Troubleshooting HP 34401a Voltage Meter Load

AI Thread Summary
The HP 34401a voltage meter is experiencing issues with high impedance sensor inputs, where connecting only the ground wire causes significant fluctuations in readings, unlike when only the red wire is connected. The user has confirmed that both leads are used for measurement, and adjusting the input resistance settings has not resolved the issue. It appears that the meter might be affecting the analyzer circuit, potentially due to a higher reference ground. Additional details about the sensor's make and model, as well as the receiving instrument, are necessary for further troubleshooting. The discussion emphasizes the need for more information to diagnose the problem effectively.
ttsky
Messages
20
Reaction score
0
Hi every one,

I have a HP 34401a meter that is causing some issues with a high impedance sensor input.

when I place only the ground wire ( black ) to the input circuit ground, the sensor reading jumps quiet dramatically, this does not happen if i place only the Red wire.

I have tried an AMPROBE 30XR-A with no issues, but it is not sensitive enough for this.

can anyone think of why this could happen. I can provide schematics of the input circuitry if needed, the sensor input is connected to an instrumentation amp.

many thanks.
 
Engineering news on Phys.org
Are you measuring Voltage?

Do you have two leads to make the measurement?
What happens if you connect both leads to the transducer?

Have you selected the input resistance in the Measurement Menu?
 
Baluncore said:
Are you measuring Voltage?

Do you have two leads to make the measurement?
What happens if you connect both leads to the transducer?

Have you selected the input resistance in the Measurement Menu?


Yes two leads, when both are connected i see a rise in sensor reading from the analyzer, and when i take both leads off the reading drops. i can recreate this with only the ground lead. i have selected between 10M and >10G ohms input resistance on the meter, both with same results. it feels as if the meter is at a slightly higher reference ground, and it is effecting the analyzer circuit. but i am not sure if that is possible with only one lead.
 
We really need more information here.
What does the sensor measure. We need to know the make / model of the sensor.
What instrument, make / model, is receiving the signal from the sensor.
 
Very basic question. Consider a 3-terminal device with terminals say A,B,C. Kirchhoff Current Law (KCL) and Kirchhoff Voltage Law (KVL) establish two relationships between the 3 currents entering the terminals and the 3 terminal's voltage pairs respectively. So we have 2 equations in 6 unknowns. To proceed further we need two more (independent) equations in order to solve the circuit the 3-terminal device is connected to (basically one treats such a device as an unbalanced two-port...
suppose you have two capacitors with a 0.1 Farad value and 12 VDC rating. label these as A and B. label the terminals of each as 1 and 2. you also have a voltmeter with a 40 volt linear range for DC. you also have a 9 volt DC power supply fed by mains. you charge each capacitor to 9 volts with terminal 1 being - (negative) and terminal 2 being + (positive). you connect the voltmeter to terminal A2 and to terminal B1. does it read any voltage? can - of one capacitor discharge + of the...