Electromagnet heat source and efficiency

AI Thread Summary
Electromagnets used in biomedical applications face challenges with heat generation, primarily from winding losses and core losses. Winding losses, caused by current flowing through the coil, can be minimized by reducing current or optimizing materials and coil design, though this may affect magnetic flux density. Core losses arise from the properties of the ferromagnetic material and include losses from magnetic domain movement and eddy currents, particularly under alternating current conditions. In static applications using direct current, winding losses are the main source of heat. Overall, strategies to enhance electromagnet efficiency focus on balancing current, resistance, and coil design to mitigate heat production.
anorred
Messages
58
Reaction score
0
So.. I'm a biomedical engineer, and a new technology that is arising is being able to control magnets that "swim" around the body using electromagnets oriented around the outside of the patients body. It has been said that electromagnets can't be used on the inside of the patient because they get too hot.

What is the source of the majority of the heat? Is it the current moving through the wire, or the current induced in the core of the electromagnet?

Are there methods to make electromagnets more efficient to reduce heat?
 
Engineering news on Phys.org
In a electromagnet one can basically divide the power losses into two categories, winding losses and core losses.

Winding losses are caused by the current flowing in the coil/winding of the electromagnet. These losses are equal to : P=I^2 R, where I is the current, and R is the resistance of the winding. To reduce the winding losses, the most effective method, due to the ^2 term, is to reduce the current. This again may reduce the magnetic flux density produced.
Or one can change the resistivity of the winding by changing the material (copper, aluminium etc.) or the length/number of turns of the winding. The latter also reduces the magnetic flux density.

Since the magnetic flux density is proportional to the current (I) times the number of turns(N), hence F=I*N, called the magnetomotive force (F). One way to reduce the winding losses, but still give a high magnetic flux density is to, keep in mind the P=I^2*R equation, increase the number of turns and reduce the current, keeping F constant. Thus, decreasing I but increasing R due to the increased length/number of turns. Observing that I is squared, follows that the power losses decreases.

Other losses associated are the core losses. These are related to the ferromagnetic material making the core (magnetic path of the electromagnet), if the magnet has a core.
https://en.wikipedia.org/wiki/Magnetic_core#Core_loss
The core losses are separated by the magnetic domain movement and eddy currents (induced currents) in the core. The magnetic domain movement are proportional to the changing magnetic flux density squared times the amplitude of the flux density.

Eddy currents are a consequence of a changing magnetic flux density in a conductive material. And the associated losses are proportional to the current density squared and the conductivity.

The above losses is mostly occurring using AC (alternating current) for the winding excitation, or if the material is moving in a magnetic field. Or if the external field is alternating.

An electromagnet can have both AC and DC winding excitation, that is application specific. In your case, the option is to use permanent magnets, so I would conclude that the electromagnet is powered by DC. But using DC excitation, the core losses is minimal if no movement or external field is applied. So in that case, I would say that the winding losses are the most pronounced source of heat.
 
Very basic question. Consider a 3-terminal device with terminals say A,B,C. Kirchhoff Current Law (KCL) and Kirchhoff Voltage Law (KVL) establish two relationships between the 3 currents entering the terminals and the 3 terminal's voltage pairs respectively. So we have 2 equations in 6 unknowns. To proceed further we need two more (independent) equations in order to solve the circuit the 3-terminal device is connected to (basically one treats such a device as an unbalanced two-port...
suppose you have two capacitors with a 0.1 Farad value and 12 VDC rating. label these as A and B. label the terminals of each as 1 and 2. you also have a voltmeter with a 40 volt linear range for DC. you also have a 9 volt DC power supply fed by mains. you charge each capacitor to 9 volts with terminal 1 being - (negative) and terminal 2 being + (positive). you connect the voltmeter to terminal A2 and to terminal B1. does it read any voltage? can - of one capacitor discharge + of the...