c1a3m1
- 1
- 0
Is it wrong to use a thermocouple to mesure the temperature of glass that will range from 20-300oC? as the range of the thermocouple is much higher, does this make it inaccurate? 

on an exponential curve? surely you refer to the indication system (thermocouple and indicator), the pair may have linearity limits?Integral said:Sam,
...you use a thermocouple which has linear response in your region of interest.
Are we confusing accuracy with resolution?sam024 said:the greater the range of the measuring instrument, the less accuracy you will have.
Every instrument is inaccurate ? only the magnitude of the error varies.c1a3m1 said:...range from 20-300oC? as the range of the thermocouple is much higher, does this make it inaccurate?
The response charts I see for T/C is very nearly linear for large portions of the temperature range. None look exponential. Could you please provide a reference? (See my Omega link.) a T/C requires a very sensitive and stable voltage source for a reference (this can be used in place of a second junction and good millivolt meter for the measurement. In this respect they can be a bit hard for a home hobbyist to use. Though you can buy (for a price) temperature compensated temperature controllers which very easy to use.mechie said:on an exponential curve? surely you refer to the indication system (thermocouple and indicator), the pair may have linearity limits?
Are we confusing accuracy with resolution?
Every instrument is inaccurate ? only the magnitude of the error varies.
If you want the best accuracy then you could consider a resistance thermometer as these are generally considered to be better for this kind of temperature range.
Thermocouples are fine if you have a decent indicator - the question now becomes ... "what kind of accuracy do you want?"