in series with the resistors. I'll build a circuit with about 500mA of current, that should be more than enough to minimize any errors that the instrument gives. Will keep you posted.
 
  • #15
zapnthund50 said:
With a fluke (on ammeter setting) in series with the resistors. I'll build a circuit with about 500mA of current, that should be more than enough to minimize any errors that the instrument gives. Will keep you posted.
That's a lot of current. Make sure you calculate the power you'll be dissipating in each resistor before you turn anything on (they have a maximum power rating).

What Fluke model would that be and what range are you using it on when measuring the 20 microamps?
 
  • #16
milesyoung said:
That's a lot of current. Make sure you calculate the power you'll be dissipating in each resistor before you turn anything on (they have a maximum power rating).

What Fluke model would that be and what range are you using it on when measuring the 20 microamps?

Yeah good point, I'll recalculate it for close to the max power that the resistors can handle. Don't want to fry anything. I'll have to get back to you on the Fluke model, its a good one though, part of the campus equipment.
 
  • #17
zapnthund50 said:
The circuit was built on this breadboard by Digilent Inc (http://store.digilentinc.com/electr...-one-usb-oscilloscope-multimeter-workstation/).
You should be aware that the breadboard system was never intended to be a 'measurement system'. It's a way of testing circuit designs in which the range of values of circuit components are assumed to swamp any of the incidental (parasitic) components. You can't tell what sort of contacts that your components / wires are making. (We still need to see what your actual connections and layout consisted of.) Half of your measuring system is hidden underneath the plastic cover. This is the sort of experiment that would probably best be carried out with wires and good-ol' crocodile clips. I have had classes of A level students, 'verifying' that Ohm's Law applies to out-of-the-drawer resistors (using that crude equipment). They were getting 'good' straight line graphs over a range of 10:1 supply volts. So it's quite possible to do a lot better than your results suggest. Doing it with visible connections would give you a chance to verity that there are no unexpected voltage drops anywhere, due to bad contacts. Fault finding can be a very satisfying exercise and can really help with gaining a deeper level of understanding.
 
  • Like
Likes Asymptotic and Tom.G
  • #18
Were you using the mA or uA setting on the meter?

This Fluke 87 meter http://support.fluke.com/find-sales/Download/Asset/2161164_6116_ENG_B_W.PDF which is fairly common as far as flukes go can't measure many of the values you are trying to measure on the mA setting. The first one is for example will have an accuracy of +-0.2% +-0.04mA. This is more than double the value you measured. The last digit has nearly no meaning except for comparative measurements.

BoB
 
  • #19
rbelli1 said:
Were you using the mA or uA setting on the meter?

This Fluke 87 meter http://support.fluke.com/find-sales/Download/Asset/2161164_6116_ENG_B_W.PDF which is fairly common as far as flukes go can't measure many of the values you are trying to measure on the mA setting. The first one is for example will have an accuracy of +-0.2% +-0.04mA. This is more than double the value you measured. The last digit has nearly no meaning except for comparative measurements.

BoB

The fluke has an automatic current range selector, there is no way to select milliamps or microamps... This was a class experiment that was checked by staff. I don't know the exact model of the fluke, but I will let you know as soon as I have access to it again.
 
  • #20
zapnthund50 said:
The fluke has an automatic current range selector,

That's a pet peeve of mine. Beginners should start out with simple analog instruments.

This will measure 20 microamps with ease. With no ambiguity as to what range it's on.

TFlipflop2.jpg


old jim
 
  • Like
Likes Asymptotic, Tom.G and sophiecentaur
  • #21
jim hardy said:
That's a pet peeve of mine.

It can be set to manual ranging if you need that.

BoB
 
  • #22
jim hardy said:
That's a pet peeve of mine. Beginners should start out with simple analog instruments.

This will measure 20 microamps with ease. With no ambiguity as to what range it's on.

View attachment 197679

old jim
I have such an analog multimeter, present from my dad! They're great, but I'm spoiled by the instant readings digital meters spit out.
 
