Calculating Photomultiplier Tube Saturation Limit

  • Thread starter Thread starter watertreader
  • Start date Start date
  • Tags Tags
    Tube
AI Thread Summary
The discussion focuses on determining the saturation limit of photomultiplier tubes (PMTs), specifically regarding the maximum light intensity they can handle without damage. Participants highlight that the primary specification for assessing saturation is the maximum current produced per luminosity, but caution that manufacturers typically provide detailed damage thresholds. One user shares their experience of damaging a PMT by exposing it to excessive light and suggests considering photodiodes as alternatives for high-intensity applications. There is also a conversation about monitoring the temperature within the circuit box housing the PMT, as elevated temperatures can accelerate aging. Overall, users emphasize the importance of careful handling and awareness of the PMT's specifications to avoid damage.
watertreader
Messages
21
Reaction score
0
Hi,

Have some query on the photomultiplier tube saturation limit(the maximum light luminosity or intensity that can be delivered to the PMT without damaging the equipment). How do we calculate it?

I have looked through the specification. The only relevant specification for the saturation problem is the maximum current produced per luminosity. Is there some other gauge we should take to measure the maximum allowable amount of input light? What's the common rating?

Wouldn't want to spoil another PMT again
 
Engineering news on Phys.org
I don't think you are able to easily calculate the damage threshold of a PMT. This type of information is usually provided by the manufacturer, and as far as I know it takes a considerable amount of light flux (relatively speaking) to damage a PMT. PMTs are used for very sensitive and low intensity applications. If you are damaging PMTs with to high of a light intensity then you should look at photodiodes or phototransistors as a possible substitute.
 
I have measured a peak pulse of more than 1 volt across 50 ohms (20 ma) out of a RCA 6810, but the average output should be less than 1 ma. Are you using a dynode or anode output? As for the cathode, you probably should not expose it to sunlight with no voltage on it, and never when any voltage is on it. What tube are you using, and what gain? How did you spoil the first one?
 
I am using the anode output, using a Hamamatsu R928... It is sited inside a circuit box that I have fabricated(with all the accessories electrical circuits for DC Measurement and an input fiber SMA fiber port)...using a gain at around 400-500V

Guess I have pumped in round 300-500uW of laser power through the fiber onto the cathode detector area(tat's why I like to know the damage threshold)... Another fatal mistake is that I have not switched off the room's light...
(Slipped my mind to off the light)

THe PMT was working for some time... but the next time round I have used them... I can't detect any DC signal anymore...

Guess I should try to be careful the next time round...
 
Just curious. Have you ever measured the temperature inside your "circuit box"?
The tubes Operating Ambient Temperature is -30 to +50 °C. Usually devices, that are operating at a temperature above their rated value, will age much quicker. Don't know if this is a problem with your tube.
 
No, I didn't. Have another one PMT(other model) in an even smaller box... seem to be working well for me...

BTW, out of interest... to monitor temperature in a box... what circuit did you used? A thermostat?
 
I would use a thermocouple with an associated meter. Or maybe just a glass thermometer.
 
Hi

I am also using Hamamatsu R928 and also exposed it accidentally to fluorescent light, with about a KV on it. After one night without voltage it resumed its capabilities.

Does someone know this device intensity limitation? Do you know at which range its response is linear with intensity?

Thanks
 