Gary_T2018 said:
That's really interesting, I am under the impression that one spine x-ray which blasts you with 2 mSv worth of X-ray dooms you for life. I'm really worried.
The 3mSv you stated for a dose for a lumbar spine might not be the dose you think it is. Typically you would want to state an average or effective dose to the irradiated field. The 3mSv is quite high for an effective dose. One these day should expect less than 1mSv as an effective dose. The exposure is determined by how much radiation is needed by the imaging device i.e., film or these days electronic digital imaging receptors. The skin dose at the entrance of the beam will be quite high and the dose will decrease about exponentially through the body. The dose will depend significantly on thickness of the body part and the kVp used depending about kVP
2.. Of course it depends linearly on the mA and time. of the exposure. Also there is a dependence on the machine age and power source.
A word of warning about estimated doses for exams, they are estimates sometimes on survey made over decades with large contributions from old technology. Even today you may have significant variations of doses from institution to institution. In the US most hospitals have a
medical physicist as consultants or on staff that can attest to the dose from various units.
The bottom line for most people is if an Xray exam is indicated for the determination of a specific diagnosis then the risk/ benefit is a no-brainer don't worry. Xray exam should not be taken on an asymptomatic person, no benefit usually. One of the links of cancer to radiation exposure was back in the 1930's or so patients where treated for various non cancerous conditions using fluoroscopy which resulted in exposure in the neighborhood of 1 Gy, and a number of cancers where produced.
gleem said:
If the DAP were cGycm2 then it would be reasonable.
Something is missing here sorry. I meant to say if the DAP were 60dSvcm2 that would be reasonable. BUT I made an error in choosing the 0.06mGy/mAs for the output by selecting the wrong data. For a real xray machines it is more like 0.20mGy/mAs at 70 kVp AND that is the dose to air at a specified distance (say 60cm) not the dose to the patient . That make the 300 dSvcm
2 very reasonable. How do you relate DAP to the effective dose to the patient? Well the DAP measures the energy in the form of radiation that is being outputted by the machine. Since almost all of it is absorbed by the patient then the DAP will be a reasonable surrogate for the effective dose to the patient.
Gary_T2018 said:
In terms us laymen could understand, the radiation does for this 1 shot/view, in mSv, would be??
There are two standard views a lateral ( xray tube from the side) and anterior-posterior xray tube above a supine patient. The doses are a bit different but around 1 mSv. again with significant variation for hospital to hospital. depending on the machine and image receptor used.