Experimentalists : sample of gas from cylinder

In summary: The TC gauge and a readout like the Hasting gauge readouts can be gotten on Ebay for around $200...It can read fairly accurately down to a few millitorr, for roughing vacuum.You probably don't need to get really serious and use a cryo pump or diffusion pump or turbo pump which can run into big bucks.
  • #1
rwooduk
762
59
Hi, are there any experimentalists here? I have gas in a cylinder. I have just been taking samples of gas from a tube connected to the outlet, but air is getting in the syringe.

So I went over to chemistry and got a round bottom flask with 3 inlets / outlets and a septum, so its like this:

5RsqaIH.jpg


I evacuate the flask with the vacuum then flow the gas in, then after about a minute I pierce the septum with the GC syringe and take in some gas.

I am still getting air in the gas sample! Does anyone know a standard way of extracting gas from a cylinder (into a GC syringe) without air getting into it?

Any help would be really appreciated. I can order glassware etc etc but I don't know what specific equipment I need.
 
Chemistry news on Phys.org
  • #2
There will be air in the syringe before you even take your sample . Even with the piston fully depressed there will be some air in the dead space and in the needle .

If you have ample supplies of test gas then suck gas in and vent to atmosphere or a collecting vessel several times before collecting the actual sample .You'll need some arrangement of lab grade valves to switch the flows .
 
  • #3
Thanks for the reply. When I put the syringe through the septum I suck gas in then empty the syringe, would that not get rid of the air in the syringe?

I have a grease seal over the inlet and outlet valves so I am pretty sure air is not getting in there, the only thing that concerns me is the septum. Perhaps when I apply the vacuum air is sucking into the previously made holes?

Thanks again for the reply, and I welcome any more advice.
 
  • #4
You also have to make sure you use vacuum grease like Apiezon or Dow Corning high vacuum grease. Also, you need to monitor the level of vacuum you are pulling out of the flask. You might think pumping to 1 torr would be ok but maybe you need to get to a few millitorr to reduce contamination. 1 atmosphere is 760 torr so at 1 torr you still have 1/760th of the contamination from atmosphere, which may contribute to analysis errors. Too bad you can't use some kind of laser analysis, that would obviate the need to extract gas at all, just beam the laser through the flask and read out the result. Not sure if that is a viable technology but worth a thought.
So no matter how many times you pre pump and such, the sample, if the flask itself is not pumped down sufficiently there will still be contamination from atmosphere so you may not be pumping enough. Like my example of pumping to 1 torr, that still leaves the possibility of 1.3 parts per thousand of air and moisture left in the mix. So you have to understand just how much contamination the analysis can tolerate to get the level of data required.

A TC gauge and a readout like the Hasting gauge readouts can be gotten on Ebay for around $200 and it can read fairly accurately down to a few millitorr, for roughing vacuum. You probably don't need to get really serious and use a cryo pump or diffusion pump or turbo pump which can run into big bucks.
You probably need a real roughing pump though, most of which can easily get down to 10 millitorr or so.
1 torr=1.3 parts per thousand. 1 milltorr, contamination now down to 1.3 parts per million so choose your weapon well!
 
  • Like
Likes rwooduk
  • #5
litup said:
You also have to make sure you use vacuum grease like Apiezon or Dow Corning high vacuum grease. Also, you need to monitor the level of vacuum you are pulling out of the flask. You might think pumping to 1 torr would be ok but maybe you need to get to a few millitorr to reduce contamination. 1 atmosphere is 760 torr so at 1 torr you still have 1/760th of the contamination from atmosphere, which may contribute to analysis errors. Too bad you can't use some kind of laser analysis, that would obviate the need to extract gas at all, just beam the laser through the flask and read out the result. Not sure if that is a viable technology but worth a thought.
So no matter how many times you pre pump and such, the sample, if the flask itself is not pumped down sufficiently there will still be contamination from atmosphere so you may not be pumping enough. Like my example of pumping to 1 torr, that still leaves the possibility of 1.3 parts per thousand of air and moisture left in the mix. So you have to understand just how much contamination the analysis can tolerate to get the level of data required.

A TC gauge and a readout like the Hasting gauge readouts can be gotten on Ebay for around $200 and it can read fairly accurately down to a few millitorr, for roughing vacuum. You probably don't need to get really serious and use a cryo pump or diffusion pump or turbo pump which can run into big bucks.
You probably need a real roughing pump though, most of which can easily get down to 10 millitorr or so.
1 torr=1.3 parts per thousand. 1 milltorr, contamination now down to 1.3 parts per million so choose your weapon well!

