Measuring the temperature of a metal surface through glass with a near-infrared pyrometer can lead to inaccuracies due to the glass's absorption of infrared radiation. The transmission of glass in the 1.45 to 1.7 micron range is typically around 90%, potentially resulting in a 2% to 4% lower reading than the actual temperature. Testing the glass's impact by comparing readings with and without the glass can help quantify this effect. For improved accuracy, calibrating the measurement using a thermocouple to determine effective emissivity is recommended. Accurate temperature measurement in this scenario requires careful consideration of these factors.
long qi
I want to measure the temperature of a metal surface surrounded by glass with a near infrared pyrometer(1.45 micron to 1.7 micron). I am not sure the influence of the glass on the measuremental accuracy. You know the infrared radiation may be absorbed by the glass to make uncertainty.
Please give me some advices on this issue.
Science news on Phys.org
Welcome to PF.
Depending on the type of glass, the transmission is likely 90%, give or take a few %, over that range. That means the reading will be about 2% to 4% lower than the actual temperature, on an absolute (eg. Kelvin) temperature scale. (Based on the T4 dependence of radiation.)
Do you have a sample of the same glass, without the metal in it? You can test the glass's effect by taking a reading, then hold the test glass in front of the pyrometer and take another reading. See how much the reading drops, and add that number to the original reading.
I've done similar measurements using infrared cameras. It's not a clean measurement- what kind of accuracy are you going for?
The best solution is to 'calibrate' the sample using a thermocouple to obtain an effective emissivity.
I need to calculate the amount of water condensed from a DX cooling coil per hour given the size of the expansion coil (the total condensing surface area), the incoming air temperature, the amount of air flow from the fan, the BTU capacity of the compressor and the incoming air humidity. There are lots of condenser calculators around but they all need the air flow and incoming and outgoing humidity and then give a total volume of condensed water but I need more than that. The size of the...
Thread 'Why work is PdV and not (P+dP)dV in an isothermal process?'
Let's say we have a cylinder of volume V1 with a frictionless movable piston and some gas trapped inside with pressure P1 and temperature T1. On top of the piston lay some small pebbles that add weight and essentially create the pressure P1. Also the system is inside a reservoir of water that keeps its temperature constant at T1. The system is in equilibrium at V1, P1, T1. Now let's say i put another very small pebble on top of the piston (0,00001kg) and after some seconds the system...
I was watching a Khan Academy video on entropy called: Reconciling thermodynamic and state definitions of entropy. So in the video it says: Let's say I have a container. And in that container, I have gas particles and they're bouncing around like gas particles tend to do, creating some pressure on the container of a certain volume. And let's say I have n particles. Now, each of these particles could be in x different states. Now, if each of them can be in x different states, how many total...