what is the depth of field of a microscope and how is it different from the depth of focus of the microscope.
Last edited by a moderator:
Andy Resnick said:Technically, 'Depth of field' refers to object space while 'depth of focus' refers to image space. In practice, both terms are often used interchangeably to refer to object space- how much of the object is in focus.
http://www.microscopyu.com/articles/formulas/formulasfielddepth.html
PAllen said:And the answer is, for microscopes I can afford, at interesting magnification, "not enough".
Andy Resnick said:Joking aside (because I feel your pain, what do you mean? I ask only because I may be able to offer some work-arounds.
Andy Resnick said:What are the numerical apertures of the objective lenses?
PAllen said:Well, I'm not sure. Here are the numbers on the 40x objective:
40/0.65
160/0.17
I assume 40 is obviously 40x (it is printed bigger than all the other numbers). This objective 100x immersion one have the front component on a spring mounting to protect from damage hitting a slide.
PAllen said:Thanks, very informative. As to stopping down the condenser, what control do I use? I have two available: 1) the Iris diaphragm 2) a knob that raises and lowers the condenser. Are either of these relevant, or is my microscope missing the control I need?
PAllen said:My condenser has an NA of 1.25, as does the 100x objective. All the objectives are labeled 160/.17.
The oil I use is labeled 1.515 rather than 1.518. It is made by Richard Allan Scientific (the oil, that is).
The microscope originally came with cedar oil, but I was advised not to use it, instead get a modern oil like the one I bought.
I note that my cover slips are labeled .13-.17 mm. Should I get better ones with smaller thickness tolerance?
There definitely seems no way to control the aperture of the condenser. Lowering it only spreads the light out; the iris comes before condenser in the light path. I would guess an aperture stop would come in front of the condenser or part of the condenser??
It is clear I don't have control over both.Andy Resnick said:Glad to hear it. First, I erred earlier- 0.17 is a #1 1/2 (1.5) coverslip, not a #1.
Stopping down the condenser- first, make sure the condenser is aligned (centered and focused; Kohler illumination)- the goal is to focus the condenser and the objective onto the same sample plane- that's what raising and lowering the condenser does. The iris diaphragm could be either the field or the aperture diaphragm, I can't say which without more information. If the condenser is focused on the sample and the iris is focused onto the sample (closing the iris results in a circle of illuminated area stopping down), then it's the field diaphragm. The aperture diaphragm, when the condenser is properly aligned, will apparently control the brightness of the illumination. It's possible that diaphragm is not located on the condenser itself, but in the body of the scope. OTOH, you may not have control over both.
I use it a couple of times a year - shouldn't be a problem. (I use the microscope more, but rarely bother with immersion). The whole thing is just one of several secondary hobbies.Andy Resnick said:You're correct on not using cedar oil- it dries out and turns into glue. The 'modern' oil is what you want, but be aware that there are nasty chemicals in it, and that contact dermatitis is a common side-effect.
.Andy Resnick said:The coverslip tolerance you have seems a little sloppy- I just get ones specified as #1.5. Some folks get really anal retentive and measure each coverslip with a micrometer on their own and only use ones with the exactly correct thickness. I think that's a bit much, but whatever...
Andy Resnick said:If you're condenser is specified as NA 1.25, are you using immersion oil on it? That is definitely an immersion condenser. Be careful- I try not to use immersion condensers because it means I have much less room for error on the objective side- much easier to break the sample.
What's the model scope?
The depth of field of a microscope refers to the thickness of the specimen that appears in focus at one time. It is the distance between the nearest and farthest points of the specimen that can be seen in sharp focus under the microscope.
The depth of field of a microscope is determined by the numerical aperture of the objective lens, the wavelength of light used, and the refractive index of the medium between the objective lens and the specimen. It can also be affected by the magnification and thickness of the specimen.
The depth of field is important in microscopy because it allows us to see a three-dimensional image of the specimen. It also helps in determining the thickness and structure of the specimen, which is crucial in many scientific fields such as biology and materials science.
Yes, the depth of field can be adjusted on a microscope by changing the aperture of the objective lens or by adjusting the focus of the microscope. Increasing the aperture will decrease the depth of field, while decreasing the aperture will increase the depth of field.
The depth of field can vary in different types of microscopes due to differences in their optical systems. For example, compound microscopes typically have a larger depth of field compared to stereomicroscopes, which have a smaller depth of field but provide a more three-dimensional image. Electron microscopes, on the other hand, have a significantly smaller depth of field compared to light microscopes due to their higher magnification and shorter wavelength of electrons.