The discussion focuses on troubleshooting data acquisition using the National Instruments PCI-6025E board with LabView. The user initially calculates a resolution of 4.9 mV based on the board's specifications but observes a resolution of approximately 2.5 mV in practice. It is identified that the board is configured in single-ended mode, which effectively doubles the resolution. Additionally, there are inquiries about how the DAQ assistant wizard manages input ranges and whether it adjusts for non-standard ranges. The conversation concludes with a suggestion to contact NI support for further clarification on software-related errors.
jamesrc
Science Advisor
Gold Member
Messages
471
Reaction score
1
Hi,

I've got a little mental block with something I'm working on and was hoping someone could help out:

I'm doing some data acquisition with a National Instruments PCI-6025E board using LabView (not a lot of experience with LabView by the way). Anyway, the board is supposed to be a 12-bit board and the input range (for the analog inputs) is supposed to be -10 to +10 V. If I do the math, my resolution should be 20/(2^12) = 4.9 mV, right? Why, then, when I read data using the program does it appear that my resolution is twice that (I see steps in the data recording spaced ~2.5 mV apart). Any hints on what I'm doing wrong?

Thanks.
 
Engineering news on Phys.org
Any chance you have it configured in single-ended mode? That will double the resolution. Or maybe it's in +/-5V range mode?

Your math appears correct for the +/-10V range.
 
That was it - it's in single ended mode.

Thanks for the help.

While I've got you and since you mentioned range, do you happen to know how the ranging works when you're using the DAQ assistant wizard (I know, this is what I get for using a black box component). I mean, I think the board specs allow for +/-50mV to +/-10V ranges; if you enter in the range you are using, will it adjust accordingly? What if you choose a range that isn't one of the board selectable ranges? Like if I enter +/-9V (not that I planned to) will it behave the same as if it were +/-10V? Thanks again.
 
Glad to help. I'm not of any help on your follow-up question unfortunately. But NI has pretty good support, so you should be able to just give them a call on Monday to ask about what kind of errors get thrown by the different parts of their software.
 
OK. Thanks again.
 
Thread 'How can I find the cleanout for my building drain?'
I am a long distance truck driver, but I recently completed a plumbing program with Stratford Career Institute. In the chapter of my textbook Repairing DWV Systems, the author says that if there is a clog in the building drain, one can clear out the clog by using a snake augur or maybe some other type of tool into the cleanout for the building drain. The author said that the cleanout for the building drain is usually near the stack. I live in a duplex townhouse. Just out of curiosity, I...
Thread 'Where is my curb stop?'
My water meter is submerged under water for about 95% of the year. Today I took a photograph of the inside of my water meter box because today is one of the rare days that my water meter is not submerged in water. Here is the photograph that I took of my water meter with the cover on: Here is a photograph I took of my water meter with the cover off: I edited the photograph to draw a red circle around a knob on my water meter. Is that knob that I drew a red circle around my meter...

Similar threads

Replies
7
Views
3K
Replies
5
Views
6K
Replies
3
Views
3K
Replies
4
Views
2K
Replies
6
Views
5K
Back
Top