Quantum Efficiency Mapping setup for solar cell charatcerization

AI Thread Summary
Collimated light is essential for accurate quantum efficiency (QE) characterization of solar cells, as the reflectivity of silicon varies with the angle of incidence. A focused beam is preferred over a Gaussian beam due to its uniform intensity, which is crucial for large area cells. While laser light is typically used for QE mapping, monochromatic light can be employed, but concerns include light intensity and potential wavelength drift during long mapping durations. The series resistance of the grid may also impact photo-current collection, though this effect can be negligible depending on the setup. The discussion emphasizes the need for careful calibration and measurement accuracy in the QE mapping process.
2lai
Messages
3
Reaction score
0
In order to build a QE measurement system, I want to confirm the following issues: (1) Is collimated light required for QE characterization of the entire solar cell? How "bad" it could be if having Gaussian beam? (2) It seems LBIC is most common for QE mapping. Is it possible to use monochromatic light instead of laser for QE mapping? I guess the restrictions may include (a) light intensity; (b) wavelength drifting of monochromator in the long mapping duration (Yes or No?) (c) the series resistance of the grid may affect photo-current since photo-generated carriers at different location will travel different distance prior to collection (is this effect negligible?)
 
Engineering news on Phys.org
Welcome to PF, I'll answer what I can.

(1)
Yes, you'd want a reasonably collimated beam, since the reflectivity of silicon will vary with angle of incidence. Collimated to within a few milliradians should definitely work, and probably within 5 degrees would work too.

Not sure why you think a Gaussian beam would be bad ... as long as it's reasonably collimated, it shouldn't matter if it's Gaussian or not.

(2)
Not sure what you mean, since a laser beam is monochromatic. Or do you mean you want to vary the wavelength? I'm not sure how much power you can get by using a monochromator as a filter on a white-light source.
 
Do you want to do this with sunlight, or with monochromatic light? The quantum efficiency varies with frequency, and is probably very poor in both the infra red and ultra violet regions. For "clean systems" like photomultipliers, QE measurement is relatively easy. For solar cells, measuring the output current is the only way to get QE, and dark current may interfere. What method do you plan to use for calibrating your light source, and how accurate a measurement do you need? How big a solar cell do you plan to use it on, and how fast do do want to do it (i.e., production line application)?
 
We plan to order a Class-B light source. The cell size is around 3x3 cm^2. The measurement system will be setup in 1-2 months.
 
Thanks Redbelly98.
(1) I should say focused beam rather than Gaussian beam. The intensity is not uniform (not collimated either) which may affect the measurement especially for large area cell.
(2) Adding a laser beam in the measurement system also add the complexity and $$$. It is also my concern the output power of monochromatic light.
 
Hi all, I have a question. So from the derivation of the Isentropic process relationship PV^gamma = constant, there is a step dW = PdV, which can only be said for quasi-equilibrium (or reversible) processes. As such I believe PV^gamma = constant (and the family of equations) should not be applicable to just adiabatic processes? Ie, it should be applicable only for adiabatic + reversible = isentropic processes? However, I've seen couple of online notes/books, and...
I have an engine that uses a dry sump oiling system. The oil collection pan has three AN fittings to use for scavenging. Two of the fittings are approximately on the same level, the third is about 1/2 to 3/4 inch higher than the other two. The system ran for years with no problem using a three stage pump (one pressure and two scavenge stages). The two scavenge stages were connected at times to any two of the three AN fittings on the tank. Recently I tried an upgrade to a four stage pump...