Detecting scintillation flash using semiconductors

Click For Summary

Discussion Overview

The discussion centers around the detection of scintillation flashes using semiconductor devices, particularly focusing on alternatives to photomultiplier tubes (PMTs) for detecting photons emitted from scintillation crystals. Participants explore various semiconductor technologies, their applications, and the feasibility of building simple detectors for specific uses, such as counting atmospheric muons.

Discussion Character

  • Exploratory
  • Technical explanation
  • Debate/contested
  • Experimental/applied

Main Points Raised

  • Some participants mention that Avalanche Photo Diodes (APDs) in Geiger mode can detect single photons and may have better quantum efficiency than PMTs.
  • There is a query about the availability of commercial radiation detectors that utilize APDs, noting that these diodes are relatively recent developments.
  • Solid-state devices, such as Germanium detectors, are discussed as alternatives to scintillation detectors, with claims that they provide superior performance and energy resolution.
  • Concerns are raised about the cooling requirements for certain Germanium detectors, which may limit their practicality for some applications.
  • One participant suggests that while Ge(Li) or HPGE detectors have excellent resolution, they may not be suitable for detecting cosmic rays due to cost and the nature of the radiation.
  • There is a discussion about the costs associated with building a detection system, including the necessary components and electronics, which may not significantly differ whether using PMTs or semiconductor alternatives.
  • Some participants argue that for detecting muons, simpler gas detectors may be more appropriate than scintillation detectors or expensive solid-state options.
  • One participant shares their experience building photon counting APD units and questions whether scintillation detectors are still superior for large area coverage.

Areas of Agreement / Disagreement

Participants express a range of opinions on the effectiveness and practicality of various detection methods, with no clear consensus on the best approach. There are competing views on the suitability of solid-state detectors versus traditional methods like PMTs and gas detectors.

Contextual Notes

Participants note limitations related to the cost of components, the need for cooling in certain detectors, and the specific requirements for detecting low levels of light from scintillation events. The discussion also highlights the varying performance characteristics of different detection technologies.

Who May Find This Useful

This discussion may be of interest to those exploring radiation detection technologies, particularly in the context of semiconductor applications, as well as individuals considering DIY projects for detecting cosmic rays or muons.

waht
Messages
1,502
Reaction score
4
Although photomultpliers are the most sensitive light detectors, are there any other radiation counters that utilize photo-transistors, or other means, to detect photons emitted form a scintillation crystal when hit by incoming radiation?
 
Engineering news on Phys.org
Avalanche Photo Diodes (APDs) in geiger mode can easily do single photon and are often better QE than PMTs
 
Cool didn't know that.

Just wondering if there are any commercial radiation detectors that already utilize these diodes? According to google, avalanche diodes were developed quite recently.
 
There are solid state devices (semiconductor detectors such as Germanium), that are fundamentally different from scintillation detectors and have no need for photomultipliers.
 
what said:
Just wondering if there are any commercial radiation detectors that already utilize these diodes? According to google, avalanche diodes were developed quite recently.
what,

Solid state radiation detectors have been the NORM for nuclear laboratories for DECADES
because they are far SUPERIOR to gas-filled [Geiger] and scintillation detectors. For example,
the solid-state detectors also give you an accurate reading of the energy of the detected radiation;
as opposed to something like a Geiger detector where the device saturates for any radiation above
the trigger level.

Commercial detectors are available from such companies as AMETEK subsidiary ORTEC.

http://www.ortec-online.com/detectors/photon/detectors.htm

Courtesy of Lawrence Berkeley National Laboratory:

http://sensors.lbl.gov/sn_semi.html

Lawrence Livermore developed a portable spectrum analyzer detector capable of determining
the identity of the radioactive nuclei for Homeland Security applications called RadScout:

https://ipo.llnl.gov/technology/profile/radscout/

This was commercialized as the ORTEC "Detective" series:

http://www.ortec-online.com/psis.htm
http://www.ortec-online.com/detective-ex.htm

Dr. Gregory Greenman
Physicist
 
Last edited by a moderator:
That's neat. I've been going over some of those germanium designs. Those ortec meters are really advanced, nice.

