Wondering about the initial signal from ultrasonic transducer

AI Thread Summary
The discussion centers on two ultrasonic transducers: an Olympus A120S (7.5 MHz) and a customized ring-type transducer (2.X MHz). The initial signal observed when powering on the pulser-receiver may be an artifact of the circuit rather than a meaningful output. The customized transducer produces a noisier and less clear signal compared to the A120S, potentially due to its geometry and lower frequency, which affects sensitivity and signal quality. Users are encouraged to compare their measurements with Olympus's waveform records and consider the compatibility of the pulser-receiver with the customized transducer. Understanding the transducer's sound field characteristics and ensuring proper equipment compatibility are crucial for improving signal clarity.
Tah
Messages
12
Reaction score
0
Hi,

I have two ultrasonic transducers and a pulser-receiver(5072PR).

One is the contact transducer, A120S (6MHz), from Olympus and the other is also contact mode transducer, but it is a customized ring-type transducer (2.X MHz) (like doughnut).

I have two questions.

First, when I turn on the pulser-receiver, the initial line signal occurs from ultrasonic transducer.
I was wondring what is this signal.

Second, the two different transducers have different signal shape as shown below.
The signal from the customized transducer look so shoddy and dirty.
Why is that? and how to improve this kind of noisy signal?Please let me know. I'm so curious about all of that.

http://postfiles16.naver.net/20130416_207/jameskimes_1366102149606wPCbI_JPEG/x.png?type=w2
 
Last edited by a moderator:
Engineering news on Phys.org
The signal you first see might just be an artefact of the circuit powering up. More details?

Can't see the picture you posted. In what way is the transducer customised?
 
Tah said:
Hi,

I have two ultrasonic transducers and a pulser-receiver(5072PR).

One is the contact transducer, A120S (6MHz), from Olympus and the other is also contact mode transducer, but it is a customized ring-type transducer (2.X MHz) (like doughnut).

I have two questions.

First, when I turn on the pulser-receiver, the initial line signal occurs from ultrasonic transducer.
I was wondring what is this signal.

Second, the two different transducers have different signal shape as shown below.
The signal from the customized transducer look so shoddy and dirty.
Why is that? and how to improve this kind of noisy signal?


Please let me know. I'm so curious about all of that.

http://postfiles16.naver.net/20130416_207/jameskimes_1366102149606wPCbI_JPEG/x.png?type=w2

1. According to the Olympus Transducer website, contact transducer part number A120S-RM operates at 7.5 MHz, and not at 6 MHz. Can you explain this discrepancy?

2. Olympus provides typical "records of the waveform and spectrum of each transducer is maintained and can be accessed for comparative or statistical studies of transducer characteristics." Have you compared these signal shapes to your measurements?

3. The driving waveforms applied to transducers may look "shoddy and dirty" to you, but what matters is the system performance. Are you receiving echoes that contain the desired signal quality?

Bobbywhy
 
Last edited by a moderator:
Bobbywhy said:
1. According to the Olympus Transducer website, contact transducer part number A120S-RM operates at 7.5 MHz, and not at 6 MHz. Can you explain this discrepancy?

2. Olympus provides typical "records of the waveform and spectrum of each transducer is maintained and can be accessed for comparative or statistical studies of transducer characteristics." Have you compared these signal shapes to your measurements?

3. The driving waveforms applied to transducers may look "shoddy and dirty" to you, but what matters is the system performance. Are you receiving echoes that contain the desired signal quality?

Bobbywhy


1.
Yes, you're right.
7.5MHz is correct for A120S, not 6Mhz.


2. Olympus provides typical "records of the waveform and spectrum of each transducer is maintained and can be accessed for comparative or statistical studies of transducer characteristics." Have you compared these signal shapes to your measurements?

I've already done lots of experiments with those two transducers to measure the thickness of an object and distinguish two or three-layered materials which have different material density in each layer.
The signal from A120S was so clear to see the results but the customized ring-type transducer was so poor from sensitiviy and lots of noise.


3. The driving waveforms applied to transducers may look "shoddy and dirty" to you, but what matters is the system performance. Are you receiving echoes that contain the desired signal quality?

I was of course receiving echoes. And as you can see the image, the A120S showed clean signal and good performance, however the signal from the customized transducer showed poor signal. It significantly overlaps the time region (x-axis in the oscilloscope) that is important for measuring the thickness of thin materials.
I'm curious that is this due to the geometry of ring-shaped transducer(single element) or lower central freuqnecy(2.X MHz)?
 
As for your first question: "First, when I turn on the pulser-receiver, the initial line signal occurs from ultrasonic transducer. I was wondring what is this signal."

Will you please be more explicit?:
1. What is an "initial line signal?
2. Which transducer is connected to the pulser-receiver?
3. Exactly where is your O'scope probe connected?
4. Are you describing the "transmitting" waveform, or the "received" echo waveform?

As for your “ring-type” customized transducer:
1. What is the manufacturer and part number?
2. Are you using the same pulser-receiver for both different transducers? If the ring-shaped transducer is resonant at 2.X MHz can you expect to drive it with a pulser for a 7.5 MHz transducer?
3. What is the sound field of this “doughnut-shaped” transducer? Specifically, where are the near-field and far-field compared to your “distance of interest”?
4. Do you know the beam width and focal zone?
5. These, plus a few other characteristics of the torriodal transducer’s sound field may explain why you see that waveform. Furthermore, it’s possible that the pulser is not compatible with that customized transducer, resulting in the observed waveform. Is the piezo-electric material the same as your A120S? Beam patterns from the simple A120S and the torriodal will be significantly different, resulting in the long-time decaying echo signal.

By the way, here is the Olympic technical manual:
http://www.olympus-ims.com/data/File/panametrics/panametrics-UT.en.pdf

Bobbywhy
 
Tah said:
Second, the two different transducers have different signal shape as shown below.
The signal from the customized transducer look so shoddy and dirty.
Why is that? and how to improve this kind of noisy signal?
Your photos initially didn't reveal themselves. So I extracted the URL and viewed them independently. When I came back to see whether I could post so they did show, I found them displayed properly.

Not sure whether to attribute that oddity to PF or naver.net

http://postfiles16.naver.net/20130416_207/jameskimes_1366102149606wPCbI_JPEG/x.png?type=w2
 
Last edited by a moderator:
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...