SR830 lock-in amp measurement problem

AI Thread Summary
The discussion revolves around measuring DC-modulated voltages with an SR830 Lock-in Amplifier while applying a DC current. The user is experiencing input overload issues when the DC current is too high, despite attempts to mitigate this with resistors. Participants suggest that using DC coupling may not be necessary and recommend switching to AC coupling, as it should allow for better signal detection at 10 kHz. They also advise implementing a bandpass filter to manage signal levels effectively. The consensus indicates that the current setup may need significant adjustments to avoid internal protection circuitry interference.
notabird
Messages
2
Reaction score
0
Hello

I am currently have difficulties trying to measure dc-modulated voltages while applying a dc current, via Sr830 Lock-in Amplifier (LIA).

Basically, I apply an 2 GHz ac current (amplitude modulated@10kHz and referenced to the LIA for lock-in measurement) and a dc current to my DUT (device under test). This is done through a biased tee. The applied modulated ac current generates a dc-modulated current @10kHz which superposes with the applied dc current, both of which goes into the LIA input. I find that when the dc current is too large, the LIA input overloads and I can't measure anything. I have tried connecting resistors to decrease the dc current drawn by the LIA but it does not seem to work. Appreciate if anyone could help! Thank you.
 
Engineering news on Phys.org
I might be missing something here, but why are you DC couplng the SR830? There is no need to apply any DC to the input.
Unless, you are for some reason referring to the 10kHz as "DC"..
Also, "DC-modulated" makes no sense.
.
Note that the input on the SR830 can go into overload if the TOTAL signal is too large; it is not just the amplitude of the 10 kHz component that matters. You could always try putting a low-pass filter (or ven better a bandpass filter centered around 10 kHz) before the input of the SR830.
 
Last edited:
  • Like
Likes davenn
f95toli said:
I might be missing something here, but why are you DC couplng the SR830? There is no need to apply any DC to the input.
Unless, you are for some reason referring to the 10kHz as "DC"..
Also, "DC-modulated" makes no sense.
.
Note that the input on the SR830 can go into overload if the TOTAL signal is too large; it is not just the amplitude of the 10 kHz component that matters. You could always try putting a low-pass filter (or ven better a bandpass filter centered around 10 kHz) before the input of the SR830.

Thanks. I am using the "DC coupling" setting on the SR830. If I use "AC coupling", I do not get the DUT signal due to excitation from the 10kHz amplitude-modulated 2 GHz ac current. But seems like a band pass filter might work, shall try that thanks.
 
notabird said:
Thanks. I am using the "DC coupling" setting on the SR830. If I use "AC coupling", I do not get the DUT signal due to excitation from the 10kHz amplitude-modulated 2 GHz ac current. But seems like a band pass filter might work, shall try that thanks.

That sounds very strange, I can't remember the cut-off for the AC-coupling but it should certainly be below a few Hz. At 10 kHz you should definitely be able to use AC coupling (I've used SR830s for many years so I am quite familiar with the model); in fact, I can't think of any reason to use DC-coupling at all in this case; any DC-offet could easily saturate the input amps (or at least force you to use less sensitivity).
I assume you are using the single-ended input?
 
According to datasheet, the SR830 is limited to 102.4kHz, the 2GHz signal is way beyond that. Looks like you are getting your AC signal demodulated by internal protection circuitry. I believe, you would need to change setup completely.
 
While I was rolling out a shielded cable, a though came to my mind - what happens to the current flow in the cable if there came a short between the wire and the shield in both ends of the cable? For simplicity, lets assume a 1-wire copper wire wrapped in an aluminum shield. The wire and the shield has the same cross section area. There are insulating material between them, and in both ends there is a short between them. My first thought, the total resistance of the cable would be reduced...
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...

Similar threads

Replies
5
Views
817
Replies
3
Views
2K
Replies
3
Views
2K
Replies
10
Views
3K
Replies
9
Views
4K
Replies
10
Views
2K
Back
Top