I'm asking because I'm exploring a possibility to build a simple detector to count muons from the atmosphere. I don't want to use geiger nor photomultiplier tubes. Instead I'd like to replace photomultiplier with some semiconductor photo-diode off the shelf, sensitive enough to pick up flash a from the scintillation crystal.

Some of those germanium detectors need to be cooled with liquid nitrogen, so I'm not sure if this is possible (off the shelf).
 
what said:
Some of those germanium detectors need to be cooled with liquid nitrogen, so I'm not sure if this is possible (off the shelf).
what,

As I recall; there are a couple types of solid-state germanium detectors; "lithium-drifted" and "intrinsic".

When I took a nuclear measurements lab course at MIT, our instructor told us he only purchased the
"intrinsic" germanium detectors. Both the intrinsic and lithium-drifted detectors need to be cooled with
liquid nitrogen in order to work. However, if the lithium-drifted detectors ever warm-up; they are ruined;
and they are not inexpensive.

I recall my professor remarking that although the intrinsic detectors were more expensive; he didn't have
to worry about a student forgetting to fill the dewar and ending up with a ruined detector. The intrinsic
ones won't work unless they are cooled - but at least if someone fails to fill the dewar, they are not
ruined.

Dr. Gregory Greenman
Physicist
 
GeLi detectors have fantastic resolution. Of course I didn't have to pay for the crystals or the liquid nitrogen!

Anyhow, as I understand this thread is really about muons.
Muons are not directly or easily detected. It more of a research project.
Here's a link to an interesting article on Muon detection:

http://www-ppd.fnal.gov/EPPOffice-w/Academic_Lectures/Denisov_%20Lecture.pdf

Is sounds like a fun, but personally I'd suggest gaining some experience with cloud chambers first.
 
Last edited by a moderator:
  • #10
Let's get back to the OP's original question and application.

The energy resolution of a Ge(Li) or HPGE detector won't help him with cosmic rays and scintillator, as everything he's looking at is minimum ionizing anyway. This isn't worth the expense.

I assume that the OP wants to avoid PMTs because of cost. I'm afraid that this will be hard to do. A tube is a few hundred dollars, but it's probably $2000 for a setup. You need a tube, and a preamp, and an amplifier-shaper-discriminator, and a HV source for the tube, and a LV source for the preamp, and all this adds up. (The cost for a second tube, of course, is smaller, since several of these components can be shared) Replace the tube by an APD or a silicon PMT, and you still need all the periphery. Different actual components, to be sure - for example the voltage requirements are different - but the total system cost for a one channel system is not grossly different.

Understand you are looking at very, very low levels of light here. Perhaps a hundred photons.
 
  • #11
Vanadium 50 said:
The energy resolution of a Ge(Li) or HPGE detector won't help him with cosmic rays and scintillator, as everything he's looking at is minimum ionizing anyway. This isn't worth the expense.

I assume that the OP wants to avoid PMTs because of cost. I'm afraid that this will be hard to do. A tube is a few hundred dollars, but it's probably $2000 for a setup. You need a tube, and a preamp, and an amplifier-shaper-discriminator, and a HV source for the tube, and a LV source for the preamp, and all this adds up. (The cost for a second tube, of course, is smaller, since several of these components can be shared) Replace the tube by an APD or a silicon PMT, and you still need all the periphery. Different actual components, to be sure - for example the voltage requirements are different - but the total system cost for a one channel system is not grossly different.

Understand you are looking at very, very low levels of light here. Perhaps a hundred photons.

Beside the cost of PMT, the electronics to interface a PMT is very simple. I'm EE major. But I guess the low level photons we're dealing with is going to be an issue.
 
  • #12
Simple, but it still costs money. You need some low-noise amplifiers, for example.

