Relationship between temperature and voltage of Type T thermocouple

AI Thread Summary
The relationship between temperature and voltage for a Type T thermocouple is not strictly linear and depends on the specific setup, including cold junction compensation. A voltage output of 1.5 mV corresponds to approximately 37.5°C (99.5°F), based on data from thermocouple tables. To determine the relationship accurately, calibration is recommended, using ice water and boiling water as reference points. The output voltage is influenced by the temperatures at the thermocouple junction and the cold junction, as well as the purity of the metals used. Understanding these factors is crucial for accurate temperature measurement with Type T thermocouples.
TheAnalogKid
Messages
10
Reaction score
0
I have a friend who was asked what temperature corresponds to an output of 1.5mV from a type T thermocouple.

What is the relationship and how do you get the relationship? This is an academic problem, so datasheets aren't really the answer I'm looking for.
 
Engineering news on Phys.org
you could calibrate the thermocouple.
put it in iced water, along with another thermometer. read the mV's and note the temp
then boil the water, read the temp, note the mV's.
i think the response is linear, but it may not be.!
 
The might help - http://www.omega.com/temperature/Z/pdf/z223.pdf and 1.496 mV corresponds to 99°F (37.2°C).

It looks like it's supposed to read ~0 V at freezing or ~32°F.

I would recommend what phlegmy suggested. Try it in ice water and boiling water as a test.

Type T (Copper Constantan) thermocouples.
http://www.veriteq.com/thermocouple-data-logger/type-t.htm

The Principles and Methods of Using Thermocouples
http://www.iotech.com/mcjun96.html

http://srdata.nist.gov/its90/download/type_t.tab (in Celsius).

Looking at this last dataset, 1.5 mV is between 1.486 and 1.528 mV, which correspond to 37 and 38°C, so 1.5 mV ~ 37.5°C (99.5°F)
 
Last edited by a moderator:
The output of a thermocouple depends on the temperature of the termocouple ( the point where the two wires are joint ) and the temperature where the two wires connect to the circuit ( the "cold junction" ). The output is a function ( linear ? , I don't know ) of the difference of temperatures. So there is no answer to your question.
 
It depends on the set up. Most TC systems have a cold junction compensation function. However, the OP should specify what the actual set up is. If there is a bare bones set up, the cold junction compensation is likely to be forgotten. Then the tables will be useless.
 
Also, the purity of the metals in the TC will make a difference in the potential it produces.
 
Hey guys. I have a question related to electricity and alternating current. Say an alien fictional society developed electricity, and settled on a standard like 73V AC current at 46 Hz. How would appliances be designed, and what impact would the lower frequency and voltage have on transformers, wiring, TVs, computers, LEDs, motors, and heating, assuming the laws of physics and technology are the same as on Earth?
While I was rolling out a shielded cable, a though came to my mind - what happens to the current flow in the cable if there came a short between the wire and the shield in both ends of the cable? For simplicity, lets assume a 1-wire copper wire wrapped in an aluminum shield. The wire and the shield has the same cross section area. There are insulating material between them, and in both ends there is a short between them. My first thought, the total resistance of the cable would be reduced...
I used to be an HVAC technician. One time I had a service call in which there was no power to the thermostat. The thermostat did not have power because the fuse in the air handler was blown. The fuse in the air handler was blown because there was a low voltage short. The rubber coating on one of the thermostat wires was chewed off by a rodent. The exposed metal in the thermostat wire was touching the metal cabinet of the air handler. This was a low voltage short. This low voltage...