Category ratings on test gear (maximum voltage)

AI Thread Summary
The discussion focuses on the meaning of voltage ratings in categories I to IV for test gear, specifically regarding a Cat III rated multimeter at 300V. It clarifies that the voltage rating typically refers to AC RMS levels, suggesting that measuring 230V RMS AC is acceptable. However, it warns against using the multimeter for peak or peak-peak measurements, as those exceed the rated limits. The user expresses caution about relying on the ratings of a UNI-T meter, implying potential concerns about its reliability. Overall, the thread emphasizes the importance of understanding voltage ratings to ensure safe usage of test equipment.
Guineafowl
Messages
865
Reaction score
404
Hi all. What does the voltage rating in each category I/II/III/IV actually mean? For example, my multimeter is Cat III rated to 300V. Is this RMS? Peak? Peak-peak?

Our supply here is 230V RMS AC, but this means the momentary peak is about + and - 330V (relative to neutral/ground) and the peak-peak is of course 660V.

Can I therefore use it to measure mains voltage at a socket? My interpretation is:

1. If RMS, yes.
2. If peak, no.
3. If peak-peak, definitely no.
 
Engineering news on Phys.org
http://content.fluke.com/promotions/promo-dmm/0518-dmm-campaign/dmm/fluke_dmm-chfr/files/safetyguidelines.pdf
 
  • Like
Likes Guineafowl
Nidum said:
http://content.fluke.com/promotions/promo-dmm/0518-dmm-campaign/dmm/fluke_dmm-chfr/files/safetyguidelines.pdf
As far as I can see the voltage rating is DC or AC RMS to ground, i.e. 230V, and implies a higher transient level, so I should be fine. That said, it is only a UNI-T meter so maybe I should take the rating with a pinch of salt...
 
Hi all I have some confusion about piezoelectrical sensors combination. If i have three acoustic piezoelectrical sensors (with same receive sensitivity in dB ref V/1uPa) placed at specific distance, these sensors receive acoustic signal from a sound source placed at far field distance (Plane Wave) and from broadside. I receive output of these sensors through individual preamplifiers, add them through hardware like summer circuit adder or in software after digitization and in this way got an...
I have recently moved into a new (rather ancient) house and had a few trips of my Residual Current breaker. I dug out my old Socket tester which tell me the three pins are correct. But then the Red warning light tells me my socket(s) fail the loop test. I never had this before but my last house had an overhead supply with no Earth from the company. The tester said "get this checked" and the man said the (high but not ridiculous) earth resistance was acceptable. I stuck a new copper earth...
I am not an electrical engineering student, but a lowly apprentice electrician. I learn both on the job and also take classes for my apprenticeship. I recently wired my first transformer and I understand that the neutral and ground are bonded together in the transformer or in the service. What I don't understand is, if the neutral is a current carrying conductor, which is then bonded to the ground conductor, why does current only flow back to its source and not on the ground path...