  • #23
zapnthund50 said:
I have such an analog multimeter, present from my dad! They're great, but I'm spoiled by the instant readings digital meters spit out.
I use them when i want precision
but their least significant digit must not be taken seriously.
 
  • #24
zapnthund50 said:
This was a class experiment that was checked by staff.
I wonder what that consisted of. What level are you working at and are the staff EE specialists? This stuff is harder than people acknowledge.
You should challenge the staff about this. 'How can they expect you to measure such low values of current with that meter?"
On the other hand, they may be wanting you to spot the problem (which you have) and to identify which of your measurements should be used to find the resistance value. (Choice of range of current values)
But I still have a problem with that definite curve you are getting (even when you ignore the lowest value). Inaccuracy at the low current range would hardly be responsible for such a systematic trend.
Did you choose the value of the resistor that you measure or was it randomly picked from a box full?
You have my sympathy. I appreciate that, in the class situation, there is seldom time to do these things 'properly' and you have to take your measurements and then pack away. Much of the time can be involved in just putting the equipment together (often for the first time) and tidying up before you have done any processing of your data. It's always a good idea to get as many measurements in as possible - in this case, using more than just one resistor - and to try graph plotting as you go. I wouldn't mind betting that you would not have had to come to PF if you had been issued with a 500Ω resistor.
 
  • #25
When experimenting
i like to plot my raw data.

Observe that your plot is volts / amps with a very small denominator.
With such a small denominator a teeny error there exaggerates error in the result. Arithmetic just behaves that way.

Let's use hindsight, which is always 20/20
Looks like your resistor turned out to be somewhere around 27k ohms, a standard 5% value.

With 0.1 volts applied to 27kΩ you ought to get 3.7 microamps .
You wrote down 20 microamps. Probably the 2 is least significant digit on your voltmeter so shouldn't be believed. Consider writing instead "Too small to measure".

With 1 volt applied you ought to get about 37 microamps .
You wrote down 50 microamps. Least significant digit on ANY meter is ±1 count, so on your plot instead of a dot, how about drawing a vertical bar from 40 to 60 microamps and labelling it "limit of resolution". Observe that 37 rounds to 40, which is how your meter would report 37 were it a dead-accurate meter.

With 2 volts applied you ought to get 74.1 microamps and you wrote down 80.
Again your meter's resolution is ± 10 microamps so a vertical bar from 70 to 90 would represent what you can measure.
Observe as your measured variable becomes larger compared to your meter's resolution your error band becomes a smaller fraction of what you're measuring.

That's why i like analog meters better than digital for beginners. That the needle is so near zero makes it obvious to the eye you're at the limit of measurability.

We had a saying in my shop: " Twice as many digits makes your errors twice as accurate.".

Form the habit of drawing error bars around any measured data point.
Try it with your experiment - plot your measured currents on a linear scale and draw a best straight line through all the error bars,
see if its slope comes any closer.to the marking on your resistor.
27K is likely marked red-purple-orange, silver stripe = ±10%, gold = ±5%

Beware of small denominators. And overly smart voltmeters.

old jim
 
  • Like
Likes zapnthund50
  • #26
jim hardy said:
Observe that your plot is volts / amps with a very small denominator...
With such a small denominator a teeny error there exaggerates error in the result.

Try it with your experiment - plot your measured currents on a linear scale and draw a best straight line through all the error bars,
see if its slope comes any closer.to the marking on your resistor.
27K is likely marked red-purple-orange, silver stripe = ±10%, gold = ±5%

Beware of small denominators. And overly smart voltmeters.

old jim

Thanks Jim, I'll try the error bar method. More and more I'm thinking what I'm seeing is an error introduced at especially low amperages, just like you're saying. The resistor wasn't a 27k, it was actually a bunch of resistors with an effective resistance of about 31k. I will try the experiment again, but this time with an 150 ohm resistance, that will give me about 33 mA at 5V, which seems like it would be high enough to minimize the meter's uncertainty at the very low amperages (the resistors we are working with are just standard 1/4 W). I'll have access to the lab equipment again on Friday, will let you know my results.
 