Hi, Many thanks for the detailed reply! Yes I have ordered some vacuum grease also I have got a couple of taps from chemistry, so it's like this:
hiMV69F.jpg


I think the issue was that without the taps after turning off the vacuum the air was rushing back in! I will look into how to monitor the pressure / vacuum level in the flask, thanks! I can't use a laser as the sample needs to be injected into a gas chromatograph to calibrate the machine. Yes, a TC gauge sounds interesting, I will look into it! Thanks again!
 
  • #6
Update, for those who have the same problem, this set up is working very well:

z3rBoi6.jpg


One goes to vacuum, other puts CO2 in.
 
  • Like
Likes Nidum
  • #7
Do you have TC gauges yet to measure vacuum levels? What you show looks ok from vacuum standpoint. The little blue knobs looks like they are attached to valves, but are unfamiliar to me. Most of the valves I use are pneumaticvalves, air pressure activated, some for high vacuum are bellows valves and some needle valves.
 
  • #8
litup said:
Do you have TC gauges yet to measure vacuum levels? What you show looks ok from vacuum standpoint. The little blue knobs looks like they are attached to valves, but are unfamiliar to me. Most of the valves I use are pneumaticvalves, air pressure activated, some for high vacuum are bellows valves and some needle valves.

To be honest we are low on funds and the above setup is working fine. We aren't really measuring the pressure levels (hence the film over the vessel incase it explodes / implodes) but its fairly safe, you just have to be careful when opening the blue taps (they are just air tight screw taps). The process is to close both taps, open vacuum (blue) tap, turn on vacuum, in less than 5 seconds the septum will become sucked in, close vacuum tap, turn off vacuum, turn on gas from cylinder, open (blue) gas VERY slightly and the septum blows up, close gas (blue) tap, turn off gas, repeat X 2... take sample from septum. It's not ideal, but when testing the gas by piercing the septum and injecting into the GC we are getting <1% air, in comparison to ~10-30% air just straight from a tube connected to the cylinder.
 
  • #9
Big impovement for sure. 1% air would translate to a vacuum level of around 7 torr, close to the atmosphere of Mars:)

Is this a student project? What do/did you hope to learn?
 
  • #10
It's for my PhD we need to calibrate the GC for our samples, unfortunately no one knows how to use the thing so it's taking a long time. The research is based on sonochemistry. Thanks for the help!
 
  • #11
That's what science is all about, learn as you go:) Good luck! Sounds like your PhD is in secure hands.
 
  • Like
Likes rwooduk

What is the purpose of collecting a sample of gas from a cylinder?

The purpose of collecting a sample of gas from a cylinder is to analyze and study its composition and properties. This can provide valuable information for research, industrial processes, or quality control.

How is a sample of gas obtained from a cylinder?

A sample of gas can be obtained from a cylinder by using a gas syringe, a gas-tight sampling bag, or a gas collection system. The method used will depend on the specific properties of the gas and the purpose of the sampling.

What are the potential hazards of collecting a gas sample from a cylinder?

The potential hazards of collecting a gas sample from a cylinder include exposure to toxic or flammable gases, high pressure in the cylinder, and risks associated with handling and transporting compressed gas cylinders. It is important to use proper safety protocols and equipment when working with gas samples.

What factors can affect the accuracy of a gas sample collected from a cylinder?

The accuracy of a gas sample collected from a cylinder can be affected by factors such as temperature, pressure, and the presence of impurities. It is important to control these variables and follow proper procedures to ensure an accurate sample is obtained.

How can the collected gas sample be analyzed and studied?

The collected gas sample can be analyzed and studied using various techniques such as gas chromatography, mass spectrometry, and infrared spectroscopy. These methods can provide information about the composition, concentration, and other properties of the gas sample.

Similar threads

  • General Engineering
Replies
10
Views
2K
Replies
7
Views
748
  • Mechanical Engineering
Replies
3
Views
955
Replies
2
Views
4K
Replies
1
Views
588
Replies
4
Views
3K
  • Materials and Chemical Engineering
Replies
18
Views
9K
  • Engineering and Comp Sci Homework Help
Replies
1
Views
730
  • Mechanical Engineering
Replies
3
Views
1K
Replies
3
Views
955
Back
Top