Which Type of Sensor is More Accurate: Thermocouple or RTD?

AI Thread Summary
The discussion centers on the accuracy of K-type thermocouples compared to RTDs, particularly in a temperature range around 90°C. K-type thermocouples typically have an accuracy of ±1.5°C to ±2.5°C at lower temperatures, while RTDs like Pt100 offer greater precision, often around ±0.2%. The accuracy calculations for a K-type sensor in the specified environment suggest a potential error range of ±3.2°C. Users note that thermocouples respond faster than RTDs but sacrifice some accuracy for a broader temperature range. For applications requiring higher precision, Pt100 sensors are recommended despite their higher cost.
Su Solberg
Messages
72
Reaction score
0
I wonder whether the accuracy of that thermometer with a type K sensor is as follow:

encironment: 90'C
Sensor type: resistance sensor
so the accuracy = +/- 90*0.005+0.5 = +/-0.95'C

Thanks for your help.
 

Attachments

  • questions.PNG
    questions.PNG
    5.3 KB · Views: 628
Science news on Phys.org
Su Solberg said:
I wonder whether the accuracy of that thermometer with a type K sensor is as follow:

encironment: 90'C
Sensor type: resistance sensor
so the accuracy = +/- 90*0.005+0.5 = +/-0.95'C

Thanks for your help.

It's a K-type thermocouple:

Usually accuracies are measured as a function of the full scale (FS). In the first instance this would be a full scale of 1350°C (from -50 to 1300). so +- 0.2% of that would be 2.7°C then plus the 0.5°C would be +- 3.2°C. Or it is +- 2.7°C + 0.5, which would be +3.2/-2.2°C So if it reads 90°C it the temperature could actually lie anywhere between 86.8°C and 93.2°C... or between 87.8°C and 93.2°C
 
redargon said:
It's a K-type thermocouple:

Usually accuracies are measured as a function of the full scale (FS). In the first instance this would be a full scale of 1350°C (from -50 to 1300). so +- 0.2% of that would be 2.7°C then plus the 0.5°C would be +- 3.2°C. Or it is +- 2.7°C + 0.5, which would be +3.2/-2.2°C So if it reads 90°C it the temperature could actually lie anywhere between 86.8°C and 93.2°C... or between 87.8°C and 93.2°C

Since the catelog said the calabration (* it has not said measured accuracy) is performed at 18-28 Celsius, may I say the accuracies is not FS any more? just the error is (0.14°C+0.5°C = 0.64°C)

Btw, it is thermocopule type sensor

Thank you very much for your help.
 
No, I don't think so. From what I know, typical k-type thermocouples have between +-1.5°C and +-2.5°C accuracy in the lower temperatures (-50°C to 350°C) and around 0.4 to 0.75% of the temperature for higher temperatures (above 350°C). So, 0.64°C seems a little low for a typical K-type.

if you want something more accurate, i would suggest a Pt100 temperature sensor, although it depends on your situation and what range you want to measure and what accuracy you really require.
 
redargon said:
No, I don't think so. From what I know, typical k-type thermocouples have between +-1.5°C and +-2.5°C accuracy in the lower temperatures (-50°C to 350°C) and around 0.4 to 0.75% of the temperature for higher temperatures (above 350°C). So, 0.64°C seems a little low for a typical K-type.

if you want something more accurate, i would suggest a Pt100 temperature sensor, although it depends on your situation and what range you want to measure and what accuracy you really require.

Thanks for your advise.

Since the Type K thermometer will have +- 1.5 to 2.5'C at -50-350'C, it is poorer than a mecury glass tube thermometer!

The application of the sensor is conduct a flow measurement for compressor by getting pressure difference between a nozzle(measure the mass flow rate and transform back to volume flow). The temperature is around 100'C.

I wonder Pt100's principle and is it much more expensive?

Thanks for your reply.

p.s. the attached is the spec. of the thermometer
 

Attachments

sorry that it took a while to get back to you, i was on holiday.

here is a interesting link for considerations whether to use a thermocouple or a pt100 (a typical type of resistance temperature detector or RTD): http://en.wikipedia.org/wiki/Resistance_thermometer#When_to_use_RTDs_or_thermocouples

Bear in mind that you can log the measurements from a thermocouple or RTD electronically if you want to measure and analyse data over a longer period experiment. Whereas, with a standard thermometer you have to either write down the observed values or set up a video camera to read the results.

Thermocouples react quicker than RTD's (from fractions of a second instead of 2-10seconds), but are less accurate (+-2% instead of 0.2%). Thermocouples also have larger temperature ranges.
 
I was watching a Khan Academy video on entropy called: Reconciling thermodynamic and state definitions of entropy. So in the video it says: Let's say I have a container. And in that container, I have gas particles and they're bouncing around like gas particles tend to do, creating some pressure on the container of a certain volume. And let's say I have n particles. Now, each of these particles could be in x different states. Now, if each of them can be in x different states, how many total...
Thread 'Why work is PdV and not (P+dP)dV in an isothermal process?'
Let's say we have a cylinder of volume V1 with a frictionless movable piston and some gas trapped inside with pressure P1 and temperature T1. On top of the piston lay some small pebbles that add weight and essentially create the pressure P1. Also the system is inside a reservoir of water that keeps its temperature constant at T1. The system is in equilibrium at V1, P1, T1. Now let's say i put another very small pebble on top of the piston (0,00001kg) and after some seconds the system...
I need to calculate the amount of water condensed from a DX cooling coil per hour given the size of the expansion coil (the total condensing surface area), the incoming air temperature, the amount of air flow from the fan, the BTU capacity of the compressor and the incoming air humidity. There are lots of condenser calculators around but they all need the air flow and incoming and outgoing humidity and then give a total volume of condensed water but I need more than that. The size of the...
Back
Top