Grounded vs Ungrounded Thermocouples for Tube-Skin Temperature Measurement

AI Thread Summary
Using grounded thermocouples in a flame tube design could lead to ground loop issues due to electrical connections between the thermocouples along the same tube. This could result in interference and inaccurate temperature readings. An instrumentation amplifier may mitigate some of these problems, but grounding one leg of the thermocouple at the readout could create additional thermocouples, complicating the measurements. The consensus suggests that ungrounded thermocouples may be a better choice to avoid these complications. Therefore, avoiding a common ground is crucial for accurate temperature profiling in this setup.
gball
Messages
2
Reaction score
0
Hello All,

We are experimenting with a flame tube design (quite literally, a flame within a tube). We are looking to get a temperature profile along this tube. So, we will need multiple thermocouples along this tube. I'm wondering whether it will be necessary to go to ungrounded (isolated) thermocouples for this since grounded thermocouples (junction connected to sheath) would all be electrically connected to each other since they are all on the same tube. Would this present a 'ground loop' problem?

Would this be a problem? Would there be interference between the thermocouples? I would really appreciate any input.

Thanks in advance,

gball
 
Engineering news on Phys.org
I think they would interfere. But it depends on your read out method. If you use an instrumentation amplifier you might be fine. But if you tie one leg of the thermocouple to ground at the read out circuit you will produce another thermocouple: "grounded leg of thermocouple 1 - contact at temperature 1 with pipe - pipe - contact at temperature 2 with pipe - grounded leg of thermocouple 2". This will shift your voltage I think.
 
0xDEADBEEF said:
I think they would interfere. But it depends on your read out method. If you use an instrumentation amplifier you might be fine. But if you tie one leg of the thermocouple to ground at the read out circuit you will produce another thermocouple: "grounded leg of thermocouple 1 - contact at temperature 1 with pipe - pipe - contact at temperature 2 with pipe - grounded leg of thermocouple 2". This will shift your voltage I think.
Many thanks! I can see what you mean now. No common ground!
 
Thread 'Physics of Stretch: What pressure does a band apply on a cylinder?'
Scenario 1 (figure 1) A continuous loop of elastic material is stretched around two metal bars. The top bar is attached to a load cell that reads force. The lower bar can be moved downwards to stretch the elastic material. The lower bar is moved downwards until the two bars are 1190mm apart, stretching the elastic material. The bars are 5mm thick, so the total internal loop length is 1200mm (1190mm + 5mm + 5mm). At this level of stretch, the load cell reads 45N tensile force. Key numbers...
I'm trying to decide what size and type of galvanized steel I need for 2 cantilever extensions. The cantilever is 5 ft. The space between the two cantilever arms is a 17 ft Gap the center 7 ft of the 17 ft Gap we'll need to Bear approximately 17,000 lb spread evenly from the front of the cantilever to the back of the cantilever over 5 ft. I will put support beams across these cantilever arms to support the load evenly