A Understanding Function Generator Noise for Calibrating Noise Measurements

Dele
Messages
3
Reaction score
0
I want to calibrate my measurement setup for noise measurements. I have taken some measurements but I need to confirm that my measurement correlates with the Johnson noise theory.

To do this, I have tried measuring resistors but I seem not to get the right results. My supervisor then recommended using a function generator to do the calibration. I have measured certain noise voltages (0.1Vpp, 0.5Vpp and 1.0Vpp). However, I don't seem to understand how these relate.

My question is this. what does the noise spectral density measured from the function generator noise output translate to? A resistance? How do I compare and what might I compare with.

P.S. I intended to compare the measured spectral densities to certain resistances as calculated from the Johnson noise equation. Is this correct??

Thank you
 
Physics news on Phys.org
It will depend on the specific function generator model you are using. I assume you are using its "noise" output? Function generators generally assume a 50 ohm load
and ssuming the noise is reasonably white (i.e Johnson-like) you should (check the manual!) be able to just calculate the total noise power directly (1Vpp into 50 ohm).

Note that it should be fairly straightforward to measure the noise power from a high value resistor at room temperature and I would trust that over the output of a function generator. There isn't really much that can go wrong if you do this; especially if you are using an instrument that can measure the PSD directly such as a FFT analyzer.
 
f95toli said:
It will depend on the specific function generator model you are using. I assume you are using its "noise" output? Function generators generally assume a 50 ohm load
and ssuming the noise is reasonably white (i.e Johnson-like) you should (check the manual!) be able to just calculate the total noise power directly (1Vpp into 50 ohm).

Note that it should be fairly straightforward to measure the noise power from a high value resistor at room temperature and I would trust that over the output of a function generator. There isn't really much that can go wrong if you do this; especially if you are using an instrument that can measure the PSD directly such as a FFT analyzer.
Thanks so much. I am strongly tending towards measuring the noise power in high value resistors ATM.

While at it, might I ask (might look stupid). The spectrum analyzer I use give a whole range of the spectrum different PSDs. In fact I see something similar to the 1/f in the result. I am therefore not sure whether to trust that also. I expect to see white noise of same power over the frequency range (100Hz) in my case.

Thanks for your response
 
How high is your resistor value? Are you using a coax cable connection to the analyzer?
If you, you might be shunting the signal by the coax cable capacitance.
 
From the BCS theory of superconductivity is well known that the superfluid density smoothly decreases with increasing temperature. Annihilated superfluid carriers become normal and lose their momenta on lattice atoms. So if we induce a persistent supercurrent in a ring below Tc and after that slowly increase the temperature, we must observe a decrease in the actual supercurrent, because the density of electron pairs and total supercurrent momentum decrease. However, this supercurrent...
Hi. I have got question as in title. How can idea of instantaneous dipole moment for atoms like, for example hydrogen be consistent with idea of orbitals? At my level of knowledge London dispersion forces are derived taking into account Bohr model of atom. But we know today that this model is not correct. If it would be correct I understand that at each time electron is at some point at radius at some angle and there is dipole moment at this time from nucleus to electron at orbit. But how...