- #1
Anisur Rahman
- 11
- 1
- TL;DR Summary
- Source Specification
Can I input the X-Ray tube voltage in MCNP source specification?
One would put in energy or energy spectrum for a photon source. I am not aware that voltage for an X-ray tube is an input.Anisur Rahman said:TL;DR Summary: Source Specification
Can I input the X-Ray tube voltage in MCNP source specification?
Even if the current value (mA) differs?Alex A said:An electron falling through a potential difference of 1V gains 1eV of kinetic energy.
So an X-ray tube powered at 100kV would be simulated with a beam of electrons of energy 0.1 MeV (SDEF ERG= are in MeV) directed at a target.
The current determines the number of particles (mA/e) per sec, were e is the unit charge.mariem said:Even if the current value (mA) differs?
No, MCNP does not allow direct input of the X-Ray tube voltage in the source specification. Instead, you need to define the energy spectrum of the X-rays generated by the tube, which can be derived from the tube voltage using separate software or analytical methods.
To define an X-Ray spectrum in MCNP, you need to use the SDEF (source definition) card along with the SI (source input) and SP (source probability) cards. These cards allow you to input the energy distribution and probability of each energy bin, effectively creating a custom energy spectrum for your X-ray source.
There are several tools available to generate an X-Ray spectrum from a given tube voltage, including SpekCalc, XOP, and IPEM Report 78. These tools take into account the tube voltage, anode material, and filtration to produce an energy spectrum that can be input into MCNP.
Yes, you can simulate the effect of different anode materials by adjusting the energy spectrum to reflect the characteristic X-rays and bremsstrahlung radiation produced by different anode materials. This requires generating or obtaining the appropriate energy spectrum for each anode material.
Filtration effects can be included by modifying the energy spectrum to account for the attenuation of X-rays as they pass through the filters. This can be done using attenuation coefficients for the filter materials and applying them to the initial spectrum, which can then be input into MCNP.