Question about Kepco Power Supply with Error Sensing

AI Thread Summary
The discussion revolves around using a Kepco BOP 20-10M power supply for I-V curve testing of a solar cell, specifically addressing concerns about error sensing and grounding. The user questions the grounding setup, fearing that current may bypass the common wire and go directly to ground, complicating the integration of an ammeter. Additionally, the user highlights the risk of tripping a breaker due to the low resistance of the solar cell when forward biased. Clarification is sought on the purpose of 1K resistors in the setup, which are explained as necessary for maintaining voltage readings on the Sense leads when relays are open, preventing the power supply from outputting maximum voltage. Overall, the conversation emphasizes the importance of proper connections and configurations in power supply applications.
Aaron William
Messages
12
Reaction score
0
Hi, here are a few newb questions about error sensing -

I have a Kepco BOP 20-10M that I have connected to a solar cell. The purpose is to sweep voltage across the cell and create I-V curves.

Attached is a screenshot from the manual showing the setup for my application - using error sensing (4-wire Kelvin measurement). If I interpret the figure correctly, Kepco wants me to ground the load on the common side. My question is: Wouldn't all of the current from the output go to ground and not through the common wire? If this is the case, why have a common wire at all?

This wouldn't work for me because the manufacturer of this system put the ammeter (more accurately, a small resistor that sends a voltage signal to my A/D card) on the common side. I could put it on the output side but I never go against manufacturer wishes :)

Also, the solar cell has very little resistance when forward biased enough. I think I will just trip the breaker when the voltage sweep gets high enough because all the output current is going directly to ground with virtually no resistance.

Second question, the screenshot also shows placing a 1K resistor from output and common to their respective sense when relays are used. I do have two relays that connect, independently, the sense circuit and the current circuit to the load. Can someone explain the purpose of those resistors?
 

Attachments

  • Kepco_BOP_Manual_Screenshot.jpg
    Kepco_BOP_Manual_Screenshot.jpg
    46.7 KB · Views: 580
Last edited:
Engineering news on Phys.org
Aaron William said:
Can someone explain the purpose of those resistors?

The 1k resistors are to connect the Sense leads to the Power leads when the relay is open.

Without the resistors, when the relay is open the Sense leads see Zero volts. At this point the power supply goes to maximum output voltage trying to get a reading on the Sense wires. Not an ideal situation!

When the relay closes, the Sense leads are connected to the load (and supply output) and all is good.
 
While I was rolling out a shielded cable, a though came to my mind - what happens to the current flow in the cable if there came a short between the wire and the shield in both ends of the cable? For simplicity, lets assume a 1-wire copper wire wrapped in an aluminum shield. The wire and the shield has the same cross section area. There are insulating material between them, and in both ends there is a short between them. My first thought, the total resistance of the cable would be reduced...
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...