Optical microscopes and white light / laser light

AI Thread Summary
The discussion centers on the use of white light versus laser light in optical microscopy. A diffuser is used to provide uniform illumination, while the field diaphragm controls the illuminated field of view, with adjustments enhancing optical performance. White light is preferred for widefield imaging due to its incoherence, which prevents issues like speckle that arise from using raw laser light. Although laser light offers high spatial and temporal coherence, it can lead to poor image quality in widefield applications without proper spatial filtering. Overall, white light is more effective for traditional optical microscopy, particularly when the human eye is the detector.
Steve Drake
Messages
53
Reaction score
1
I have a few questions regarding an optical microscope and their white light sources...

So white light generally first hits a diffuser, some kind of ground glass lens. What is the purpose of this?

Then the light goes through a field diaphragm, which we can open and close. I have heard that closing of the field diaphragm can slightly increase the spatial coherence of the light, why is this?

Lastly, why is white light used and not laser light? Wouldn't the almost perfect spatial and temporal coherence of laser light mean we get much better images? With that said, how does white light form an image as good as it does being so spatially and temporally incoherent?

Sorry for the barrage of questions, just been trying to wrap my head around this stuff.

Thanks
 
Science news on Phys.org
The diffuser provides uniform illumination.

Laser light is monochromatic. White light will reveal all the colors of the object under the microscope.

Some specialized imaging applications do use laser light.

But white light is better when the detector is the human eye.
 
Steve Drake said:
I have a few questions regarding an optical microscope and their white light sources...
So white light generally first hits a diffuser, some kind of ground glass lens. What is the purpose of this?

The diffuser plate is there in case the user doesn't know how to establish Kohler illumination. I have all my students read and walk through the process as written out here:

http://www.well.ox.ac.uk/_asset/file/zeiss-guide-to-microscopy-from-the-very-begiining.pdf

We remove the diffusers, as they waste light.

Steve Drake said:
Then the light goes through a field diaphragm, which we can open and close. I have heard that closing of the field diaphragm can slightly increase the spatial coherence of the light, why is this?

There's some confusion here: the field diaphragm controls the field of view that is illuminated by the source. The aperture diaphragm controls the numerical aperture of the condenser illumination. Stopping down the aperture diaphragm will increase the contrast and depth of field and decrease the resolution. To be sure, slightly stopping down both diaphragms will slightly improve the optical performance (less glare, for example) because you aren't pushing the limits of the design.

Steve Drake said:
Lastly, why is white light used and not laser light? Wouldn't the almost perfect spatial and temporal coherence of laser light mean we get much better images? With that said, how does white light form an image as good as it does being so spatially and temporally incoherent?

Widefield imaging works best with incoherent light. Kohler illumination is perfectly incoherent, phase contrast and DIC imaging start with incoherent light and increase the coherence in a controlled way to achieve optical processing. Using 'raw' laser light for wide-field imaging results in terrible images- there is an intolerable amount of speckle due to the low spatial coherence of the beam. Using a spatial filter will increase the spatial coherence and give better images than the raw light, but that's a pointless exercise. Again, if you are not doing widefield imaging (confocal, near field, etc.) then laser light is better as it can be better focused and coincidentally can be used with fluorescence techniques.
 
Hi Courtney, Andy, thanks for your replies. Some questions,

Andy Resnick said:
Widefield imaging
I am not 100% familiar with this term, is optical microscopy widefield imaging?

Andy Resnick said:
Using 'raw' laser light for wide-field imaging results in terrible images- there is an intolerable amount of speckle due to the low spatial coherence of the beam.
By laser light I meant very high temporally and spatially coherence lasers.
Andy Resnick said:
there is an intolerable amount of speckle due to the low spatial coherence of the beam
I thought speckle only comes from a coherent source interacting with a random array of scatterers, can you clarify what you mean please?

Essentially what I am trying to understand is if you replaced a traditional white light source on an optical microscope with a laser, and sent it through a collimator, what kind of image would you obtain vs white light?

Thanks
 
Steve Drake said:
I am not 100% familiar with this term, is optical microscopy widefield imaging?

'Widefield' means that the field of view is larger than a point, it's used to distinguish 'normal' microscopy from say, scanning methods.

Steve Drake said:
By laser light I meant very high temporally and spatially coherence lasers.
A note of caution- the 'raw' laser beam does indeed have high temporal coherence, but typically only moderate spatial coherence- it's set by the size of the exit window. To increase the spatial coherence of the laser light, you spatially filter the beam- essentially creating a new source that is much smaller in extent, by using a pinhole aperture.

Steve Drake said:
I thought speckle only comes from a coherent source interacting with a random array of scatterers, can you clarify what you mean please?

That's close enough- the point is that the size of the speckle is related to the coherence area. Clearly, magnifying the object will also magnify the speckle.

Steve Drake said:
Essentially what I am trying to understand is if you replaced a traditional white light source on an optical microscope with a laser, and sent it through a collimator, what kind of image would you obtain vs white light?

Again, it depends of the details of the sources and how they are aligned within the optics. Goodman's "fourier optics" book has a fantastic image comparing the two approaches, but I couldn't find it on the interwebnet. Here's an alternative:

http://www.nature.com/nphoton/journal/v6/n6/images/nphoton.2012.90-f3.jpg
 
Thread 'A quartet of epi-illumination methods'
Well, it took almost 20 years (!!!), but I finally obtained a set of epi-phase microscope objectives (Zeiss). The principles of epi-phase contrast is nearly identical to transillumination phase contrast, but the phase ring is a 1/8 wave retarder rather than a 1/4 wave retarder (because with epi-illumination, the light passes through the ring twice). This method was popular only for a very short period of time before epi-DIC (differential interference contrast) became widely available. So...
Hello ! As we know by definition that: "Constructive interference occurs when the phase difference between the waves is an even multiple of π (180°), whereas destructive interference occurs when the difference is an odd multiple of π." But my question is in the case of destructive interference, what happens to the energy carried by the two electromagnetic waves that annihilate, the energy carried by the electromagnetic waves also disappears, or is transformed into some other type of...
I am currently undertaking a research internship where I am modelling the heating of silicon wafers with a 515 nm femtosecond laser. In order to increase the absorption of the laser into the oxide layer on top of the wafer it was suggested we use gold nanoparticles. I was tasked with modelling the optical properties of a 5nm gold nanoparticle, in particular the absorption cross section, using COMSOL Multiphysics. My model seems to be getting correct values for the absorption coefficient and...

Similar threads

Back
Top