  • #27
.
0.
sophiecentaur said:
I wonder what that consisted of. What level are you working at and are the staff EE specialists?

This is one of the last classes for my AS, as I prepared to take my BA for EE. It's called Circuits, and quite a few mechanical engineers are taking it as well, as basic circuit knowledge is recommended for their profession too. The staff there, unfortunately, are not EE specialists (they did bring one into give a talk once), although the lab technicians seems proficient at helping us through each experiment (granted these are simple circuits). The professor teaching this class got his doctorate in mechanical...go figure lol. I expect to have some different results when I try the experiment again with 150 ohms.
 
  • Like
Likes sophiecentaur
  • #28
zapnthund50 said:
The professor teaching this class got his doctorate in mechanical.
Ah yes. He would never have let you do a Hooke's Law experiment with a lead bar. We all develop intuition in our fave subjects.
 
  • Like
Likes jim hardy
  • #29
sophiecentaur said:
We all develop intuition in our fave subjects.

Now THERE is a profound observation... Its converse explains why computers are so counter-intuitive to me...

old jim
 
  • Like
Likes sophiecentaur
  • #30
jim hardy said:
That's a pet peeve of mine. Beginners should start out with simple analog instruments.

This will measure 20 microamps with ease. With no ambiguity as to what range it's on.

old jim

I'm in agreement with Jim. Other benefits (often, with mixed blessings) ...
  • It forces one to think twice about what scale the meter is dialed to, and which jacks the probe leads are plugged into before connecting the meter to a live circuit. Attempting a 480V measurement on the "R X1" range tends to make the leads disappear into a brilliant plasma cloud, and scares the bejeezus out of anyone within earshot.
  • You'll discover the principle of parallax error by learning how to properly read an analog meter, and by extension, will be able to accurately read pressures gauges, and other pointer-type instrumentation.
  • Although digital meters have many advantages, certain measurements are best done with an analog VOM, for instance, "bumping" the shunt and series fields in a compound DC motor to determine their polarity in relation to one another. Do this with an autoranging digital and all you'll see is a brief jumble of numbers, but an analog meter movement clearly reveals the direction of the inductive kick.
  • All voltmeters load the circuit they are measuring - think of the meter as a resistor in parallel with whatever the leads are connected to. Digital meters are high impedance (check your meter specs, but for a typical Fluke, 10 megohms springs to mind). The best that high-end analog meters such as Jim's Simpson 260 or equally venerated Triplett 630 can do is 20,000 ohms per volt (and the cheapest meters could be as bad as 1,000 ohms/volt). A 20KΩ/V meter on the 1 volt range "looks" like a 20KΩ resistor connected into the circuit.
 
  • #31
It is possible that there are no analogue meters available in the OP's lab. Shelf space can be tight in prep rooms and non-specialists can easily turf stuff out which has a value they just couldn't conceive.
 
  • #32
zapnthund50 said:
I expect to have some different results when I try the experiment again with 150 ohms.
Just keep an eye on power, E^2/R

5 volts across 150 ohms would be 25/150 = 1/6 watt so you should be okay. A single 1/4 watt resistor would get noticeably warm, though .
Use all your senses. An old engineer i knew from Tennessee always asked this technical question "Does it fry spit yet?"
Overheated resistors will emit crackling sounds just before the smoke comes out.

These being pretty low ohm resistors they'll have lots of carbon if they are that composition, and carbon having a negative coefficient of temperature you might see some nonlinearity opposite direction from that of metal .

http://hyperphysics.phy-astr.gsu.edu/hbase/Tables/rstiv.html
 
Last edited:
  • #33
jim hardy said:
Just keep an eye on power, E^2/R

5 volts across 150 ohms would be 25/150 = 1/6 watt so you should be okay. A single 1/4 watt resistor would get noticeably warm, though .
Use all your senses. An old engineer i knew from Tennessee always asked this technical question "Does it fry spit yet?"
Overheated resistors will emit crackling sounds just before the smoke comes out.