I built a 432 channel system, and it ended up with a per-channel cost of $400. It was $150 per tube, $150 for on-detector electronics, and $40k for off-detector electronics. This was generally considered quite cheap.
 
  • #13
Vanadium 50 said:
The energy resolution of a Ge(Li) or HPGE detector won't help him with cosmic rays and scintillator, as everything he's looking at is minimum ionizing anyway. This isn't worth the expense.

Indeed, detecting muons with Ge xtals is pretty silly (and expensive). What you need is a relatively large detection volume, and you don't need precise detection, so I would opt for a gas detector. No need to go through scintillation. A big tube with a wire in it under HV, and you don't need much. You will need the gas filling (and hence access to a vacuum pump and such) and a tight container. But then for a simple Geiger tube, you don't even need much electronics (a HV power supply, some resistors, a HV capacitor and then a speaker or a pulse counter).
 
  • #14
I built photon counting geiger mode APD units 10years as a grad student for about $500.
They had a fibre pigtailed APD from EG+G Canda that was about $250 at the time, a little single stage peltier cooler to get the APD down to 250k and some circuitry that quenched the diode, and output TTL pulses- it counted photons upto about 3Mhz. The unit was about the size of an altoids tin.


I'm guessing that scintillation detectors are still better than semiconductors if you want large area coverage ?
 
  • #15
mgb_phys said:
I built photon counting geiger mode APD units 10years as a grad student for about $500.
They had a fibre pigtailed APD from EG+G Canda that was about $250 at the time, a little single stage peltier cooler to get the APD down to 250k and some circuitry that quenched the diode, and output TTL pulses- it counted photons upto about 3Mhz. The unit was about the size of an altoids tin.

How did you interface the fiber to a scintillation crystal?
 
  • #16
what said:
How did you interface the fiber to a scintillation crystal?
I didn't - these were for astronomy.

I haven't worked with particle physics detectors, but the ones I have seen are large (1m long) complicated curved bits of plastic that end in an output face a couple of cm square.
You could couple that into a fibre fairly easily. The fibres that came with these diodes were 200-250um multimode (don't remember the NA).

I really just meant that APDs are very easy to work with compared to PMTs.
 
  • #17
mgb_phys said:
I didn't - these were for astronomy.

I haven't worked with particle physics detectors, but the ones I have seen are large (1m long) complicated curved bits of plastic that end in an output face a couple of cm square.
You could couple that into a fibre fairly easily. The fibres that came with these diodes were 200-250um multimode (don't remember the NA).

I really just meant that APDs are very easy to work with compared to PMTs.

I think I saw such a paddle in fermilab, but it was coupled with a PMT whose surface area of contact was many times bigger than that of a fiber. I'm going to do more research on APDs.
 
  • #18
what said:
I think I saw such a paddle in fermilab, but it was coupled with a PMT whose surface area of contact was many times bigger than that of a fiber.
That's the advantage of PMTs of course.
I wouldn't have though it was too difficult to couple to an APD, you can get large area ones but sensitivity and time response go down with area.
Obvious ways to couple to a scintillator would be either a microscope objective or I suppose (having never tried it!) get a similar shaped bit of plastic, heat it to softening point and draw it out to neck down to a thin stand. Cleave the end and join to the fibre with index matching glue and polish the other end and stick that to the scintillator with index matching gel.
 

Similar threads

  • · Replies 25 ·
Replies
25
Views
5K
  • · Replies 24 ·
Replies
24
Views
3K
  • · Replies 3 ·
Replies
3
Views
2K
  • · Replies 4 ·
Replies
4
Views
3K
  • · Replies 0 ·
Replies
0
Views
902
  • · Replies 1 ·
Replies
1
Views
1K
  • Poll Poll
  • · Replies 1 ·
Replies
1
Views
5K
  • · Replies 3 ·
Replies
3
Views
3K
  • · Replies 14 ·
Replies
14
Views
3K
Replies
8
Views
5K