Medical OSL Dosimetry and Build Up Region

AI Thread Summary
The discussion centers on the relationship between photon energy, dose measurement, and the use of bolus material in dosimetry. It explains that the build-up region, where kerma exceeds dose, is crucial for understanding how 6 MV photons interact with tissue, particularly at the depth of Dmax (approximately 1.5 cm). The necessity of using bolus material, such as water, is clarified; it is not essential for measurements but is employed to accurately assess the dose at Dmax rather than at the surface. This approach helps achieve charged particle equilibrium, which is vital for reliable dose verification against treatment planning systems. The discussion also highlights the challenges of measuring surface dose due to steep dose gradients and uncertainties in treatment planning predictions, as well as the potential perturbation effects of the OSL detector itself. Overall, while OSL measurements can be taken without bolus, using it enhances accuracy in specific contexts.
AndresPB
Messages
28
Reaction score
1
I understand the following image as follows:

B9781416053163002513_f248-003-9781416053163.jpg


There is a build up region that varies with photons energy, the Dmax for 6 MV photons is around 1.5 cm. The build up region means that kerma > dose in that region, meaning that photons are creating electrons and when the Dmax is reached it is going to be the point where the photons make the most electrons and then after this point the dose will begin to decrease because the intensity of the beam (photons) will diminish because of body attenuation.

I am having troubles relating this explanation with the way Optical Stimulated Light Dosimetry and TLD's work. Isn't it suppose that this kind of dosimetry absorb photons to work (Photons excite electrons in the conduction band of the crystal and they move to the valence band)? Then why is it necessary to use some kind of material for the "build up" region to measure dose in surface of a patient?
With this I mean: In many cases, medical physicists use a bolus equivalent of 1.5 cm of water to simulate this "build-up" region and they put the detector underneath it. Why is this necessary if what we are measuring are photons and not electrons?

Thanks a lot!
 

Attachments

  • B9781416053163002513_f248-003-9781416053163.jpg
    B9781416053163002513_f248-003-9781416053163.jpg
    10.5 KB · Views: 1,611
Biology news on Phys.org
Interesting question.

Adding bolus is not "necessary" as a matter of need for the measurements. It's done when one is interested in measuring the Dmax dose, as opposed to the surface dose.

If you're trying to say, make a measurement on a patient to verify a calculation made by your treatment planning system, it's much easier to verify dose to a point where charged particle equilibrium exists. Most modern treatment planning systems will calculate dose at Dmax and downstream very accurately. But look at your graph over the first few mm from the surface. The slopes are rather steep. The steep slopes, plus, uncertainties in your source model (electron contamination, low energy spectra, etc.), plus partial volume effects on your planning CT, can lead to high uncertainties in treatment planning system predictions of surface dose. And you also have to account of the physical properties of your OSL detector and its casing perturbing the beam. This makes it difficult to expect a dosimetry match at the surface.

You can still make an OSL detector measurement without any bolus or buildup material. And there are many reasons why someone might want to do this.

Does that make sense?
 
Last edited:
  • Like
Likes jim mcnamara
https://www.discovermagazine.com/the-deadliest-spider-in-the-world-ends-lives-in-hours-but-its-venom-may-inspire-medical-miracles-48107 https://en.wikipedia.org/wiki/Versutoxin#Mechanism_behind_Neurotoxic_Properties https://www.sciencedirect.com/science/article/abs/pii/S0028390817301557 (subscription or purchase requred) he structure of versutoxin (δ-atracotoxin-Hv1) provides insights into the binding of site 3 neurotoxins to the voltage-gated sodium channel...
Popular article referring to the BA.2 variant: Popular article: (many words, little data) https://www.cnn.com/2022/02/17/health/ba-2-covid-severity/index.html Preprint article referring to the BA.2 variant: Preprint article: (At 52 pages, too many words!) https://www.biorxiv.org/content/10.1101/2022.02.14.480335v1.full.pdf [edited 1hr. after posting: Added preprint Abstract] Cheers, Tom