These being pretty low ohm resistors they'll have lots of carbon if they are that composition, and carbon having a negative coefficient of temperature you might see some nonlinearity opposite direction from that of metal .

http://hyperphysics.phy-astr.gsu.edu/hbase/Tables/rstiv.html
Right, I purposefully chose the 150 value because it was under 1/4 watt for the experiment. Hmm, I don't want ohmic heating to skew my results ... maybe I'll build an equivalent resistance from a couple resistors. Thanks for pointing that out!
 
  • #34
sophiecentaur said:
It is possible that there are no analogue meters available in the OP's lab. Shelf space can be tight in prep rooms and non-specialists can easily turf stuff out which has a value they just couldn't conceive.
That's correct, sadly all they have are digital flukes.
 
  • #35
zapnthund50 said:
I purposefully chose the 150 value because it was under 1/4 watt for the experiment. I don't want ohmic heating to skew my results

Try a larger value instead to keep yourself under half the wattage of the resistors. Also power them then immediately measure as quickly as you can to lessen any heating effects.

zapnthund50 said:
maybe I'll build an equivalent resistance from a couple resistors.

That is a good idea if you need it. For best results keep them separated or in one layer if possible. Twenty resistors paralleled by twisting them all together tightly in a bunch does not get you twenty times the power.

BoB
 
  • Like
Likes Asymptotic
  • #36
When you re-run the experiment, connect a second Fluke meter (set up for voltage) directly across the resistor leads as close to the resistor bodies as possible, and record both these voltages, and voltages read by the on-board Digilent voltmeter. As mentioned upthread, each plug-in breadboard connection introduces an additional resistance (hence, voltage drop) that affects measurements. It would be interesting to see if both voltages track well with one another.

I'm guessing the apparent resistance non-linearity was mostly due to operating beyond the Fluke ammeter's accuracy specs, and won't reoccur during this second run-through using a lower resistance load.

Nevertheless, one thing that stood out looking at the Digilent Electronics Explorer is it has several regulated power supplies and other heat-generating components which I suppose are tucked away under the breadboard itself. A thermoelectic, "Seebeck effect" junction is formed where ever two dissimilar conductors are in contact with one another (this is the basic effect used in thermocouple-based temperature measurement) and it happens within resistors, too. How much voltage offset is generated depends on the materials involved; for every degree C difference between one side of a resistor and the other it could be as high as 400 microvolt/°C for carbon composition while 20 uV/°C is typical for metal film resistors. This may be a factor if one side of the resistor were in contact with the (presumably, warmer) breadboard, while the other side was in open air.
 
  • #37
Results are in! The meter the school provides is a Fluke 175 True RMS Meter. The experiment was done with 6 100-ohm resistors, hooked up to give an an equivalent resistance of 149 ohms, this ensured that each used only a fraction of power, and eliminated the heating variable. Voltage was measured across the ends of the resistor chain (instead of relying on the round number in the computer).

I've attached a picture of the data, along with graphs generated, for all interested to see. The linear fit is excellent, supporting ohms law. The effect of resistance approaching the effective resistance as a limit is ... well, see for yourself. Thoughts?

Edit: note that the scale of the graph of R examines a small portion and is exagerated (ie. about 144-149 ohms), instead of just 0-150 ohms. Thus, the trend is almost gone, but is still clearly present.
 

Attachments

  • ohms law exp.jpg
    ohms law exp.jpg
    46.5 KB · Views: 501
Last edited:
  • Like
Likes jim hardy
  • #38
zapnthund50 said:
I've attached a picture of the data, along with graphs generated, for all interested to see. The linear fit is excellent, supporting ohms law. The effect of resistance approaching the effective resistance as a limit is ... well, see for yourself. Thoughts?
Attached Files:
Your datapoint 1 sure points out the rule "Don't count on that last digit!"
The readings of 0.085V and 0.59mA, when "corrected" in their last digit to 0.086V and 0.58mA yield 148.28 Ohm. In line with the other calculated resistances. Keep this in mind during your journey through Engineering.

