The discussion revolves around displaying slow-changing signals on a RIGOL oscilloscope, specifically when the signal is around 5V ±100mV. The user faces challenges with AC-mode not capturing slow changes and DC-mode being limited to a 2V offset when using <500mV/DIV settings. Suggestions include using a low-pass filter to create a reference signal or employing a power supply as a reference voltage. Additionally, a difference amplifier setup is recommended to help manage the offset while maintaining signal integrity. The conversation emphasizes practical solutions for effectively visualizing and processing the signal data.
Markus Hoeg
Messages
5
Reaction score
0
Hello,

I have a very specific problem and didn't find any way to solve it.
My signal is very slow changing at 5V +-100mV
When I want to display this signal on my RIGOL Oscilloscope, the problem is, that with AC-mode I don't see those slow changes and in DC-mode the offset range is limited to 2V when displaying <500mV/DIV.
My changes are around 100mV so my display settings need to be 100mv/DIV.
Is there any way to display this kind of signals?

Thanks in advance
Markus
 
Engineering news on Phys.org
Welcome to the PF, Markus.

Do you have a non-changing 5V reference signal? If so, use two 'scope probes and do a math subtraction (like A = Ch1-Ch2) to be able to get rid of the offset...
 
  • Like
Likes Averagesupernova
No I don't have any reference signal. The signal is created by a photodetector after light passes through a fibre. So the voltage depends on the light passing through the fibre, which is slow-changing.
 
Then you could make one, with a very-lowpass-filter. Or use a power supply that you set to the average voltage. Are you digitizing the signal and doing any post-processing?
 
Okay I will give that one a try, Thanks!
Yes I need to send the data via USB to MATLAB afterwards.
 
You can probably use the power supply of the photodetector as a reference voltage, perhaps through a voltage divider if needed.
 
The power supply of the photodetector is constant, while the output DC component is changing for every measurement depending on the light through the fiber. That would mean that for every measurement I have to adjust the voltage divider.
 
Insert a difference amplifier after the photodetector:
03044.png

Connect V2 to the +5V supply and V1 to the photodetector output. The value of R depends on the choice of Op-Amp, but I would recommend something around 0.1 -1MΩ.
 

Attachments

  • 03044.png
    03044.png
    1.8 KB · Views: 439
Thank you for your useful instructions :smile:
 
  • #10
How about this? Depending in the DC level from the detector, you may want the Zener to be 3.3V. The 2.2k resistor value is just a starting suggestion, the actual value depends on the maximum available output current of the detector and the leakage current of the Zener.

upload_2018-11-27_18-52-9.png


Cheers,
Tom
 

Attachments

  • upload_2018-11-27_18-52-9.png
    upload_2018-11-27_18-52-9.png
    1.9 KB · Views: 445
  • Like
Likes Markus Hoeg