How does a calibrated Hall Probe work?

  • Thread starter Thread starter Yoriko
  • Start date Start date
  • Tags Tags
    Probe Work
AI Thread Summary
A calibrated Hall probe measures magnetic field strength (B) and requires connection to a meter for direct readings. The probe typically operates with a regulated DC voltage of 5V or 6V and outputs a differential voltage proportional to the magnetic field strength, usually between 0.2 mV and 2 mV per Gauss. Probes can vary in design, with some including built-in amplifiers and temperature compensation. Calibration accuracy is generally around ±20% unless specifically hand-calibrated. For precise measurements, users can create a small electromagnet for self-calibration if needed.
Yoriko
Messages
9
Reaction score
0
I'm supposed to plan an experiment using a calibrated hall probe to investigate the relationship between Magnetic field strength (B) of a coil and the distance (x) from the centre of the coil.

I'm supposed to be able to take direct readings from the calibrated hall probe to calculate B, but how to? Is the hall probe supposed to be connected to something else?
 
Last edited:
Physics news on Phys.org
There are several implementations of the linear Hall Effect probes used for magnetic field measurement. The simplest architecture is similar to a Wheatstone bridge, with two terminals for applied voltage, and two for a differential voltage output (floating outputs). The applied dc input voltage should be a regulated voltage, usually 5V or 6V. Use a 7805 or 7806 IC regulator if necessary.The differential output voltage is usually about 0.2 mV per Gauss to 2 mV per Gauss, with an output impedance of several k-ohms. The output voltage usually has a temperature coefficient. These are 4-terminal probes. Use a floating digital voltmeter (battery operated is best) with 1 mV or better resolution, and a 10 meg input impedance.

Some linear probes have a built-in differential amplifier (with some voltage gain) with temperature compensation. These are usually 3-terminal probes, with the output voltage measured with respect to the negative input voltage terminal. For a 6V input probe, the output voltage for zero Gauss is 3V.

Unlike some other probes, like proton nuclear magnetic resonance (NMR), the Hall Effect probes are usually only about ±20% calibration, unless specifically hand-calibrated.

Bob S
 
Bob S said:
There are several implementations of the linear Hall Effect probes used for magnetic field measurement. The simplest architecture is similar to a Wheatstone bridge, with two terminals for applied voltage, and two for a differential voltage output (floating outputs). The applied dc input voltage should be a regulated voltage, usually 5V or 6V. Use a 7805 or 7806 IC regulator if necessary.The differential output voltage is usually about 0.2 mV per Gauss to 2 mV per Gauss, with an output impedance of several k-ohms. The output voltage usually has a temperature coefficient. These are 4-terminal probes. Use a floating digital voltmeter (battery operated is best) with 1 mV or better resolution, and a 10 meg input impedance.

Some linear probes have a built-in differential amplifier (with some voltage gain) with temperature compensation. These are usually 3-terminal probes, with the output voltage measured with respect to the negative input voltage terminal. For a 6V input probe, the output voltage for zero Gauss is 3V.

Unlike some other probes, like proton nuclear magnetic resonance (NMR), the Hall Effect probes are usually only about ±20% calibration, unless specifically hand-calibrated.

Bob S

:confused: Hm I'm not taking so high level physics yet so I don't really understand. So basically the hall probe is connected to a Gauss meter so direct readings of magnetic field strength can be obtained?
 
Yoriko said:
I'm supposed to plan an experiment using a calibrated hall probe to investigate the relationship between Magnetic field strength (B) of a coil and the distance (x) from the centre of the coil.

I'm supposed to be able to take direct readings from the calibrated hall probe to calculate B, but how to? Is the hall probe supposed to be connected to something else?

It needs to be connected to a meter. Most probes are calibrated to match a given meter.

for instance:

http://www.lakeshore.com/mag/ga/gm410po.html
 
Last edited by a moderator:
MagnetDave said:
It needs to be connected to a meter. Most probes are calibrated to match a given meter.

for instance:

http://www.lakeshore.com/mag/ga/gm410po.html

Ok thanks! :smile:
 
Last edited by a moderator:
Yoriko said:
:confused: Hm I'm not taking so high level physics yet so I don't really understand. So basically the hall probe is connected to a Gauss meter so direct readings of magnetic field strength can be obtained?
You could get one of these sensors from AllegroMicro (sometimes available as a free sample) and read it out with a standard digital voltmeter:

http://www.allegromicro.com/en/Products/Part_Numbers/1301/1301.pdf

If you need an absolute calibration better than than the standard specification, make a small electromagnet and do it yourself.

Bob S
 
Hello everyone, Consider the problem in which a car is told to travel at 30 km/h for L kilometers and then at 60 km/h for another L kilometers. Next, you are asked to determine the average speed. My question is: although we know that the average speed in this case is the harmonic mean of the two speeds, is it also possible to state that the average speed over this 2L-kilometer stretch can be obtained as a weighted average of the two speeds? Best regards, DaTario
The rope is tied into the person (the load of 200 pounds) and the rope goes up from the person to a fixed pulley and back down to his hands. He hauls the rope to suspend himself in the air. What is the mechanical advantage of the system? The person will indeed only have to lift half of his body weight (roughly 100 pounds) because he now lessened the load by that same amount. This APPEARS to be a 2:1 because he can hold himself with half the force, but my question is: is that mechanical...
Some physics textbook writer told me that Newton's first law applies only on bodies that feel no interactions at all. He said that if a body is on rest or moves in constant velocity, there is no external force acting on it. But I have heard another form of the law that says the net force acting on a body must be zero. This means there is interactions involved after all. So which one is correct?