How can I design high-quality EEG electrodes?

AI Thread Summary
Advanced EEG electrode design requires understanding both the basic principles and specific techniques for high-quality production. Key considerations include electrode size, material selection, and the importance of using conductive gels for optimal signal capture. Safety requirements are critical, necessitating protective current limiters and isolation amplifiers to prevent interference and ensure patient safety. Modern designs often utilize disposable electrodes and advanced wiring techniques to minimize noise and maintain signal integrity. Resources from companies like Analog Devices provide valuable insights into the latest technologies and best practices in EEG electrode design.
pierce15
Messages
313
Reaction score
2
I am looking for a source (or several) on advanced EEG design techniques, in particular how the highest quality electrodes are made. I understand the basic principles of how an electrode works but have no knowledge at all about design techniques. Does anyone know of something that would help me?

Also I was not sure if this should go in the learning materials section, sorry in advance if this is the wrong section.
 
Engineering news on Phys.org
pierce15 said:
I am looking for a source (or several) on advanced EEG design techniques, in particular how the highest quality electrodes are made. I understand the basic principles of how an electrode works but have no knowledge at all about design techniques. Does anyone know of something that would help me?

Also I was not sure if this should go in the learning materials section, sorry in advance if this is the wrong section.
I did a google search on Biomedical Instrumentation Design, and got lots of good hits. Maybe try something similar?

Also, keep in mind that biomedical instruments (especially those that attach to the body via electrodes) must meet strict safety requirements in their design. As you read about the designs, look for the safety requirements and how they are being met in the design.
 
The first thing to do is to clean the skin and apply a conductive gel with the electrode. The size of the electrode will be important. Too big and it will pick up noise and influence other nearby electrodes, too small and the series resistance will be higher.
These days the electrodes are disposable.

If AC coupling is used there should be little electrolytic corrosion at the electrode. The problem with AC coupling will be “de-blocking”, that is quickly getting the right voltage across the capacitor to bring the amplifier voltage into range, while still permitting low-frequency signals to pass during observation.

Back when I was in the EEG field, protective current limiters were used on all electrodes. They were once made from a resistor with an N-channel and a P-channel depletion mode FET; as a bidirectional constant current circuit. The FETs were rated to the peak voltage of the local supply.
Analog Devices have been building isolation amplifiers since then. As an example take a look at;
http://www.analog.com/en/products/landing-pages/001/7b/7b31-isolated-voltage-input.html#Introduction

Now we might isolate the entire analogue front end, powered through an isolated DC-DC converter. All the amplifiers and A-D converters would be floating and communicating back to the display through opto-isolators or an optic fibre.

You will need to select wires to the electrodes that are not microphonic.
Where the electrode wire was screened coax, the screen was not grounded but driven by a low-impedance voltage-follower from the signal so as to eliminate the cable capacitance and restore the frequency response.

The instrumentation amplifiers in the front end were once referenced to a floating average of all signals. Now we might consider looking at the difference signals wanted and convert them independently of the average. You will need to find out what electrode positions are being used these days, and what signals must be extracted from them.

Analog Devices have expertise in EEG system component design. Go to http://www.analog.com and search 'EEG'. You will find good information like;
http://www.analog.com/media/en/technical-documentation/white-papers/ECG-EEG-EMG_FINAL.pdf
http://www.analog.com/en/applications/markets/healthcare/clinical-monitoring/eeg-measurement.html
 
  • Like
Likes berkeman
Thread 'I need a concave mirror with a focal length length of 150 feet'
I need to cut down a 3 year old dead tree from top down so tree causes no damage with small pieces falling. I need a mirror with a focal length of 150 ft. 12" diameter to 36" diameter will work good but I can't think of any easy way to build it. Nothing like this for sale on Ebay. I have a 30" Fresnel lens that I use to burn stumps it works great. Tree service wants $2000.
Hi all, i have some questions about the tesla turbine: is a tesla turbine more efficient than a steam engine or a stirling engine ? about the discs of the tesla turbine warping because of the high speed rotations; does running the engine on a lower speed solve that or will the discs warp anyway after time ? what is the difference in efficiency between the tesla turbine running at high speed and running it at a lower speed ( as fast as possible but low enough to not warp de discs) and: i...