It's not limited to just Electrical stuff, ALL measurements have this problem. With mechanical measurement, or any analog measurement for that matter, you get a similar effect with different people doing the reading. Perverse, isn't it?

It appears the meter(s) have opposite calibration issues, although minor. In this case on the low ranges the voltage reading is low and the current reading is high; the combination leading to a lower calculated resistance value. If they are separate meters it would be interesting to see what the low-value readings are if the meters are swapped.
 
  • Like
Likes Asymptotic, zapnthund50 and jim hardy
  • #39
Tom.G said:
Your datapoint 1 sure points out the rule "Don't count on that last digit!"

It appears the meter(s) have opposite calibration issues, although minor..

The first datapoint may very well be iffy, good point. Is it normal for meters to be oppositely calibrated ... that is, reading low on the voltage, and high on the current? I know the effect we're looking at must be generated by the meter, and it seems tied in with the amount of current generated, since the effect was quite pronounced in the first experiment (current on the order of 0.1 milliamps), and almost gone in the second (order of 10 milliamps)... this makes me suspect that something is eating up current and giving us this effect. But how is the meter doing this? It couldn't be the shunt resistor, could it? Guess I might never know lol. Anyway, narrowing the effect down to the instrument is a great start.
 
  • #40
I take it then that both the voltage and current wer taken with the same meter. It is very common for a meter to have a slightly different reading for the same input when on different ranges. The two main causes are
  1. Range changing is done by selecting different resistors within the meter and resistors have a tolerance
  2. The zero adjust may be off a little bit, which effectively adds or subtracts a constant from each reading. For readings near zero, this leads to a large relative error.
It looks like this particular meter has both. Not at all unusual, they aren't perfect.
 
  • Like
Likes Asymptotic
  • #41
Tom.G said:
  • Range changing is done by selecting different resistors within the meter and resistors have a tolerance
  • The zero adjust may be off a little bit
Yes, it was the same meter used for measuring both current and voltage. How does range changing for an ammeter work ... are different shunt resistors selected? Would range changing or a moving zero adjust be able to produce the effect shown, though? It appears to mimic the equation A - e-x ... thanks for your input.
 
  • #42
It occurred to me that the inner workings of the fluke need current to operate, such as powering the LCD. This might cause a tiny voltage drop in the inner workings of the meter, and if this remained uncorrected, could show up in very small readings (where we are measuring tiny voltages). Of course if the meter is able to correct for this, all bets are off.
 
  • #43
zapnthund50 said:
It occurred to me that the inner workings of the fluke need current to operate, such as powering the LCD. This might cause a tiny voltage drop in the inner workings of the meter, and if this remained uncorrected, could show up in very small readings (where we are measuring tiny voltages). Of course if the meter is able to correct for this, all bets are off.

These effects are incorporated into the meter's basic percentage and LSD (Least Significant Digit) accuracy specification. What can be important (but doesn't appear to be involved in your case) is internal burden resistance, which changes depending on current range. For this Fluke 175 the accuracy spec is +/-1.0%, and 3 LSD for all ranges (10.00A, 6.000A, 400.0 mA, and 60.00 mA). All of your current measurements were less than 60 mA; it is likely the meter autoranged to the 60 mA scale. Note that the burden specifications for a Fluke 175 are 37 mV/A on the 6 and 10 amp range, and 2 mV/mA on the lower ranges.

A basic difference between analog and digital meters is an analog meter movement responds to current (more current = more magnetic flux = greater pointer displacement; for a Simpson 260, the movement is 50 microamps full scale) while a digital is at it's heart a voltmeter. When configured for current measurement, a DMM measures voltage dropped across an internal 'burden' resistor placed between the test leads. A burden voltage spec of 37 mV/A suggests a 0.037 ohm resistor is used on the Fluke 175's 6 and 10 amp ranges, and 2 ohm resistor for the 60 and 400 mA ranges. Voltage drop across the burden resistor is scaled as current (using Ohm's law, I=E/R), and displayed by the meter.

