Acetylene Safety: Safe Transfer & Storage Tips

  • Thread starter Thread starter rwooduk
  • Start date Start date
  • Tags Tags
    Safety
AI Thread Summary
The discussion centers around the need for acetylene to calibrate a gas chromatograph (GC). The user seeks advice on acquiring and safely transferring acetylene, considering options like obtaining it from another department or generating it from calcium carbide. Generating acetylene by adding water to calcium carbide is noted as a standard procedure, although safety precautions are emphasized due to the flammable and potentially explosive nature of acetylene. The conversation highlights the importance of purity for calibration purposes, with users questioning whether pure calcium carbide would yield pure acetylene. The method of capturing acetylene in a balloon for transfer is suggested, along with concerns about transporting it in a syringe. The user aims to establish a calibration curve for acetylene analysis to quantify its presence in samples containing short-chain hydrocarbons. Overall, safety, purity, and effective transfer methods are the main topics of discussion.
rwooduk
Messages
757
Reaction score
59
Hi, I need to get some acetylene so I can calibrate our GC, the university has some in another department which I could possibly ask to have. I only need to inject very small quantities into the GC. I was wondering if anyone is familiar with this gas?

If I ask the other department for it, I need to suggest how I will transfer it and store it safetly, therefore I would really appreciate some ideas on this. Could I dissolve it in a liquid? But I need a pure sample for our GC so we can determine if there is any in our sample.

Thanks for any help as I am not a chemist.
 
Chemistry news on Phys.org
No idea about pure. But my first instinct is to make it from calcium carbide.
 
  • Like
Likes rwooduk
Borek said:
No idea about pure. But my first instinct is to make it from calcium carbide.

Many thanks for the reply! You mean instead of going to the other department, make it instead? Is it an easy proceedure? I can get any chemicals I need through the University, but as I say I'm really not a chemist. Would it be dangerous?
 
This is pretty easy and pretty standard procedure - calcium carbide is a solid, just add some water and it will nicely bubble. Standard precautions will be perfectly adequate - the leftover is caustic and the acetylene is flammable (perhaps even explosive is some concentrations), so gloves/glasses and ventilation are required, but there is nothing extraordinary here.

As I said - I have no idea about purity of the acetylene made this way, a lot will depend on whether the carbide is a pure one, or not.

From what I remember acetylene is not easy to transport - if pressurized it can easily react with itself, the reaction can be highly exothermic. Thus it is not transported as a gas, but rather dissolved (under pressure) in acetone. Thus making it in situ can be easier than other approaches.
 
  • Like
Likes rwooduk
Ok, that's great, thanks very much for your help.
 
Why do you need acetylene to calibrate a GC? What does calibrate mean anyway? Are you making a calibration curve for acetylene analysis? What are you running as samples and standards?
 
Since GC is a separation technique purity is really not an issue, but you would need UPC acetylene. I really don't know of any way to get acetylene into crimped vial, without air. Perhaps you can fill a baloon, put a piece of tape on it, and insert your syringe through the tape to keep the baloon from popping.
 
We need to know the retension time for acetylene so that when we inject our sample (which will likely contain the substance) we will be able to identify how much is there by its peak, which will be in the same place as a pure sample. This would be identification and calibration using the pure would tell us the amount. Sample is water containing short chained hyrocarbons and some other products.

Yes I think the bubble idea will work, there's a youtube video here:



Thanks for the comments
 
ok we finally have our GC ready to start calibrating, I asked above about the purity of the acetylene via this method (balloon over water calcium carbide mix), does anyone else have any idea of the purity?

^^ would pure calcium carbide give pure acetylene?

also just one more thing on safety, would I be safe taking it into a syringe (as suggested above) and running around with it in a syringe to the location of the GC?
 