Wheatstone bridge offset-nulling circuit

AI Thread Summary
A Wheatstone bridge is being used for a cantilever beam measurement system with two 120Ω strain gauges. To null the system before adding weight, the user seeks guidance on incorporating a variable resistor to account for tolerances and sensor response differences. It's suggested to measure the actual resistances of the strain gauges first to determine any discrepancies. Based on the measurements, appropriate fixed resistors can be selected to maintain the same ratio, with a small variable resistor added for fine-tuning. Experimentation is encouraged to achieve optimal nulling results.
nitronick
Messages
1
Reaction score
0
Hi all, I’m making a simple cantilever beam measurement system using 2 strain gages (120Ω).

The two gages make up one leg of my Wheatstone bridge; the other leg will be made up using 120Ω resistors.

Because of tolerances etc I will need to null the system before I add any weight to the end of the beam.

So how is best to go about this, I’ve been looking about on Google but all I can find is to add a variable resistor into the circuit, but I’m not sure where and what value to use.

Any help will be much appreciated.
 
Engineering news on Phys.org
Probably about 2x the tolerance of your devices. You're mostly nulling-out the error, which is the difference in the response of the two sensors and any temperature related fluctuations. But an experiment might be the best way to figure it out...
 
You could start by measuring the actual resistances of the strain gauges. If there is a difference of n% then the two fixed resistors could be picked to have the same ratio. Only a small value of variable resistor would be necessary then. But the numbers would depend on what your measurements tell you.
 
While I was rolling out a shielded cable, a though came to my mind - what happens to the current flow in the cable if there came a short between the wire and the shield in both ends of the cable? For simplicity, lets assume a 1-wire copper wire wrapped in an aluminum shield. The wire and the shield has the same cross section area. There are insulating material between them, and in both ends there is a short between them. My first thought, the total resistance of the cable would be reduced...
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...