A Fluke 87 uses the same principle, and very nearly the same burden resistors for these ranges (they are spec'd as 0.03 V/A, and 1.8 mV/mA), but adds two microamp ranges (6000 uA, and 600.0 uA) at 100 uV/uA burden. This means a Fluke 87 uses a 100 ohm resistor when the microamp ranges are dialed in. 100 ohms may be enough added resistance to significantly affect circuit operation.
 
  • Like
Likes zapnthund50
  • #44
Tom.G said:
Your datapoint 1 sure points out the rule "Don't count on that last digit!"
The readings of 0.085V and 0.59mA, when "corrected" in their last digit to 0.086V and 0.58mA yield 148.28 Ohm. In line with the other calculated resistances. Keep this in mind during your journey through Engineering.

It's not limited to just Electrical stuff, ALL measurements have this problem. With mechanical measurement, or any analog measurement for that matter, you get a similar effect with different people doing the reading. Perverse, isn't it?

It appears the meter(s) have opposite calibration issues, although minor. In this case on the low ranges the voltage reading is low and the current reading is high; the combination leading to a lower calculated resistance value. If they are separate meters it would be interesting to see what the low-value readings are if the meters are swapped.

This link to an article by John Gyork at Design World goes into digital meter accuracy in greater detail. It's too bad this is one of your last classes because it is an excellent introductory exercise in metrology - the science of measurement - and serves as a good jumping-off point to understanding the tangled web between precision, accuracy, resolution, repeatability, reproducibility, and uncertainty.

Extending on the rule Tom quoted, "Don't count on that last digit!", I've calculated your current measurements against Fluke 175 meter specifications.

Fluke_175_Accuracy1.jpg


  • Calculated resistance variation is within specified current meter accuracy.
  • I've used measured voltages in resistance calculations as though they are precisely dead-on, but in actuality they too have their own +/- percentage and LSD accuracy limitations, and would have to be worked out to get a more precise idea of the error budget.
  • The first data point of 0.59 mA is only about 1% of 60 mA full scale. If this were an analog meter movement the pointer would have barely moved from zero.
  • Take note how much more the LSD matters at lower values.
  • 0.59 mA +/- 1% basic accuracy is between 0.584 and 0.596 mA (rounded to a resolution of 0.58 and 0.60 above), but it's the least significant digit spec of '3' added to and subtracted from them that makes a bigger difference (0.55 to 0.63 mA). Calculated resistance at 0.085 V are from 135.8 ohms to 153.4 ohms over this current range. The 147Ω test resistor value falls in between them.
 
  • Like
Likes jim hardy and zapnthund50
  • #45
Asymptotic said:
I've calculated your current measurements against Fluke 175 meter specifications.

View attachment 199069

You are an excellent resource, thanks for running those calculations! The fact that errors grow larger as we near the beginning of the range of the 60 mA scale is particularly interesting. I suppose my last question is, how is the error curve generated? That we may never know. It seems curious that it resembles a function in the family of A - e-x. At any rate, I'm completely satisfied that we are seeing an error introduced by the meter. Thanks to all for your great input!
 
  • #46
Glad we've been of help.
It isn't only that instrument-derived error is greater at the extremes, but everything about low signal measurement is touchier, and makes one pay attention to even the smallest of details in order to account for observations.

Just putting this out there for future readers of this thread. There are few better ways to understand instrument capabilities and limitations that to build them, or barring that, to understand what design considerations go into the process. Joseph J. Carr wrote two excellent books on precisely this, and each are available used for under $20 (although new copies can be more than $200).

"How to Design and Build Electronic Instrumentation"
"Elements of Electronic Instrumentation and Measurements (3rd Edition)"
 
  • #47
It's important to understand your test equipment lest it fool you..

Great thread, guys! Look how much came out of his seemingly simple lab experiment.
A computer simulation wouldn't have taught nearly so much.

We learn best by doing.

old jim
 
  • Like
Likes Asymptotic, Tom.G and zapnthund50